*Solution tank Empty | 紧急停止 | Add alkaline washing solution to the reservoir .向桶内加碱性清洗液。 |
Drain Tank Full 排水桶满了 | E.STOP 紧急停止 | Waste liquid volume in the reservoir increased than the fixed level桶内废液增加到比固定水平高时。 |
* Drain Tank Full | | Remove waste liquid.弃去废液。 |
LEVEL SNS 水平系统 Short Sample 样品少 | WARNING 警告 | Sample level sensor could not detect sample level.样品水平传感器不能探测样品水平。 |
*Samp.No.xx[xxx]样品编号 | | Add sample and test again.加样品并再测一次。 |
Short Reagent-1 R1少 | WARNING警告 | R1 level sensor could not detect R1 level. R1水平传感器不能探测R1水平。 |
*Samp.No.xx[xxx]样品编号 | | Add R1 and test again.加R1并再测一次。 |
Short Reagent-2 R2 少 | WARNING警告 | R2 level sensor could not detect R1 level. R2水平传感器不能探测R2水平。 |
*Samp.No.xx[xxx]样品编号 | | Add R2 and test again. 加R2并再测一次。 |
DETECTOR 探测器 | E.STOP | Internal communication error occurred. 发生内部通讯错误 |
Comm. Buffer Full 通讯缓冲满了 | 紧急停止 | “READY”operation after complete stop.If frequently occurs, service call. 完全停机后,作“Ready”,如果反复出现相同报警,联系工程师。 |
Alarm name 报警名称 (Unit/motion)部件/监测 | Level 水平 | Alarm content ,cause and action 报警内容、原因和处理方法 |
H/W | E.STOP | Analyzer power switch ON while PC s ON. PC开机时,分析仪电源键也是开。 |
Reset When running 重复运行时复位 | 紧急停止 | (After power failure or once switch. OFF analyzer)(断电或关机) |
Encoder(STF)error 向电码一样错误 | E.STOP | Sampler rotation trouble occurred.(Caused by electric noise) 样品台转动发生问题(电噪音引起的) |
| 紧急停止 | “READY”operation after complete stop.If frequently occurs, service call. 完全停机后,作“Ready”,如果反复出现相同 报警,联系工程师。 |
H/W Encoder (RCT) error 电码一样的(反应盘)错误 | E.STOP | Reaction tray rotation trouble occurred. (Caused by electric noise) 试剂盘旋转发生问题(由于电噪音)。 |