找回密码
 立即注册
搜索
yeec近年来原创帖合集 本站基础知识下载汇总 yeec网站学习币充值链接 学习中心正式上线

[转帖]CHEMIX-180软件升级指南.

[复制链接]
nybosai 发表于 2008-5-25 15:03 | 显示全部楼层 |阅读模式

注册登录才能更好的浏览或提问。

您需要 登录 才可以下载或查看,没有账号?立即注册

×
 

 

Procedures of updating the software for CHEMIX-180

 

Before Updating

1. Save the current system parameters.

(System Parameters) à (FD) à Save user parameters to FD

2. Print Inventory details (2/2 page) by hard copy, if necessary.

Because after updating the software, the current inventory is all deleted, and it is not possible to cover the parameter saving by FD.

3. Write down the current software version.

(System Parameters) à (System) à  Program version

4. If possible, save the “DB” folder in the “Analyzer” folder to “temp” folder of ”C” drive. In current situation, it's not possible to restore the data results to new version after updating.

(Ctrl +Alt + .dot) à (Power off) à ( Desk top screen) à (Explorer) à (“C” drive) à (Program files) à (Analyzer) à save “DB” folder to “temp” folder

 

Updating

. Return to the Desk top on the Win. NT.

(Ctrl +Alt + .dot) à (Power off) à ( Desk top screen)

 

(Updating the Mechanical Control Software by 2 FDs )

1.       Procedures of FFFTP

 Before the following procedures, do check if the icon of “A/B” is selected on the above side of the screen. “A/B” means being available for both ASCII and BINARY.

( Double crick "FFFTP") à (CA-180) à  (Connect) à (the right side of the screen, "update" folder) à  open (/home/kogata/update)

(Insert the first FD) à ( the left side of the screen,) à  ( "A drive" double click) à ( kogata_01.tar select) à ( "↑" upload) 

à ( kogata_02.tar select) à ( "↑" upload) à (Insert the second FD) à ( the left side of the screen,) à ( kogata_03.tar select) à ( "↑" upload) à 

à ( kogata_04.tar select) à ( "↑" upload) à Completed

 

 

2. Procedures of TELNET

( Double click "TELNET") à login : ( kogata ) à password : ( qnx ) à 

kogata# : ( cd update) à  kogata# : ( kgt_update) à after several seconds completed à  kogata# : ( logout )

3. Confirmation the updating

( Double click "FFFTP") à (CA-180) à  (Connect) à (the right half screen, "update" folder) à  open (/home/kogata/update) à Confirm the 4 executable files being changed

 

(Updating the User Interface Software by CD)

1. Delete the current software

(Start) à ( settings) à ( Add/Remove Programs) à (Next) à ( Program Maintenance) à ( Remove) à completed à ( Finish)

2. Install the new software

( Insert the CD) à Automatically appears the message à ( Input "Analyzer" in "User Name") à (Next) à (Next) à ( Install ) à Completed à ( Finish)

 

Loading the current System Parameters by FD

1. Restart the main unit and the computer

2. Load the System Parameters

(System Parameters) à (FD) à the lowest line "System Parameters ( Former generation) Load parameters" à Completed

3. Restart the main unit and the computer

Confirm the chemistry and the calibration parameters

4. Registration of the reagents on the "inventory" screen, if necessary

( Run Monitor) à ( Inventory ) à ( Page: 2/2 ) à ( Double crick "*" in the column of "Barcode") à ( Enter the number of 8 or 14 digits)


yeec维修网视频培训资料购买链接
BeckmanCoulter DXA系列培训资料
Ortho VITROS 系列培训资料
Ortho enGen_ThermoFisher TCA 实验室自动化系统培训资料
Roche Cobas 实验室自动化系统培训资料
Roche Cobas modular系列分析仪培训资料
Horiba-ABX Yumizen系列培训资料
DiaSorin Liaison系列培训资料
Advia2120培训资料
Inpeco-Aptio系列培训资料
Atellica Solution系列培训资料
Siemens Immunoassay系列培训资料 西门子化学发光系列
SIEMENS Advia系列培训资料 西门子生化系列
Toshiba/Abbott系列培训资料 东芝雅培生化系列
Abbott Architect 系列培训资料 雅培生化化学发光系列
ACL TOP 系列培训资料 沃芬TOP血凝系列
BeckmanCoulter Immunoassay系列培训资料 贝克曼化学发光系列
BeckmanCoulter DXH 系列培训资料 贝克曼DXH血球系列
BeckmanCoulter自动样品处理系统介绍性培训资料 贝克曼前后处理流水线系列
BeckmanCoulter AU系列培训资料 贝克曼AU生化系列
BeckmanCoulter DXC系列培训资料 贝克曼DXC生化系列
LaboSpect003/008/AS 7100/7180分析仪培训资料
Horiba-ABX系列培训资料 Horiba-ABX血球系列
Sysmex 血凝系列培训(CA/CS)
Sysmex 尿液分析系列培训(UF1000/5000/UC3500)
Sysmex 血球系列培训(KX21/POCH/XS/XT/XE)
Sysmex XN系列培训(XN-L/XN1000/XN2000/XN3000/XN9000)
Sysmex HISCL系列培训
可直接淘宝店铺购买https://yeec.taobao.com,或咨询手机/微信:13991827712,QQ:67708237
 

LIXM 发表于 2008-5-25 22:24 | 显示全部楼层

要是中文就好了

123@qwe 发表于 2018-11-13 20:50 | 显示全部楼层
学习的
您需要登录后才可以回帖 登录 | 立即注册

本版积分规则

QQ|申请友链|手机版|小黑屋|加入QQ群|yeec维修网

GMT+8, 2025-4-2 15:31 , Processed in 0.737464 second(s), 34 queries .

Powered by Discuz! X3.5

© 2001-2025 Discuz! Team.

快速回复 返回顶部 返回列表