Flow Status Interpretation on the Cell-Dyn 1200
I. RBC and WBC flow status are contained at the bottom of the Data Review screen (under the Utility menu):
RBC flow status
II. Interpretation of the flow error status is as follows (RBC and WBC flow status are interpreted individually):
|
Flow Errors |
|
|
Cadence Error |
Voltage Error |
|
Status code |
Cadence CV |
Voltage CV |
Voltage delta |
Voltage range |
00 |
|
|
|
|
01 |
|
|
|
X |
02 |
|
|
X |
|
03 |
|
|
X |
X |
04 |
|
X |
|
|
05 |
|
X |
|
X |
06 |
|
X |
X |
|
07 |
|
X |
X |
X |
08 |
X |
|
|
|
09 |
X |
|
|
X |
0A |
X |
|
X |
|
0B |
X |
|
X |
X |
0C |
X |
X |
|
|
0D |
X |
X |
|
X |
0E |
X |
X |
X |
|
0F |
X |
X |
X |
X |
|
|
|
|
|
|
|
|
|
|
Definitions:
Cadence CV: coefficient of variation for RBC or WBC count cadence exceeds 10% or 20% respectively
Voltage CV: coefficient of variation for RBC or WBC electrode voltage cadence exceeds 2% (for both)
Voltage Delta: difference in RBC or WBC electrode voltage between adjacent cadence periods exceeds 2.0V
Voltage Range: absolute RBC or WBC electrode voltage exceeds range of ±30% nominal electrode voltage (Vnom). DC bias of 2.17V is removed from acquired electrode voltage before comparison.
Vnom = Rnom * 1.01mA * [RBC or WBC current set point] / 4096
Rnom = 22.4kW, 22.9kW for RBC, WBC streams respectively
Note: Flow error analysis aborted if electrode voltage does not exceed 1.5V. Cadence CV analysis is aborted if raw cell count does not exceed 2000.
III. See related document on diagnosis of flow error.