找回密码
 立即注册
搜索
yeec近年来原创帖合集 本站基础知识下载汇总 yeec网站学习币充值链接 学习中心正式上线

[转帖]EMO乙醚汽化器拆卸保养图示

[复制链接]
郑振寰 发表于 2010-2-28 13:24 | 显示全部楼层 |阅读模式

EMO

Service Instruction
 
created by Frank Weithöner
  Table of contents:

-
Special Tools
- Disassembling Mixing Chamber
- Assembling Mixing Chamber
- Adjustment Rotor / Level Indicator Unit
- Temperature Compensating Unit
- Filler Unit
- Leakage Test
- Closing Mechanism (Release Valve)

 

Special Tools


Before starting make sure that all tools are available. Besides a set of normal blade srew drivers you will need some special tools.
 
1. For opening the EMO:
    Allen key 3/32'' (or 2.3mm)
2. For leakage test:
    Pressure gauge with plugs
3. Special EMO tools:
    - Setting gauge
    - Tool for taking out springs
    - Tool for taking out the rotor
    - Tool for taking out the filler
    - Tool for rotor counter nut
    - Tool for taking out the filler
4. Alcohol, Vaseline
 

Here the special EMO tools:
 
For calibration a setting gauge is needed. It is a simple piece of bent metal with a diameter of 2.5mm.
An allen key of the same size also works fine.
 
(down the original tool)

For turning the springs out of the pointer such a tool helps you alot. It is an old srew driver (diameter 3mm) with a slit.
If you are experienced you can try to turn out the springs with  strong forceps. In my opinion it is faster and easier to build such a tool than getting out the springs with forceps.

For holding the counter nut of the rotor while adjusting you need a socket with a handle. It's a nice tool but not really necessary. You also can hold a simple 8mm socket with pliers.

This tool helps you to lift out the rotor. Unfortunately it is difficult to make and does not work so well.

By now I have no idea how to create a better tool. Maybe you have an idea...
 
The inner diameter is 28mm.
To grab into the notches of the rotor shaft there must be at least one nose. In this case it is a 5mm bolt placed half in the ring.

This a tool for taking out the filler unit.
It is difficult to make (and to decribe) but you will need it only if you really want to take out the filler...

Do you really want to take out the filler and build it?
Here are the measurements:
 
Take a 1" pipe (4 mm strong) and saw a 9mm cut. Now you need 2 little noses with a hight of approximately 1.5mm. The easiest way to take material from the pipe is with a file and leaving the two noses blank.

The handle is not important but of course it must be higher than the filler.
Disassembling Mixing Chamber


Step 1
 
Clean the whole unit to make sure that no dirt gets inside while working with the open unit.
 
Ether would be perfect, but alcohol also works.

Step 2
 
Remove the handle on top.
(2 screws)

Step 3
 
Remove cover.
(2 screws)

Step 4
 
Lift out the spring.

Step 5
 
DO NOT TOUCH THE 3 SCREWS.

Step 6
 
Remove the screw of the pointer.

It is only a cover screw and hides an allen key srew which you have to remove later.

Step 7
 
Remove the 3 springs. A special tool is needed.
See Tool Section.
 
Even with the special tool it is a very tricky procedure.
Because there is no thread you have to turn the springs anti-clockwise to reduce their diameters while pulling them out.
That sounds easier than it is...

Step 8
 
Now the pointer turns free, but the rotor is still attached.
Turn the lever now to 6%.
Now through the hole of the lever an allen key screw appears (3/32'').
Remove the screw.

Step 9
 
Remove the lid.
(8 screws)
 
Don't forget the screw in the center under the lever.

Step 10
 
Now you have to lift out the rotor.
Loosen the nut with the socket key until the end.
Don't take out the screw completely.

Step 11
 
Now screw the axle in the center clockwise.
The rotor must come out slowly.
 
If the rotor is stuck lubrication around the rotor helps a lot.
Give the oil time to penetrate.
The perfect lubrication is a mixture oil and ether.

Step12
 
The rest you have to do with the special wrench.
Put on the wrench and move the rotor from side to side and lift it simultaneously.
Use the wrench only. Do not take a pair of pliers. You will destroy the shaft.

Step 13
 
Clean the rotor and the wall with ether.
 
For heavy dirt and particles use metal polish and a cloth. NEVER use sand paper or a file. You will create massive leaks!
Step 14
 
Now clean the mixing box chamber and the gasket with ether or alcohol.
 
Check gasket for damages.
Step 15
 
Check the closing mechanism of proper functionality.
 
Warning: Do not press the button too deep. If you do so the gasket will come out of the button and you have to dismantle the whole mechanism to bring it back in place.
If this already has happened see the Closing Mechanism section.
Step 16
 
Lubricate the rotor very slightly with vaseline.

Assembling Mixing Chamber


Step 17
 
Set the rotor back in place.
If this is difficult even with vaseline turn the rotor and try another position. Note the cuttings of the rotor and the inlet and outlet of the chamber and take the original position.

Step 18
 
Screw out the screw (anti-clockwise) and push the rotor down.
The exact positioning of the rotor comes later.

Step 19
 
Clean the mixing chamber.
 
Inspect the vaporizer chamber.
If it is very dirty inside fill in ether, stir up and wash it out through the filler unit.
 
Replace the seal.

Step 20
 
Replace the lid.
(8 screws  and
1 allen key screw with
1 cover screw of the pointer)

Step 21
 
And now the tricky part: Replacing the springs.
Turn the pointer and rotor into the correct position so that the holes for the springs appear.
Now put the springs in. Use the special tool, turn anti-clockwise to reduce the diameter of the springs and press them down.

Step 22
 
Remember, there are no threads so do not turn clockwise! You will open the springs and then the springs get damaged.
If you already have a damaged spring with a little luck and skill you may repair it.
Adjustment Rotor / Level Indicator Unit


Step 23a
 
For adjusting the rotor you first have to take out the level indicator. Therefore you have two options. First the correct one:
 
Just loose the 3 screws
(2-4 turns)
and pull the whole unit out.
That is probably difficult because the unit usually is stuck. So turn it right and left and rock it slightly. Use your hands only - no tools.

Step 23b
 
If it is not possible to take out the whole unit turn out the screws completely and lift only the upper part of the unit out of the EMO.


But watch out for the thin plastic washer.
The washers holt the distant tubes in place. They are not so important but must not fall into the main chamber.
 
Clean and check everything, especially the gusket.
 
Continue with the rotor adjustment.
Step 24
 
The adjustment is made by setting the vertical height of the rotor.
Turn the pointer to approximately '4'. Insert the setting gauge in the slit between the rotor and housing. The rotor is correct ajusted when the setting gauge fits without backlash.
To reduce the slit turn the ajustment screw clockwise. To open the slit loose the screw and press the rotor down.
Step 25
 
After the adjustment secure the screw and pointer by screwing the counter nut tighter. Make sure that the screw is not turning.
 
Replace the cover of the pointer.
(2 screw)
Temperature Compensating Unit

Step 26
 
Remove the Temperature Compensating Unit.
(3 screws)
Be careful. The glass ball is fragile.
Look for dirt and check gasket.
If possible do only a visual inspection - you can destroy more than you can repair.
 
Put it back into its place carefully.
Filler Unit

Step 27
 
Take out the Filler Unit with the special tool.

Step 28
 
Clean everything and check the function.
The valve should move easily.
Check the gasket.
 
Put the unit back into the EMO.
Leakage Test

Step 29
 
For the leakage test you need a pressure gauge and 2 rubber plugs. An old Sphygmomanometer works very well.
 
Note: Before using the test equipment check the pressure gauge itself for leakage.
Step 30
 
Connect pressure gauge on one side and lock the other side with a plug. Set the pointer between 1 and 20. Do not leave the pointer on the release valve.
Pump up the system with approximately 150 mmHg. The pressure must be stable, but a pressure loss of up to 10 mmHg in 10 sec. is acceptable.
If the test is ok press the release valve down. Air must be released now.
Step 31
 
But most likely you will have leakage problems ;-)
 
Do the following:
1. Check your test equipment
2. Screw all screws tight
3. Is the filler closed?
4. Take a closer look to the
    release valve. It is always a
    source of problems.
 
Tip: A soap solution spread out with a brush on all the gaskets is a very good indicator for leakages. Even the smallest leakages will immediately create bubbles.
For the solution just take some drops of dishwashing liquid.
Closing Mechanism (Release Valve)

Step 32
 
The closing mechanism is the most sensitive part of The EMO. If possible leave it in the lid.
 
But if you have problems (probably leakages) here are images that helps you to dismantle and mantle.
Step 33
 
For removing the closing mechanism turn the two screws for 2 or 3 turns only.
 
The third screw is for adjusting.
Do not turn.
Step 33
 
Two little hooks inside will swing in and leave the unit free.
Step 34
 
Clean everything carefully and watch for damages.
The major problems are the gaskets. Clean it with alcohol and put some vaseline on it.
 
Start with reassembling the silver knob. First put the gusket around the knob and insert the knob into the frame from the inside.

Step 35
 
As I mentioned above the biggest problems are the gaskets - the rubber gaskets as well as the little plastic washers under the heads of the crews. They are definitely damaged. If  tightening the screws with force does not help, you may try liquid gasket maker - but not too much.
Work carefully that no gasket maker gets into the chamber.
Step 36
 
If you use gasket maker don't forget to seal the screw heads.
Step 37
 
Remove excessive gasketmaker immediately after assembling and wait until rubber has dried before doing another leakage test.
 
Don't forget to adjust the closing mechanism after reassembling. You do this with the adjustment screw. Check in close and open position.

 

本帖子中包含更多资源

您需要 登录 才可以下载或查看,没有账号?立即注册

×
看贴要回是本分,有问必答是人才,解决问题回贴是公德.
医疗设备维修.维修咨询(请尽可能在论坛提问),协助维修,上门服务.
电话:13991827712

yeec维修网视频培训资料购买链接
BeckmanCoulter DXA系列培训资料
Ortho VITROS 系列培训资料
Ortho enGen_ThermoFisher TCA 实验室自动化系统培训资料
Roche Cobas 实验室自动化系统培训资料
Roche Cobas modular系列分析仪培训资料
Horiba-ABX Yumizen系列培训资料
DiaSorin Liaison系列培训资料
Advia2120培训资料
Inpeco-Aptio系列培训资料
Atellica Solution系列培训资料
Siemens Immunoassay系列培训资料 西门子化学发光系列
SIEMENS Advia系列培训资料 西门子生化系列
Toshiba/Abbott系列培训资料 东芝雅培生化系列
Abbott Architect 系列培训资料 雅培生化化学发光系列
ACL TOP 系列培训资料 沃芬TOP血凝系列
BeckmanCoulter Immunoassay系列培训资料 贝克曼化学发光系列
BeckmanCoulter DXH 系列培训资料 贝克曼DXH血球系列
BeckmanCoulter自动样品处理系统介绍性培训资料 贝克曼前后处理流水线系列
BeckmanCoulter AU系列培训资料 贝克曼AU生化系列
BeckmanCoulter DXC系列培训资料 贝克曼DXC生化系列
LaboSpect003/008/AS 7100/7180分析仪培训资料
Horiba-ABX系列培训资料 Horiba-ABX血球系列
Sysmex 血凝系列培训(CA/CS)
Sysmex 尿液分析系列培训(UF1000/5000/UC3500)
Sysmex 血球系列培训(KX21/POCH/XS/XT/XE)
Sysmex XN系列培训(XN-L/XN1000/XN2000/XN3000/XN9000)
Sysmex HISCL系列培训
可直接淘宝店铺购买https://yeec.taobao.com,或咨询手机/微信:13991827712,QQ:67708237
 

您需要登录后才可以回帖 登录 | 立即注册

本版积分规则

QQ|申请友链|手机版|小黑屋|加入QQ群|注销账号|yeec维修网

GMT+8, 2024-11-22 02:50 , Processed in 0.838011 second(s), 35 queries .

Powered by Discuz! X3.5

© 2001-2024 Discuz! Team.

快速回复 返回顶部 返回列表