一台Baxter Flo-Gard 6201 输液泵,开机报错“Common failure F-94”,除了开关机键以外,其他按键均失效。
由于没有维修手册,对这个故障代码一无所知,所以一筹莫展。抱着试试看的态度,在网上搜索了一下,居然大获成功,在国外的一个维修论坛上(www.ebme.co.uk),有人给予了详细的解答。处理过程如下:
How to clear the "Common failure F-94" error? (this is a start-up check-sum error that usually arises by having rubbish in the time and date addresses, due to all system batteries being totally depleted. The solution is to restore time and date to current, or at least valid, values):-
1) Whilst holding down "Panel lock" button at rear of unit, hold down "Stop" and “On/off" together at the front panel. This should bring up "Modify Config" in the top window.
2) Step through config options by pressing "Next" (which should now be lit) to get to "Time".
3) Enter correct time using the number keys.
4) Press green "Pri Start" key to store changed time.
5) Step to "Date", and enter correct date using the number keys.
6) Press green "Pri Start" key to store changed date.
7) Once valid entries have been made, switch off at "On/off".
8) Re-start the unit. Error code should have been cleared!
这段英文简单易懂,我就不翻译了。按照上面的提示一步一步做就可以了,没用两分钟,故障排除了。
既然解决方法来源于网络,本人不敢独吞,特意发在这里与大家共享。互联网真是无敌…… |