找回密码
 立即注册
搜索
yeec近年来原创帖合集 本站基础知识下载汇总 yeec网站学习币充值链接 学习中心正式上线

[转帖]OMV流量计拆卸保养图示

[复制链接]
郑振寰 发表于 2010-2-28 13:28 | 显示全部楼层 |阅读模式

OMV
Service Instruction
 
created by the participants of the
CCBRT Anaesthesia Maintenance Training Course 2008
Step 1

Remove pointer.
(1 screw)
Step 2

Remove label and scale.
(2 screws)
Step 3

Remove cover.
(3 screws)
Step 4

Mark position of inlet or outlet cone.
Also mark the position of the rack.
Step 5

Remove the rack.
(2 brass screws)

Important: Open the 2 counter nuts and
screws from the sides first. Otherwise you will demage the 2 brass screws.
Step 6

Remove inlet cone.
(4 screws)

Remove regulator assembly.
(2 gaskets)
Step 7

Check the regulator and the gaskets.
Clean everything with ether or alcohol.
Step 8

Remove outlet cone.
(4 screws, 1 gasket)
Step 9

Remove slide valve by pushing from outlet end of the housing.
If you have difficulties use a wooden or plastic tool.

Important: No metal tool! Metal will damage the valve's surface.
Step 10

Clean the device with alcohol or ether.
Obstinate dirt can be removed with metal polish, if you wash away the polish afterwards with alcohol carefully.
Before assembling you may lubricate the valve lightly with vaseline.
Step 11

Unscrew the regulator housing.
(4 screws)
Step 12

Remove the regulator housing.
Step 13

Clean the regulator with ether or alcohol.
Check every moving part for smoth motion.
Step 14

Probably you need a new O-ring. If you don't have the original you may take another one. Put a bigger O-ring in and tighten the 4 srews of the clamp ring lighty just to keep the O-ring in position.
Cut it now accurately with a sharp knife. Work carefully, otherwise you will have leakage problems. Put a bit vaseline on the O-ring before inserting the regulator back to the housing.
Step 15

Check the vaporizing chamber and the metal mesh for damages and clean everything.
Step 16

Remove the lever of the halothane insert
mechanism.
(1 screw)
Step 17

Clean the parts.
Step 18

Only in case of a leakage continue dismanteling the rest of the mechanism.
Hold the sleeve and unscrew the inner
screw.

Important: To loose – turn clockwise
Step 19

While assembling be sure that the lever
covers the hole of the white plastic
nozzle properly.
Step 20

After reassembling and tighten all srews
check the pointer setting.
Move the pointer between 50 and 60
and insert the setting gauge through the
outlet.
Now move the pointer to the left. It
should indicate 35 (+/- 1mm).
Step 21

The original setting gauge is shown in
the drawing above.
The same result you get with a modified
4mm allen key. The original diameter is
3.17 mm. But even a normal 3 mm allen key works fine.

Step 22

Finally the leak test. Block outlet connector and connect an air source and a pressure gauge to the inlet.
(An old sphygmomanometer works very well)
Pump air into the system until approximately 200 mmHg. Close the airway. The pressure will fall slowly. The time for falling until
190 mmHg should be at least 30 seconds. Do the test with pointer in 'OFF' and in '3.5' position.
Step 23

Leaks become visible by brushing a little soap water around the connections.
Bubbles will appear.

Note: Around the brass rack on top of the regulator only use ether or alcohol. The slider gets stick after the soap
solution is dried.

Finally some dos and don'ts
 
-  Store and transport the OMV with pointer in 'O' position.
–  Do not expose the vaporizer to full sun.
–  PENLON recommends a maintenance every 6 month.

本帖子中包含更多资源

您需要 登录 才可以下载或查看,没有账号?立即注册

×
看贴要回是本分,有问必答是人才,解决问题回贴是公德.
医疗设备维修.维修咨询(请尽可能在论坛提问),协助维修,上门服务.
电话:13991827712

yeec维修网视频培训资料购买链接
BeckmanCoulter DXA系列培训资料
Ortho VITROS 系列培训资料
Ortho enGen_ThermoFisher TCA 实验室自动化系统培训资料
Roche Cobas 实验室自动化系统培训资料
Roche Cobas modular系列分析仪培训资料
Horiba-ABX Yumizen系列培训资料
DiaSorin Liaison系列培训资料
Advia2120培训资料
Inpeco-Aptio系列培训资料
Atellica Solution系列培训资料
Siemens Immunoassay系列培训资料 西门子化学发光系列
SIEMENS Advia系列培训资料 西门子生化系列
Toshiba/Abbott系列培训资料 东芝雅培生化系列
Abbott Architect 系列培训资料 雅培生化化学发光系列
ACL TOP 系列培训资料 沃芬TOP血凝系列
BeckmanCoulter Immunoassay系列培训资料 贝克曼化学发光系列
BeckmanCoulter DXH 系列培训资料 贝克曼DXH血球系列
BeckmanCoulter自动样品处理系统介绍性培训资料 贝克曼前后处理流水线系列
BeckmanCoulter AU系列培训资料 贝克曼AU生化系列
BeckmanCoulter DXC系列培训资料 贝克曼DXC生化系列
LaboSpect003/008/AS 7100/7180分析仪培训资料
Horiba-ABX系列培训资料 Horiba-ABX血球系列
Sysmex 血凝系列培训(CA/CS)
Sysmex 尿液分析系列培训(UF1000/5000/UC3500)
Sysmex 血球系列培训(KX21/POCH/XS/XT/XE)
Sysmex XN系列培训(XN-L/XN1000/XN2000/XN3000/XN9000)
Sysmex HISCL系列培训
可直接淘宝店铺购买https://yeec.taobao.com,或咨询手机/微信:13991827712,QQ:67708237
 

您需要登录后才可以回帖 登录 | 立即注册

本版积分规则

QQ|申请友链|手机版|小黑屋|加入QQ群|注销账号|yeec维修网

GMT+8, 2024-11-24 13:19 , Processed in 0.685026 second(s), 35 queries .

Powered by Discuz! X3.5

© 2001-2024 Discuz! Team.

快速回复 返回顶部 返回列表