=msg 501 Meaning: Developer temperature is 2 C too high Reason: -No circulation in developer bath (belt that drives the pumps fell off. Pump is sticking) -Temperature measuring system defective -Temperature sensor 9E1 defective -Temperature card 15GS1 defective -CPU card 14GS1 defective -Triac card 12GS1 defective -Environment temperature too high =msg 502 Meaning: Desired temperature in developer bath not reached. Reason: Temperature error -> heating off, possibly no regeneration -If temperature increase in 12 minutes < 1 K, Triac card 12GS1 defective. -Overtemperature protection 10BE1 switched: Impeller housing of circulation pump clogged due to some kind of sludge Þ clean pump at every preventive maintenance (will take > 1 hour) or flood tanks with cleaning chemicals from time to time. -Heating element defective =msg 503 Meaning: Temperature in fixer is > 41C Reason: No circulation in fixing bath (belt that drives the pumps fell off. Pump is sticking Þ replace pump) Temperature sensor defective CPU board defective Triac board defective =msg 504 Meaning: Desired temperature in fixer not reached (< 25C) Reason: Temperature error -> heating off, possibly no replenishment -Triac card 12GS1 defective -Overtemperature protection 10BE2 switched: Impeller housing of circulation pump clogged due to crystallization Þ clean pump at every preventive maintenance (will take > 1 hour) or flood tanks with cleaning chemicals from time to time. -Heating element defective -Temperature measuring system defective =msg 505 Meaning: The temperature increase in the dryer within 10 minutes is less than 0,5K Reason: One or more dryer bulbs defective Cross flow fan defective Thermal cut-out on the IR heaters activated Temperature sensor or control defective =msg 506 Meaning: Processor drive speed error (91/92GS at CX160 IA; GS4 at CX330) Reason: Motor speed too low resulting from to high torque, possibly -film jam -Defective motor control board -Fuse on motor control baord defective -Fuses on distributor board defective -Distributor board defective -CPU defective =msg 507 Meaning: Developer level not reached Reason: Replenisher pump defective Triac board defective Sensor defective =msg 508 Meaning: Fixer level not reached Reason: Replenisher pump defective Triac baord defective Sensor defective =msg 509 Meaning: Water level not reached Reason: Anti-algae solenoid valve defective water supply solenoid valve defective Triac board defective Sensor defective Water with too little conductivity =msg 511 Meaning: No developer replenishment. Tank empty. Reason: Empty detection responds =msg 512 Meaning: No fixer replenishment. Tank empty. Reason: Empty detection responds. =msg 513 Meaning: Waste tank full Reason: Sensor for "tank full" responds. Reason: Board for level sensors defective. =msg 516 Meaning: Light sensors 804MC / 806MC for processor steamgate endposition didn`t detect endposition: Steamgate fails to open. Reason: -Steamgate drive 850M1 or corresponding board 856GS defective or not connected -No power --> check Si1 on 502 GS -Light sensor 852MC / 806MC not connected or defective -Filmjam in steamgate or jammed steamgate =msg 521 Meaning: Chemix mixer empty Reason: Empty detection responds. =msg 522 Meaning: Exhaust fan out of order. Reason: Fuse 1 A or power supply defective. Cure: Change fuse to 1.25 A or exchange power supply. Reason: Exhaust clogged Reason: Exhaust hose too long. Decrease of pressure to high. Cure: Use shorter hose, bigger diameter or additional fan Test: Take off hose at LR3300P. =msg 523 Meaning: Bottom tray full. Reason: Hoses seem to be clogged. Reason: Anti algae valve defective. =msg 525 Meaning: No pump strokes are counted by microswitch at developer pump after the developer replenisher pump switched on (emergency input is still possible until a level error is indicated) Reason: Case1: Pump motor is turning - Microswitch out of adjustment or defective. Cure: Check the switch in the Easy Service Program: INPUT - SENSORS - REPDSWI - Cable to the microswitch interrupted or loose contact (BU23 on GS30, distributor board). - Cables between GS30, GS2, triac board GS7 or CPU GS5 not properly connected. Cure: Check all cables from GS30 to the GS5 (CPU) of the film processor. Reason: Case 2: Pump motor is not turning: Cure: Check fuse SI9 on GS2. Check triggering on the triac board GS7 LED D2/7. If it is off, either the CPU is defective or the cable connection between triac board and distributor board GS1 is defective. Reason: Case 3: Pump motor defective: Cable to the power supply defective. = msg 526 Meaning: No pump strokes are counted by microswitch at fixer pump after the fixer replenisher pump switched on (emergency input is still possible until a level error is indicated) Reason: Case 1: Pump motor is turning: - Microswitch out of adjustment or defective. Cure: Check the switch in the Easy Service Program: INPUT - SENSORS - REPFSWI - Cable to the microswitch interrupted or loose contact (BU23 on GS30, distributor board). - Cables between GS30, GS2, triac board GS7 or CPU GS5 not properly connected. Cure: Check all cables from GS30 to the GS5 (CPU) of the film processor. Reason: Case 2: Pump motor is not turning: Cure: Check fuse SI1 on GS30. Check triggering on the triac board GS30 LED D5. If it is off, either the CPU or the cable connection between GS30 and CPU GS5 is defective. Reason: Case 3: Pump motor defective: Cable to the power supply defective. =msg 600 Meaning: Laser calibration not successful (This error message causes a startup cycle). This message is only displayed with old calibration boards installed: SN<1036 and <5027. Exception: SN1022, 1024, 1028, 1029, 1031, 1034, 5022. Main reason status 8/96: a)D/A converter on laser control board 134GS defective. Check output of laser control board like described in error 600. Replace D/A converter if output not o.k. b) Mirror #1 polluted. Remark: First of all test if laser beam is present in the film plane with the service program (repair - tests - print engine - AOM test - bright). If beam is not present in the filmplane check if laser tube works: insert a firm piece of paper between laser tube and pre scan cover. No beam present: Reason: 402GS Power supply for laser tube defective Cure: Check fuse 400 Si2. Check voltage 2,5 kV by connecting a spare high voltage power supply 402 GS at the laser tube. Reason: 300EL HE-NE laser defective Cure: Replace HE-NE laser by changing "pre scan unit" Reason: 308GS AOM defective. To check: Service software: AOM bright. a) check output of AOM driver. Peak to peak 15V, 80Mhz with AOM bright. If o.k. ---- >b) b) dismount prescan cover. Put white piece of paper behind AOM. Several small spots should be visible. If not: AOM defective. Cure: Replace pre scan unit. Adjust AOM driver afterwards. Beam present: Reason: 300M1 polarized filter motor or 314GS I/O Bus laser calibration board defective Cure: Check polarized filter motor with service software. Check voltage (8V ripple) for polarized filter motor on 314GS laser calibration board. Remark: Replace 314GS laser calibration board and 322 GS Cal sensor always together Cure: Check fuses Si1/Si2 in 130GS Power Supply Print Engine Cure: Check voltage on 130GS Power Supply Print Engine: +/- 18V DC Reason: 118GS AOM-driver board or 134GS I/O Bus laser control board (D/A converter!) defective. Cure: Check +/-18 V DC and fuses Si1 and Si2 on 130GS power supply print engine Cure: Check +25 V DC and fuse Si3 on 130GS power supply print engine. If fuse is blown check the current supplied to the AOM driver. If the current is higher than 400 mA exchange the AOM driver and adjust the new board. Cure: Check 134 GS I/O Bus laser control board output: - connect a T-piece at video plug on 118GS AOM driver - check voltage with oscilloscope and service program: (repair - tests - printengine - AOM test) AOM offÞ appr. -360mV (+/-60 mV) AOM bright Þ appr. +460mV (+/-60mV) Change D/A converter if voltage is not reached. Cure: Check output (RF) of 118GS AOM driver board with oscilloscope - switch laser imager off - plug in a T-connector at 118GS AOM driver output - switch laser imager on and check voltage with oscilloscope and service program: AOM off Þ appr. 0V, 80 MHz AOM bright Þ appr. 15 V, 80 MHz Reason: 300MG1 lifting magnet, 322GS Cal sensor or 314GS laser calibration board defective. Cure: Check lifting magnet with the service program. Calibrate with the service program to different values up to 254. Remark: Replace 314GS laser calibration board and 322GS cal sensor always together. =msg 601 Meaning: Laser calibration not successful during filmrun (see also error logbook for previous errors!) (This error causes a startup cycle) Main reason status 8/96: a)D/A converter on laser control board 134GS defective. Check output of laser control board like described in error 600. Replace D/A converter if output not o.k.) b) Mirror #1 polluted. See also "error 637". c)For additional info, see ‘’special reasons on LR3300/5200 with polygon mirror’’ below Þ If message is only displayed in the sub menu "error log" in the service menu: Reason: Temporary error during calibration. Laser imager starts a warm up cycle and tries to calibrate again. Cure: If it was a temporary error it was eliminated by the warm up cycle.If it was not a temporary error the next error message in the error logbook will point to the defective part. Þ If message is on the display: Reason: D/A converter on laser control board 134GS defective Cure: Change D/A converter Reason: Exact reason is not recognized by the device software. Cure: Check complete print engine with the service software (tests and adjustments), see error 600. Reason: Wrong message due to a mixup of different tasks at startup. No real hardware error. Disappears after reset only for SW <LTP18XX. Cure: Upgrade device SW to LTP1805 or higher. Þ If message is on the display in combination with error 636 or 637 in the sub menu "error log" in the service menu see error 636 or 637 =msg 601 on LR3300 / LR5200 with polygon mirror LR ‘’high res’’ measures the laser brightness with the BOL (begin of line) sensor. in case, the polygon mirror does not turn, you also get error 601! This has nothing to do with calibration! How to check, whether the polygon motor turns. Remove black cap over green LED on polygon control board wiht some tweezers. Green LED must be on. Reason: Polygon motor does not turn: Polygon or polygon control board defective. Cable not connected. 24 V missing. Cure: Check cables. Exchange polygon or exchange polygon control board. Exchange 24V power supply. Reason: Shutter doesn’ t open. Shutter or laser calibration board defective. Cable to shutter not connected. Cure: Check cabling to shutter, Replace laser calibration board or shutter. =msg 604 Meaning: Image FIFO of FOI board timeout: For 5 minutes no image data were transmitted, despite print order was sent by controller. This is an internal error only. The laser imager gets the print command for this film once more. (This error is displayed in the error ringbuffer and the logbook only) Reason: Problems with FOI Board on MG3000 output (should be rev. E..F). Cure: Check AMDI protocol on both sides: At laser imager and at Controller. Check whether there are intermittent transmission problems. Check cable connection. Replace FOI board (at Controller) if necessary. Reason: Due to SW problems in the Controller the image data could not be sent in time. Upgrade Controller SW (to MeGa3P31 or higher) =msg 605 Meaning: Laser imager transmits three consecutive "lock errors" to the controller. Reason: The oscillating frequency of the galvo is wrong or disturbed. Cure: Replace galvo (replace it always together with 310 GS galvo control board) Cure: Check the + 15V DC on 132GS scan timing board and 310GS galvo control board Cure: Check the + 18V on 130GS (power supply print engine), 134GS (I/O bus laser control board) and 132GS (scan timing board). Cure: Vibration for the period of transmission of data. Temporary fault or permanent fault due to floor vibrations. =msg 606 Meaning: "FOI violation error". This is an internal error only. The laser imager gets the print command for this film once more. (This error is displayed in the error ringbuffer and the logbook only) Reason: Interrupt during image data transmission. Cure: Check FOI cable. Short circuit it at one end and look at AMDI protocol. =msg 607 Meaning: "Start of image too soon" This is an internal error only. The laser imager gets the print command for this film once more. (This error is displayed in the error ringbuffer and the logbook only) Reason: Controller sent already image data, but laser imager was not ready. Problem with synchronization. Image transmission before start of image request. Cure: Check revisions of laser imager and controller FOI boards: Laser Imager: rev. D and higher. Controller: rev. E and higher. Check FOI boards. Short circuit it at one end and look at AMDI protocol. The commands <DC2> and <DC4> should be visible. Replace FOI board. =msg 608 Meaning : "FIFO (of FOI board at laser imager) full at end of image" (=EFI error) This is an internal error only. The laser imager gets the print command for this film once more. (This error is displayed in the error ringbuffer and the logbook only) Reason : Imager exposed the right number of pixels, but there are still some pixels left in the FIFO of the FOI board "data underrun". (either too many data were sent or too less data were read). Cure : Check FOI cable. Short circuit it at one end and look at AMDI protocol. Check FOI boards. The commands <DC2> and <DC4> should be visible with AMDI protocol. Reason : Problems at switching from "border value" to "image info" in case the text feature in the border (terminal: lctrl>manufacturer) is switched on. Only possible as of LTP1405. (How to recognize: There is a small white textline on the top left border diplaying filmtype and so on) Cure : Switch off the textline. Press <ENTER> at display "TERMINAL: >lctrl>manufacturer, so that no manufacturer is entered, and so this feature is switched off. Reason: : in combination with a „ghost image“ (= image seems to be repeated 2 cm on the right): Disturbed BOL (begin of line) signal. Cure: : Upgrade device SW to LTP1805 or higher. =msg 609 Meaning: „EFI_D_UR_ERROR“ = Image data FIFO empty during transmission (data underrun). Reason: FOI transfer speed too low. FOI at LR High Res and MG3000 lower than rev. F (must be ³ F at both sides!) MG3000 FOIXXX11 parameter transfer_mode not on 1 (must be 1). AMDI motherboard not rev. 3 (at least). Monitor version on MG3000 not 4.22 or higher. Cure: Exchange FOI board at LR High Res and at MG3000 to rev. F or higher. Install MON4.22 or higher on MG3000. Change FOI parameter transfer_mode to 1. Check AMDI motherboard: Must be rev. 3 or higher. Reason: : in combination with a „ghost image“ (= image seems to be repeated about 2 cm on the right): Disturbed BOL (begin of line) signal. Cure: : Upgrade device SW to LTP1805 or higher. =msg 610 Meaning: "Select" function of 134GS (laser control board) not detected during I/O bus node check Reason: I/O bus IC "select" on 134GS (laser control board) defective. Cure: Replace laser control board . =msg 611 Meaning: "Timing" function of 134GS (laser control board) not detected during I/O bus node check Reason: I/O bus IC "timing" on 134GS (laser control board) defective. Cure: Replace laser control board. =msg 612 Meaning: "FOI" function of 134GS (laser control board) not detected during I/O bus node check Reason: I/O bus IC "FOI" on 134GS (laser control board) defective. Cure: Replace laser control board. =msg 613 Meaning: "Test pattern" function of 134GS (laser control board) not detected during I/O bus node check Reason: I/O bus IC "test pattern" on 134GS (laser control board) defective. Cure: Replace laser control board. =msg 614 Meaning: 314GS (Laser calibration board) not detected during I/O bus node check Reason: 314GS (Laser calibration board) defective. Cure: Replace Laser calibration board Reason: Cable between 314GS (Laser calibration board) and 126GS (I/O bus distributor) not connected/interrupted. Cure: Connect cable. Check cable for damage. Reason: Coding cannot be red out due to wrong R23. Note: This problem only occurs for cold laser imager and disappears normally after RESET. Cure: Short circuit R23 or replace board =msg 615 Meaning: 04GS (stepper motor board for the scan drive) not detected during I/O bus node check. Reason: 304GS (scan drum drive board) defective. Cure: replace scan drum drive board Reason: Cable between 304GS (scan drum driven board) and 126GS (I/O bus distributor) not connected / interrupted. Cure: Connect cable. Check cable for damage. Reason: Wrong coding of 304GS Cure: Adjust code switch to 2. =msg 616 Meaning: 324GS (magnet control board) for 300MG1 not detected during I/O bus node check Reason: 324GS (magnet control board) defective. Cure: Replace magnet control board Reason: Cable between 324GS (Laser calibration board) and 126GS (I/O bus distributor) not connected / interrupted. Cure: Connect cable. Check cable for damage. =msg 620 Meaning: FPGA on LPE_CTRL-HR cannot be loaded Reason: defective LPE_CTRL-HR Cure: Reset the printer. Check IO Bus cable connection between RK188 and Laser Control board 134GS. Eventually change 134GS (IOB Laser control-HR board) =msg 621 Meaning: Polygon rotates with incorrect speed (most likely it is 0) Reason: Start signal on pin 4 (plug J4; polygon motor controller 310GS) not present. Cure: Check voltage between pin 4 and 7 at J4 on 310 GS. 0V Æ problem of 310GS or polygon (hall feedback sensors!) +5V Æ problem of laser control board 134GS or cable between laser control board and polygon motor controller. Remark:+5V on pin 4 is the signal generated on laser control board to switch off the polygon motor! To switch the motor on , the signal is changed to 0V. Reason: Missing 24V of power supply 106GS Cure: Check presence of 24V at plug J2, polygon motor controller 310GS.
|