找回密码
 立即注册
搜索
yeec近年来原创帖合集 本站基础知识下载汇总 yeec网站学习币充值链接 学习中心正式上线

[转帖]电刀相关背景知识--英文

[复制链接]
郑振寰 发表于 2010-3-4 12:25 | 显示全部楼层 |阅读模式

Electrosurgery

Electrosurgery is the application of a high-frequency electric current to biological tissue as a means to cut, coagulate, desiccate, or fulgurate tissue.[1][2][3][4][5][6][7] (These terms are used in specific ways for this methodology—see below). Its benefits include the ability to make precise cuts with limited blood loss. Electrosurgical devices are frequently used during surgical operations helping to prevent blood loss in hospital operating rooms or in outpatient procedures. [8]

In electrosurgical procedures, the tissue is heated by an electric current. Although electrical devices may be used for the cauterization of tissue in some applications, electrosurgery is usually used to refer to a quite different method than electrocautery. The latter uses heat conduction from a probe heated by a direct current (much in the manner of a soldering iron), whereas electrosurgery uses alternating current to directly heat the tissue itself.

Often electrosurgery is mistakenly referred to as diathermy. Unlike Ohmic heating by electric current passing through the conductive tissue in conventional electrosurgery, diathermy means dielectric heating, produced by rotation of molecular dipoles in high frequency alternating electric field. This effect is most widely used in microwave ovens which operate at GHz frequencies.

Electrosurgery is commonly used in dermatological, gynecological, cardiac, plastic, ocular, spine, ENT, orthopedic, urological, neuro- and general surgical procedures.

Electrosurgery is performed using an Electrosurgical Generator (also referred to as Power Supply or Waveform Generator) and a handpiece including one or several electrodes, sometimes referred to as an RF Knife. The apparatus when used for coagulation in surgery is still often referred to informally by surgeons as a "Bovie," after the inventor.

Contents

 
  • 1 History
  • 2 Tissue heating by electric current
  • 3 Electrical stimulation of neural and muscle cells
  • 4 Common electrode configurations
  • 5 Electrosurgical modalities
  • 6 Electrosurgical waveforms
  • 7 Prevention of unintended burns in patients
  • 8 Notes
  • 9 See also
  • 10 Manufacturers of Electrosurgical Equipment
  • 11 External links

 History

Development of the first commercial electrosurgical device is credited to Dr. William T. Bovie, who worked on it from 1914 to 1927 while employed at Harvard University[8][9] The first use of an electrosurgical generator in operating room occurred on October 1, 1926. The surgery was performed by Dr. Harvey Williams Cushing.

 Tissue heating by electric current

When voltage is applied across the material it produces electric field which exerts force on charged particles. A flow of free charge carriers – electrons and ions - is called electric current. In metals and semiconductors the charge carriers are primarily electrons, whereas in liquids the charge is carried predominantly by ions. Electrical conduction in biological tissues is primarily due to the conductivity of the interstitial fluids, and thus is predominantly ionic. Transition between the electronic and ionic conduction is governed by electrochemical processes at the electrode-electrolyte interface. Value of electric current, I, is determined by the applied voltage, V, and material’s resistance, R, according to Ohm's law:

Electric current of a constant polarity is referred to as Direct Current (DC). A current of alternating polarity is referred to as Alternating Current (AC). Its frequency is measured in cycles/second or Hertz (Hz).

Current flowing through a resistor causes the generation of Joule heating. In other words, the resistance of the tissue converts the electric energy of the voltage source into heat (thermal energy) which causes the tissue temperature to rise. The deposited electric power (energy per time) can be calculated using:

where P represents the electric power, typically measured in Watts.

In absence of heat conduction, the rate of temperature rise, dT/dt, in a heated object is proportional to the deposited power P, and inversely proportional to its heat capacity, which is in turn proportional to the mass m of the object and its specific heat capacity c:

Larger amount of heat is required to increase the temperature of a heavier object. Thus when heat is generated in a small region of an object, the temperature of that localized region will rise much faster than if the same amount of heat is evenly dispersed over the entire object.

Current density, j is a measure of the concentration of electric current. A higher current density results in a higher concentration of Joule heating. Power density generated by electric current in the material, p is proportional to the square of the current density, and to the material's resistivity, g:

In absence of heat conduction, the rate of local temperature rise is proportional to the power density, p, produced in that region of tissue, and inversely proportional to its specific heat capacity and density ρ [10].

 Electrical stimulation of neural and muscle cells

Neural and muscle cells are electrically-excitable, i.e. they can be stimulated by electric current. In human patients such stimulation may cause acute pain, muscle spasms, and even cardiac arrest. Sensitivity of the nerve and muscle cells to electric field is due to the voltage-gated ion channels present in their cell membranes. Stimulation threshold does not vary much at low frequencies (so called rheobase - constant level). However, the threshold starts increasing with decreasing duration of a pulse (or a cycle) when it drops below a characteristic minimum (so called chronaxie). Typically, chronaxie of neural cells is in the range of 0.1 - 10 ms, so the sensitivity to electrical stimulation (inverse of the stimulation threshold) decreases with increasing frequency in the kHz range and above. (Note that frequency of the alternating electric current is an inverse of the duration of a single cycle). To minimize the effects of muscle and neural stimulation, electrosurgical equipment typically operates in the radio frequency (RF) range of 100 kHz to 5 MHz.

Operation at higher frequencies also helps minimizing the amount of hydrogen and oxygen generated by electrolysis of water. This is especially important consideration for applications in liquid medium in closed compartments, where generation of gas bubbles may interfere with the procedure. For example, bubbles produced during an operation inside an eye may obscure a field of view.

 Common electrode configurations

There are several commonly used electrode configurations or circuit topologies:

In bipolar configuration the voltage is applied to the patient using a pair of similarly-sized electrodes. For example, special forceps, with one tine connected to one pole of the AC generator and the other tine connected to the other pole of the generator. When a piece of tissue is held by the forceps, a high frequency electric current flows from one to the other forceps tine, heating the intervening tissue.

In monopolar configuration the patient lies on top of the return electrode, a relatively large metal plate or a flexible metalized plastic pad which is connected to the return electrode of the AC source. The surgeon uses a pointed probe to make contact with the tissue. The electric current flows from the probe tip, through the body to the return electrode, and then back to the electrosurgical generator. Since electric current spreads from the pointed electrode as it enters the body the current density is rapidly (quadratically) decreasing with distance from the electrode. Since the rate of heating is proportional to the square of current density, the heating occurs in a very localized region, only near the probe tip. On an extremity such as a finger, there is limited cross-sectional area for the return current to spread across, which might result in higher current density and some heating throughout the volume of the extremity.

There is also a common intermediate configuration, when both electrodes are located on the same probe, but the return electrode is much larger than the active one. Since current density is higher in front of the smaller electrode, the heating and associated tissue effects take place only (or primarily) in front of the active electrode, and exact position of the return electrode on tissue is not critical. Sometimes such configuration is called sesquipolar, even though the origin of this term in Latin (sesqui) means a ratio of 1.5 [11].

Relatively low-powered high frequency electrosurgery can be performed on conscious outpatients with no return electrode at all [12]. Operating with no return electrode is possible, because at the very high frequencies and low currents, the self-capacitance of the patient's body (which is between the patient's body and the machine's return potential) is large enough to allow the resulting displacement current to act as a return path. One example of such a machine is called a hyfrecator.

An accidental additional return path through an earth-ground provides a danger of a burn at a site far away from the probe electrode, and for this reason single-electrode devices are used only on conscious patients who would be aware of such complications, and only on carefully insulated tables.

 Electrosurgical modalities

In cutting mode electrode touches the tissue, and sufficiently high power density is applied to vaporize its water content. Since water vapor is not conductive under normal cirumstances, electric current cannot flow through the vapor layer. Energy delivery beyond the vaporization threshold can continue if sufficiently high voltage is applied (> +/-200 V) [10] to ionize vapor and convert it into a conductive plasma. Vapor and fragments of the overheated tissue are ejected, forming a crater [13]. Electrode surfaces intended to be used for cutting often feature a finer wire or wire loop, as opposed to a more flat blade with a rounded surface.

Coagulation is performed using waveforms with lower average power, generating heat insufficient for explosive vaporization, but producing a thermal coagulum instead.

Electrosurgical desiccation occurs when the electrode touches the tissue open to air, and the amount of generated heat is lower than that required for cutting. The tissue surface and some of the tissue more deep to the probe dries out and forms a coagulum (a dry patch of dead tissue). This technique may be used for treating nodules under the skin where minimal damage to the skin surface is desired.

In fulguration mode, the electrode is held away from the tissue, so that when the air gap between the electrode and the tissue is ionized, an electric arc discharge develops. In this approach the burning to the tissue is more superficial, because the current is spread over the tissue area larger than the tip of electrode. [14] Under these conditions, superficial skin charring or carbonization is seen over a wider area than when operating in contact with the probe, and this technique is therefore used for very superficial or protrusive lesions such as skin tags. Ionization of an air gap requires voltage in the kV range.

Besides the thermal effects in tissue, electric field can produce pores in the cellular membranes - a phenomenon called electroporation. This effect may affect cells beyond the range of thermal damage.

 Electrosurgical waveforms

Different waveforms can be used for different electrosurgical procedures. For cutting, a continuous single frequency sine wave is often employed. Rapid tissue heating leads to explosive vaporization of interstitial fluid. If the voltage is sufficiently high (> 400 V peak-to-peak)[10] the vapor sheath is ionized, forming conductive plasma. Electric current continues to flow from the metal electrode through the ionized gas into the tissue. Rapid overheating of tissue results in its vaporization, fragmentation and ejection of fragments, allowing for tissue cutting[10]. In applications of a continuous wave the heat diffusion typically leads to formation of a significant thermal damage zone at the edges of the lesion. Open circuit voltage in electrosurgical waveforms is typically in the range of 300 - 10,000 V peak-to-peak.

Higher precision can be achieved with pulsed waveforms [10][13]. Using bursts of several tens of microseconds in duration the tissue can be cut, while the size of the heat diffusion zone does not exceed the cellular scale. Heat accumulation during repetitive application of bursts can also be avoided if sufficient delay is provided between the bursts, allowing the tissue to cool down [13]. The proportion of ON time to OFF time can be varied to allow control of the heating rate. A related parameter, duty cycle, is defined as the ratio of the ON time to the period (the time of a single ON-OFF cycle). In the terminology of electrical engineering, this process of altering an amplitude of a periodic waveform is called modulation.

For coagulation, the average power is typically reduced below the threshold of cutting. Typically, sine wave is turned ON and OFF in a rapid succession. The overall effect is a slower heating process, which causes tissue to coagulate. In simple coagulation/cutting mode machines, the lower duty cycle typical of coagulation mode is usually heard by the ear as a lower frequency and a rougher tone, than the higher frequency tone typical of cutting mode with the same equipment.

Many modern electrosurgical generators provide sophisticated waveforms with power adjusted in real time, based on changes of the tissue impedance.

 Prevention of unintended burns in patients

For high power surgical uses during anesthesia the monopolar modality relies on a good electrical contact between a large area of the body (typically at least the entire back of the patient) and the return electrode. If contact with the return pad is insufficient, severe burns (3rd degree) can occur in areas of poor contact with the return pad, or with metal objects in contact with Earth-ground serving as an unintended (capacitative) return path.

To prevent unintended burns, the skin is cleaned and a conductive gel is used to enhance contact. Proper electrical grounding practices must be followed in the electrical wiring of the building. It is also recommended to use a newer electrosurgical unit that includes alarms for ground circuit interruption. Grounding pads should always have full contact with the skin and be placed on the same side of the body and close to the body part where the procedure is occurring.

If there is any metal in the body of the patient, the grounding pad is placed on the opposite side of the body from the metal and be placed between the metal and the operation site. This prevents current from passing selectively through metal on the way to ground. For example, for a patient who has had a right sided hip replacement who is scheduled for surgery, the grounding pad is placed on the left side of the body on the lateral side of the lower abdomen, which places the grounding pad between the location of the metal and the surgical site and on the opposite side from the metal. If there is metal on both sides of the body, the grounding pad is placed between the metal and the procedure site when possible. Common grounding pad locations include lateral portions of the outer thighs, abdomen, back, or shoulder blades. [8]

The use of the bipolar option does not require the placement of a grounding pad because the electrical current does not flow through the bulk of the body.

Electrosurgery should only be performed by a physician who has received specific training in this field and who is familiar with the techniques used to prevent burns.

 Notes

  1. ^ Hainer BL, "Fundamentals of electrosurgery", Journal of the American Board of Family Practice, 4(6):419-26, 1991 Nov-Dec.
  2. ^ Electrosurgery for the Skin, Barry L. Hainer M.D., Richard B. Usatine, M.D., American Family Physician (Journal of the American Academy of Family Physicians), 2002 Oct 1;66(7):1259-66.
  3. ^ A Simple Guide to the Hyfrecator 2000 Schuco International (London) Ltd.
  4. ^ Boughton RS, Spencer SK, "Electrosurgical fundamentals", J Am Acad Dermatol, 1987 Apr;16(4):862-7.
  5. ^ Bouchier G, "The fundamentals of electro-surgery. High frequency current generators", Cah Prothese, 1980 Jan;8(29):95-106. In French.
  6. ^ Oringer MJ, "Fundamentals of electrosurgery", J Oral Surg Anesth Hosp Dent Serv, 1960 Jan;18:39-49.
  7. ^ Reidenbach HD, "Fundamentals of bipolar high-frequency surgery", Endosc Surg Allied Technol, 1993 Apr;1(2):85-90.
  8. ^ a b c McCauley, Genard. (2003). “Understanding Electrosurgery.” Bovie Aaron Medical.
  9. ^ Sheldon V. Pollack FRCPC, Alastair Carruthers FRCPC, Roy C. Grekin MD (2000) "The History of Electrosurgery", Dermatologic Surgery 26 (10), 904–908.
  10. ^ a b c d e On Mechanisms of Interaction in Electrosurgery. New Journal of Physics. 10: 123022 (2008).
  11. ^ US Patent 3987795. Electrosurgical devices having sesquipolar electrode structures incorporated therein
  12. ^ see page 6
  13. ^ a b c Electrosurgery with Cellular Precision. IEEE Transactions on Biomedical Engineering, 55(2):838-841 (2008)
  14. ^ Electrosurgery for the Skin. Barry L. Hainer M.D., Richard B. Usatine, M.D., American Family Physician (Journal of the American Academy of Family Physicians), 2002 Oct 1;66(7):1259-66. See illustration.

 See also

  • Cryosurgery
  • Laser surgery
  • Electrocautery
  • Diathermy
  • microwave minimaze procedure

 Manufacturers of Electrosurgical Equipment

  • Aaron Medical Industries, a Bovie Company, manufacturer of the Aaron series of devices.
  • adeor Medical Technologies GmbH, manufacturer of the adeor RF4 Radiofrequency Surgical Unit.
  • Aesculap, a division of B. Braun, manufacturer of the BipoJet Bipolar Product Line, as well as other mono and bipolar devices.
  • Electrosurgical Instruments Bipolar Forceps, Monopolar, Cables, Electrosurgical Pencils.
  • Elliquence, LLC, manufacturer of the Surgi-Max 4.0 MHz Radiofrequency Surgical Unit
  • Ellman International, Inc, manufacturer of the Surgitron Dual Frequency 4.0 MHz device.
  • ERBE Elektromedizin GmbH, manufacturer of the VIO series of devices.
  • Gyrus ENT, manufacturer of the Plasmacision and Plasmaknife devices.
  • MEGADYNE Medical Products, manufacturer of the MEGA 2000 of device.
  • PEAK Surgical,Manufacturer of the PEAK Surgical System: Pulsar generator and PlasmaBlade handprobes.
  • Conmed Corporation, manufacturer of the Conmed Hyfrecator 2000.
  • WEM, manufacturer of the SS-601MCa and many other devices.
  • Valleylab, a division of Tyco Healthcare Group LP, manufacturer of the Force FX and Force EZ devices.

 External links

  • A Simple Guide to the Hyfrecator 2000, Richard J Motley, Schuco International Ltd. a primer for low-powered outpatient dermatological devices, such as the Hyfrecator 2000 device.
  • Electrosurgery for the Skin, Barry L. Hainer M.D., Richard B. Usatine, M.D., American Family Physician (Journal of the American Academy of Family Physicians), 2002 Oct 1;66(7):1259-66.
  • Electrosurgical Generator Testing Online Journal of he Biomedical Engineering Association of Ireland (BEAI), May 1997.
  • Monopolar Electrosurgery Thermal Management System Kristin Cermak, Erica Gorbutt, Brian Schielke, Trey Souchock *Practical Manual of Electrosurgery at The Laser Training Institute of the Professional Medical Education Association, Inc, Ohio.
  • Principles of Electrosurgery at Valleylab, a manufacturer of electrosurgical equipment.
  • Update on Electrosurgery, Judith Lee, Contributing Editor, Outpatient Surgery Magazine, February, 2002.
  • Electrosurgical Instruments Bipolar Forceps, Monopolar, Cables, Electrosurgical Pencils

本帖子中包含更多资源

您需要 登录 才可以下载或查看,没有账号?立即注册

×
看贴要回是本分,有问必答是人才,解决问题回贴是公德.
医疗设备维修.维修咨询(请尽可能在论坛提问),协助维修,上门服务.
电话:13991827712

yeec维修网视频培训资料购买链接
BeckmanCoulter DXA系列培训资料
Ortho VITROS 系列培训资料
Ortho enGen_ThermoFisher TCA 实验室自动化系统培训资料
Roche Cobas 实验室自动化系统培训资料
Roche Cobas modular系列分析仪培训资料
Horiba-ABX Yumizen系列培训资料
DiaSorin Liaison系列培训资料
Advia2120培训资料
Inpeco-Aptio系列培训资料
Atellica Solution系列培训资料
Siemens Immunoassay系列培训资料 西门子化学发光系列
SIEMENS Advia系列培训资料 西门子生化系列
Toshiba/Abbott系列培训资料 东芝雅培生化系列
Abbott Architect 系列培训资料 雅培生化化学发光系列
ACL TOP 系列培训资料 沃芬TOP血凝系列
BeckmanCoulter Immunoassay系列培训资料 贝克曼化学发光系列
BeckmanCoulter DXH 系列培训资料 贝克曼DXH血球系列
BeckmanCoulter自动样品处理系统介绍性培训资料 贝克曼前后处理流水线系列
BeckmanCoulter AU系列培训资料 贝克曼AU生化系列
BeckmanCoulter DXC系列培训资料 贝克曼DXC生化系列
LaboSpect003/008/AS 7100/7180分析仪培训资料
Horiba-ABX系列培训资料 Horiba-ABX血球系列
Sysmex 血凝系列培训(CA/CS)
Sysmex 尿液分析系列培训(UF1000/5000/UC3500)
Sysmex 血球系列培训(KX21/POCH/XS/XT/XE)
Sysmex XN系列培训(XN-L/XN1000/XN2000/XN3000/XN9000)
Sysmex HISCL系列培训
可直接淘宝店铺购买https://yeec.taobao.com,或咨询手机/微信:13991827712,QQ:67708237
 

 楼主| 郑振寰 发表于 2010-3-4 12:27 | 显示全部楼层

Cauterization

The medical practice or technique of cauterization is a medical term describing the burning of part of a body to remove or close off a part of it in a process called cautery, which destroys some tissue[1], in an attempt to mitigate damage, remove an undesired growth, or minimize other potential medical harmful possibilities such as infections, when antibiotics are not available. The practice was once widespread and is still used in remote regions of the world such as central Africa for treatment of wounds. Its utility before the advent of antibiotics was effective on several levels:
   
•  useful in stopping severe blood-loss,
   
•  to close amputations,
   
•  useful in preventing infections, including complications from septicaemia.

Actual cautery is a term referring to the white-hot iron—a metal generally heated only up to a dull red glow—that is applied to produce blisters, to stop bleeding of a blood vessel, and other similar purposes.[2]

The main forms of cauterization used today in the first world are electrocautery and chemical cautery—where both are, for example, prevalent in the removal of unsightly warts. Cautery can also mean the branding of a human, either recreational or forced. Accidental burns can be considered cauterization as well.

Contents

  • 1 History
  • 2 Electrocautery
  • 3 Chemical cautery
  • 4 Nasal cauterization
  • 5 Religious beliefs
  • 6 See also
  • 7 References and notes
  • 8 External links

 History

Hot cauters were applied to tissues or arteries to stop them from bleeding.

Cauterization was used to stop heavy bleeding, especially during amputations. The procedure was simple: a piece of metal was heated over fire and applied to the wound. This would cause tissues and blood to heat rapidly to extreme temperatures in turn causing coagulation of the blood thus controlling the bleeding, at the cost of extensive tissue damage.

Cautery is described in the Hippocratic Corpus.[3] The cautery was employed for almost every possible purpose in ancient times: as a ‘counter-irritant’, as a haemostatic, as a bloodless knife, as a means of destroying tumours, etc.[4] Later, special medical instruments called cauters were used to cauterize arteries. These were first described by Abu al-Qasim al-Zahrawi (Abulcasis) in his Kitab al-Tasrif.[5] Abu al-Qasim al-Zahrawi also introduced the technique of ligature of the arteries as an alternative to cauterization. This method was later improved and used more effectively by Ambroise Paré.

 Electrocautery

Electrocauterization is the process of destroying tissue using heat conduction from a metal probe heated by electric current (much like a soldering iron). The procedure is used to stop bleeding from small vessels (larger vessels being ligated) or for cutting through soft tissue. Unlike electrocautery, electrosurgery is based on generation of heat inside tissue, using electric current passing through the tissue itself.

Electrocauterization is preferable to chemical cauterization because chemicals can leach into neighbouring flesh and cauterize outside of the intended boundaries.[6]

Ultrasonic coagulation and ablation systems are also available.

 Chemical cautery

Many chemical reactions can destroy tissue and some are used routinely in medicine, most commonly for the removal of small skin lesions (i.e. warts or necrotized tissue) or hemostasis. The disadvantages are that chemicals can leach into areas where cauterization was not intended. For this reason, laser and electrical methods are preferable, where practical. Some cauterizing agents are:

  • Silver nitrate: Active ingredient of the lunar caustic, a stick that traditionally looks like a large match-stick. It is dipped into water and pressed onto the lesion to be cauterized for a few moments.
  • Trichloroacetic acid
  • Cantharidin: An extract of the blister beetle that causes epidermal necrosis and blistering; used to treat warts.

 Nasal cauterization

If a person has been having frequent nose bleeds, it is most likely caused by an exposed blood vessel in their nose. Even if the nose is not bleeding at the time, it is cauterized to prevent future bleeding. The different methods of cauterization include burning the affected area with acid, hot metal, lasers, or silver nitrate. Such a procedure is naturally quite painful. Sometimes liquid nitrogen is used as a less painful alternative, though it is less effective. In the few countries that permit the use of cocaine for medicinal purposes, it is occasionally used topically to make this procedure less uncomfortable, cocaine being the only local anesthetic which also produces vasoconstriction, making it ideal for controlling nosebleeds.

 Religious beliefs

Some followers of Islam believe that cauterization is prohibited, as can be found in Sahih Bukhari.[7]

Volume 7, Book 71, Number 584 Narrated Ibn 'Abbas: (The Prophet said), "Healing is in three things: A gulp of honey, cupping, and branding with fire (cauterizing)." But I forbid my followers to use (cauterization) branding with fire."

Volume 7, Book 71, Number 587: Narrated Jabir bin Abdullah: I heard the Prophet saying, "If there is any healing in your medicines, then it is in cupping, a gulp of honey or branding with fire (cauterization) that suits the ailment, but I don't like to be (cauterized) branded with fire."

In spite of the above, several cases of thermal cauterization are seen in some Muslim countries.

 See also

  • Diathermy
  • Singe

 References and notes

  1. ^ "Dictionary definition, retrieved: 2009-03-07.". https://dictionary.reference.com/browse/cautery?qsrc=2888. 
  2. ^ Robinson, Victor, Ph.C., M.D. (editor) (1939). "Actual cautery". The Modern Home Physician, A New Encyclopedia of Medical Knowledge. WM. H. Wise & Company (New York). , page 16.
  3. ^ The Presocratic Influence upon Hippocratic Medicine
  4. ^ Surgical Instruments from Ancient Rome
  5. ^ Mohamed Kamel Hussein (1978), The Concise History of Medicine and Pharmacy (cf. Mostafa Shehata, "The Father Of Islamic Medicine: An International Questionnaire", Journal of the International Society for the History of Islamic Medicine, 2002 (2): 58-59 [58])
  6. ^ See Mr R McElroy for details of various operations and the unintended effects of chemical cauterization
  7. ^ "USC-MSA Compendium of Muslim Texts". University of Southern California. https://www.usc.edu/dept/MSA/fundamentals/hadithsunnah/bukhari/071.sbt.html. Retrieved on 2007-11-12. 

 External links

  • Valleylab division of Tyco Healthcare, explaining the basics of electrosurgery
  • Examples of Cauterizing the Wound in Cinema - Daily Film Dose

本帖子中包含更多资源

您需要 登录 才可以下载或查看,没有账号?立即注册

×
看贴要回是本分,有问必答是人才,解决问题回贴是公德.
医疗设备维修.维修咨询(请尽可能在论坛提问),协助维修,上门服务.
电话:13991827712
xuzhaoxian 发表于 2011-2-17 09:29 | 显示全部楼层
谢谢版主!关注你哦!
您需要登录后才可以回帖 登录 | 立即注册

本版积分规则

QQ|申请友链|手机版|小黑屋|加入QQ群|注销账号|yeec维修网

GMT+8, 2024-11-25 14:13 , Processed in 0.761018 second(s), 35 queries .

Powered by Discuz! X3.5

© 2001-2024 Discuz! Team.

快速回复 返回顶部 返回列表