找回密码
 立即注册
搜索
yeec近年来原创帖合集 本站基础知识下载汇总 yeec网站学习币充值链接 学习中心正式上线

科华贝克曼生化参数

[复制链接]
郑振寰 发表于 2005-12-2 23:52 | 显示全部楼层 |阅读模式

注册登录才能更好的浏览或提问。

您需要 登录 才可以下载或查看,没有账号?立即注册

×

上海科华临床生化诊断试剂在美国贝克曼(BECKMAN)系列

SYNCHRON CX4/CX5/CX7/CX4△/CX5△/CX5Pro/CX7△/CX9ALX生化分析仪上测定参数(一)

Chemistry Name

ALT

AST

γ-GT

ALP

α-HBDH

α-AMY

LDH

CK

CK-MB

CK-MB()

HDL-C

LDL-C

Test Name

KLT/U31A

KST/30

KGT/36

KLP/35

KBH/57

KAM/10

KLD/33

KCK/32

KCB/40

KCB/40

HDLC

LDLC

Reaction Type

RATE 2

RATE 2

RATE 2

RATE 2

RATE 2

RATE 2

RATE 2

RATE 2

RATE 2

RATE 2

RATE-1

Endpoint 2

Units

U/L

U/L

U/L

U/L

U/L

U/L

U/L

U/L

U/L

U/L

mmol/L

mmol/L

Decimal Precision

×

×

×

×

×

×

×

×

×

×

×.××

×.××

Reaction Direction

NEGATIVE

NEGATIVE

POSITIVE

POSITIVE

NEGATIVE

POSITIVE

POSITIVE

POSITIVE

POSITIVE

POSITIVE

POSITIVE

POSITIVE

Calculation Factor *

10634

10634

2842

6865

12679

7727

12679

20859

12679

10634

*

*

Math Model

LINEAR

LINEAR

LINEAR

LINEAR

LINEAR

LINEAR

LINEAR

LINEAR

LINEAR

LINEAR

LINEAR

LINEAR

No. of Calibrators

0

0

0

0

0

0

0

0

0

0

1

1

Calibrator #

-

-

-

-

-

-

-

-

-

-

#

#

Cal Time Limit (hr)

0

0

0

0

0

0

0

0

0

0

336hr

336hr

Primary Wavelength

340

340

410

410

340

410

340

340

340

340

600

600

Secondary Wavelength

380

380

520

520

380

520

380

380

380

380

700

700

Reagents Primary Inject A

B

150

150

200

200

192

200

192

200

192

200

200

200

75

75

50

50

48

0

48

50

48

[None] vol

[None]vol

[None]vol

Reagents (Secondary Inject) B

[None] vol

[None] vol

[None] vol

[None] vol

[None] vol

[None] vol

[None] vol

[None] vol

[None] vol

[None] vol

67

67

Add Time (sec)

0

0

0

0

0

0

0

0

0

0

656

656

Sample Volume

9

9

20

4

8

4

8

5

8

8

3

3

REAGENT BLANKStart Read

End Read

120

120

200

200

120

200

120

120

120

120

592

592

300

300

300

300

300

300

300

300

300

300

624

624

REACTION : Start Read

End Read

60

60

60

60

60

60

60

120

120

60

360

592

180

180

240

240

180

180

180

240

240

180

520

624

Usable Range :LOWER LIMIT

UPPER LIMIT

0

0

0

0

0

0

0

0

0

0

0.08

0.26

1000

1000

350

700

1000

1200

1000

2000

500

250

3.40

22.44

REAGENT BLANK:Low ABS Limit

High ABS Limit

0.400

0.400

-0.100

-0.100

0.400

-0.100

-0.100

-0.100

-0.100

-0.100

-1.500

-1.500

1.500

1.500

0.700

0.700

1.500

1.000

1.000

1.000

1.000

0.600

1.500

1.500

REACTION : Low ABS Limit

High ABS Limit

0.300

0.300

-1.500

-1.500

0.300

-1.500

-1.500

-1.500

-1.500

-1.500

-1.500

-1.500

1.500

1.500

1.500

1.500

1.500

1.500

1.500

1.500

1.500

1.500

1.500

1.500

SUBSTRATE DEPLETION

Initial Rate

-99.999

-99.999

99.999

99.999

-99.999

99.999

99.999

99.999

99.999

99.999

99.999

99.999

Delta ABS

0.1

0.1

1.5

1.5

1.5

1.5

1.5

1.5

1.5

1.5

1.5

1.5

注*:请用户根据实际情况校准确认。 版号:20030327

上海科华临床生化诊断试剂在美国贝克曼(BECKMAN)系列

SYNCHRON CX4/CX5/CX7/CX4△/CX5△/CX5Pro/CX7△/CX9ALX生化分析仪上测定参数(二)

Chemistry Name

TC

TG

GLU

CREA()

CREA()

BUN

UA

TBIL(重氮)

DBIL(重氮)

CaAR3

IP

TP

ALB

Test Name

KTC/44A

KTG/42B

KGL/06A

KRE/03C

KRE/03C

KBU/05A

KUA/41A

KTB/11A

KDB/12A

KCA

KIP/43A

KTP/07A

KAL/08A

Reaction Type

Endpoint 2

Endpoint 2

Endpoint 2

Endpoint 2

Rate2

RATE 2

Endpoint 2

Endpoint 2

Endpoint 2

Endpoint 2

Endpoint 2

Endpoint 2

Endpoint 2

Units

mmol/L

mmol/L

mmol/L

μmol/L

μmol/L

mmol/L

μmol/L

μmol/L

μmol/L

mmol/L

μmol/L

g/L

g/L

Decimal Precision

×. ××

×. ××

×. ××

×

×

×. ×

×

×. ×

×. ×

×. ×

×. ×

×

×

Reaction Direction

POSITIVE

POSITIVE

POSITIVE

POSITIVE

POSITIVE

NEGATIVE

POSITIVE

POSITIVE

POSITIVE

POSITIVE

POSITIVE

POSITIVE

POSITIVE

Calculation Factor*

*

*

*

*

*

*

*

*

*

*

*

*

*

Math Model

LINEAR

LINEAR

LINEAR

LINEAR

LINEAR

LINEAR

LINEAR

LINEAR

LINEAR

LINEAR

LINEAR

LINEAR

LINEAR

No. of Calibrators

1

1

1

1

1

1

1

1

1

1

1

1

1

Calibrator #

#

#

#

#

#

#

#

#

#

#

#

#

#

Cal Time Limit (hr)

336hr

336hr

336hr

336hr

336hr

336hr

336hr

336hr

336hr

336hr

336hr

336hr

336hr

Primary Wavelength

560

560

520

560

520

340

560

560

560

660

340

560

600

Secondary Wavelength

700

700

700

700

700

410

700

700

700

700

700

700

700

Reagents (Primary Inject) A

B

150

150

150

210

200

180

150

210

210

180

200

200

327

75

75

75

[None] vol

0

60

75

[None] vol

[None] vol

90

0

0

0

Reagents (Secondary Inject) B

[None] vol

[None] vol

[None] vol

70

[None] vol

[None] vol

[None] vol

70

70

[None] vol

[None] vol

[None] vol

[None] vol

Add Time (sec)

0

0

0

620

0

0

0

620

620

0

0

0

0

Sample Volume

3

3

3

7

20

3

6

10

20

3

4

4

3

REAGENT BLANKStart Read

End Read

260

260

260

580

60

180

260

580

580

260

260

120

120

300

300

300

600

180

300

300

600

600

300

300

300

300

REACTION Start Read

End Read

300

300

580

580

60

60

300

580

580

300

300

580

50

360

360

600

600

120

180

360

600

600

360

360

600

70

Usable Range LOWER LIMIT

UPPER LIMIT

0

0

0

0

0

0

0

0

0

0

0

0.00

0.00

25.85

11.30

27.75

8840

1760

36

4720

342

342

3.75

4.0

100.00

60.00

REAGENT BLANKLowABS Limit

High ABS Limit

-0.100

-0.100

-0.100

0.100

-1.500

-1.500

-0.100

-0.100

-0.100

-0.100

0.100

-0.100

-0.100

0.500

0.500

0.500

0.500

1.500

1.500

0.500

0.500

0.500

0.500

0.500

0.500

0.500

REACTIONLow ABS Limit

High ABS Limit

-1.500

-1.500

-1.500

-1.500

-1.500

-1.500

-1.500

-1.500

-1.500

-1.500

-1.500

-1.500

-1.500

1.500

1.500

1.500

1.500

1.500

1.500

1.500

1.500

1.500

1.500

1.500

1.500

1.500

SUBSTRATE DEPLETION

Initial Rate

99.999

99.999

99.999

99.999

99.999

-99.999

99.999

99.999

99.999

99.999

99.999

99.999

99.999

Delta ABS

1.5

1.5

1.5

1.5

1.5

1.5

1.5

1.5

1.5

1.5

1.5

1.5

1.5

注* :标准品浓度,应按每批标示值输入。 版号:20030327


上海科华临床生化诊断试剂在美国贝克曼(BECKMAN)系列

SYNCHRON CX4/CX5/CX7/CX4△/CX5△/CX5Pro/CX7△/CX9ALX生化分析仪上测定参数(三)

Chemistry Name

apoA1(单)

apoA1(双)

apoB(单)

apoB(双)

Lp(a)

Test Name

APOA

APOA

APOB

APOB

LPA

Reaction Type

Endpoint 2

Endpoint 2

Endpoint 2

Endpoint 2

Endpoint 2

Units

g/L

g/L

g/L

g/L

mg/L

Decimal Precision

×.××

×.××

×.××

×.××

×

Reaction Direction

POSITIVE

POSITIVE

POSITIVE

POSITIVE

POSITIVE

Calculation Factor

*

*

*

*

*

Math Model

1

1

1

1

1

No. of Calibrators

5

5

5

5

5

Calibrator # 1

0.00

0.00

0.00

0.00

0

Calibrator # 2

a

a

a

a

a

Calibrator # 3

b

b

b

b

b

Calibrator # 4

c

c

c

c

c

Calibrator # 5

d

d

d

d

d

Primary Wavelength

340

340

340

340

340

Secondary Wavelength

700

700

700

700

700

Sample Volume

3

3

3

3

15

Reagents Primary Inject A

B

300

225

300

225

250

[None]vol

[None]vol

[None]vol

[None]vol

[None]vol

Reagents (Secondary Inject) B

[None]vol

75

[None]vol

75

25

Add Time (sec)

0

620

0

620

620

REAGENT BLANK Start Read

End Read

320

560

320

560

560

336

600

336

600

600

REACTION Start Read

End Read

300

560

300

560

560

340

600

340

600

600

Usable Range LOWER LIMIT

UPPER LIMIT

0.35

0.35

0.25

0.25

0

2.50

2.50

2.40

2.40

1500

REAGENT BLANK Low ABS Limit

High ABS Limit

-1.500

-1.500

-1.500

-1.500

-1.500

1.500

1.500

1.500

1.500

1.500

REACTION Low ABS Limit

High ABS Limit

-1.500

-1.500

-1.500

-1.500

-1.500

1.500

1.500

1.500

1.500

1.500

SUBSTRATE DEPLETION

Initial Rate

99.999

99.999

99.999

99.999

99.999

Delta ABS

1.5

1.5

1.5

1.5

1.5

注* :校准(参考)血清浓度,应按每批标示值输入参数表,定标时应注意校准(参考)血清浓度从低到高依次排列,若更换批号应重新定标。 版号:20030327

上海科华临床生化诊断试剂在美国贝克曼(BECKMAN)系列

SYNCHRON CX4/CX5/CX7/CX4△/CX5△/CX5Pro/CX7△/CX9ALX生化分析仪上测定参数(四)

Chemistry Name

IgG(?)

IgA

IgM

C3

C4

CRP

PA

Test Name

IGG

IGA

IGM

C3

C4

CRP

PA

Reaction Type

Endpoint 2

Endpoint 2

Endpoint 2

Endpoint 2

Endpoint 2

Endpoint 2

Endpoint 2

Units

g/L

g/L

g/L

g/L

g/L

mg/L

mg/L

Decimal Precision

×.×

×.×

×.×

×.×

×.×

×

×

Reaction Direction

POSITIVE

POSITIVE

POSITIVE

POSITIVE

POSITIVE

POSITIVE

POSITIVE

Calculation Factor

*

*

*

*

*

*

*

Math Model

1

1

1

1

1

1

1

No. of Calibrators

5

5

5

5

5

5

5

Calibrator # 1

a

a

a

a

a

a

a

Calibrator # 2

b

b

b

b

b

b

b

Calibrator # 3

c

c

c

c

c

c

c

Calibrator # 4

d

d

d

d

d

d

d

Calibrator # 5

e

e

e

e

e

e

e

Primary Wavelength

700

340

340

340

340

340

340

Secondary Wavelength

-

700

700

700

700

700

700

Sample Volume

3

3

3

3

3

17

7

Reagents Primany Inject A

B

200

200

200

200

200

200

200

[None]vol

[None]vol

[None]vol

[None]vol

[None]vol

[None]vol

[None]vol

Reagents (Secondary Inject)

75

75

50

75

75

50

75

Add Time (sec)

620

620

620

620

620

620

620

REAGENT BLANK Start Read

End Read

560

560

560

560

560

560

560

600

600

600

600

600

600

600

REACTION Start Read

End Read

560

560

560

560

560

560

560

600

600

600

600

600

600

600

Usable Range LOWER LIMIT

UPPER LIMIT

3.00

0.50

0.30

0.10

0.025

0.0

30

50.00

8.24

4.98

3.30

0.70

200.0

800

REAGENT BLANKLow ABS Limit

High ABS Limit

-1.500

-1.500

-1.500

-1.500

-1.500

-1.500

-1.500

1.500

1.500

1.500

1.500

1.500

1.500

1.500

REACTION Low ABS Limit

High ABS Limit

-1.500

-1.500

-1.500

-1.500

-1.500

-1.500

-1.500

1.500

1.500

1.500

1.500

1.500

1.500

1.500

SUBSTRATE DEPLETION

Initial Rate

99.999

99.999

99.999

99.999

99.999

99.999

99.999

Delta ABS

1.5

1.5

1.5

1.5

1.5

1.5

1.5

注* :校准(参考)血清浓度,应按每批标示值输入参数表,定标时应注意校准(参考)血清浓度从低到高依次排列,若更换批号应重新定标。 版号:20030327

看贴要回是本分,有问必答是人才,解决问题回贴是公德.
医疗设备维修.维修咨询(请尽可能在论坛提问),协助维修,上门服务.
电话:13991827712

yeec维修网视频培训资料购买链接
BeckmanCoulter DXA系列培训资料
Ortho VITROS 系列培训资料
Ortho enGen_ThermoFisher TCA 实验室自动化系统培训资料
Roche Cobas 实验室自动化系统培训资料
Roche Cobas modular系列分析仪培训资料
Horiba-ABX Yumizen系列培训资料
DiaSorin Liaison系列培训资料
Advia2120培训资料
Inpeco-Aptio系列培训资料
Atellica Solution系列培训资料
Siemens Immunoassay系列培训资料 西门子化学发光系列
SIEMENS Advia系列培训资料 西门子生化系列
Toshiba/Abbott系列培训资料 东芝雅培生化系列
Abbott Architect 系列培训资料 雅培生化化学发光系列
ACL TOP 系列培训资料 沃芬TOP血凝系列
BeckmanCoulter Immunoassay系列培训资料 贝克曼化学发光系列
BeckmanCoulter DXH 系列培训资料 贝克曼DXH血球系列
BeckmanCoulter自动样品处理系统介绍性培训资料 贝克曼前后处理流水线系列
BeckmanCoulter AU系列培训资料 贝克曼AU生化系列
BeckmanCoulter DXC系列培训资料 贝克曼DXC生化系列
LaboSpect003/008/AS 7100/7180分析仪培训资料
Horiba-ABX系列培训资料 Horiba-ABX血球系列
Sysmex 血凝系列培训(CA/CS)
Sysmex 尿液分析系列培训(UF1000/5000/UC3500)
Sysmex 血球系列培训(KX21/POCH/XS/XT/XE)
Sysmex XN系列培训(XN-L/XN1000/XN2000/XN3000/XN9000)
Sysmex HISCL系列培训
可直接淘宝店铺购买https://yeec.taobao.com,或咨询手机/微信:13991827712,QQ:67708237
 

 楼主| 郑振寰 发表于 2005-12-2 23:54 | 显示全部楼层

上海科华临床生化诊断试剂在美国贝克曼(BECKMAN)LX20全自动分析仪上测定参数(一)上

Test Name

KALT

KAST

KGGT

KALP

KHBDH

KAMY

KLDH

KCK

KCKM

Chemistry Parameters

Reaction Type

RATE-2

RATE-2

RATE-2

RATE-2

RATE-2

RATE-2

RATE-2

RATE-2

RATE-2

Units

U/L

U/L

U/L

U/L

U/L

U/L

U/L

U/L

U/L

Decimal Precision

X

X

X

X

X

X

X

X

X

Reaction Direction

NEGATIVE

NEGATIVE

POSITIVE

POSITIVE

NEGATIVE

POSITIVE

POSITIVE

POSITIVE

POSITIVE

Math Model

LINEAR

LINEAR

LINEAR

LINEAR

LINEAR

LINEAR

LINEAR

LINEAR

LINEAR

Primary Wavelength

340

340

410

410

340

410

340

340

340

Secondary Wavelength

380

380

520

520

380

520

380

380

380

Calculation Factor *

-10634

-10634

2842

6865

-12679

7727

12679

20859

12679

No. of Calibrators

0

0

0

0

0

0

0

0

0

Setpoints 1

0

0

0

0

0

0

0

0

0

Cal Time Limit

0

0

0

0

0

0

0

0

0

Processing Parameters

First Inject Component

A

A

A

A

A

A

A

A

A

Dispense Volume (μl)

150

150

200

200

240

200

240

200

240

Inject Time (s)

-

-

-

-

-

-

-

-

-

Second Inject Component

B

B

B

B

B

NONE

B

B

B

Dispense Volume (μl)

75

75

50

50

60

0

60

50

60

Add Time (s)

-180

-180

-180

-180

-180

-180

-180

-180

-180

Third Inject Component

NONE

NONE

NONE

NONE

NONE

NONE

NONE

NONE

NONE

Dispense Volume (μl)

-

-

-

-

-

-

-

-

-

Add Time (s)

-

-

-

-

-

-

-

-

-

Sample Volume (μl)

9

9

20

4

10

4

10

5

10

注*:请用户根据实际情况校准确认。 版号:20020926

上海科华临床生化诊断试剂在美国贝克曼(BECKMAN)LX20全自动分析仪上测定参数(一)下

Test Name

KALT

KAST

KGGT

KALP

KHBDH

KAMY

KLDH

KCK

KCKM

Reagent Blank Start Read

-80

-80

-80

-80

-80

-80

-80

-80

-80

End Read

-10

-10

-10

-10

-10

-10

-10

-10

-10

Reaction 1 Start Read

60

60

60

60

60

60

60

120

120

End Read

180

200

240

240

180

180

180

240

240

Usable Result Range Lower Limit

0.000

0.000

0.000

0.000

0.000

0.000

0.000

0.000

0.000

Upper Limit

1000.000

1000.000

350.000

700.000

1000.000

1200.000

1000.000

2000.000

500.000

Error Detection Limits

Blank ABS Low Limit

-1.500

-1.500

-1.500

-1.500

-1.500

-1.500

-1.500

-1.500

-1.500

ABS High Limit

2.200

2.200

2.200

2.200

2.200

2.200

2.200

2.200

2.200

Rate Low Limit

2.200

2.200

-1.500

-1.500

-1.500

-1.500

-1.500

-1.500

-1.500

Rate High Limit

-1.500

-1.500

2.200

2.200

2.200

2.200

2.200

2.200

2.200

Mean Deviation

2.200

2.200

2.200

2.200

2.200

2.200

2.200

2.200

2.200

Reaction 1 ABS Low Limit

-1.500

-1.500

-1.500

-1.500

-1.500

-1.500

-1.500

-1.500

-1.500

High Limit

2.200

2.200

2.200

2.200

2.200

2.200

2.200

2.200

2.200

Rate Low Limit

2.200

2.200

-1.500

-1.500

2.000

-1.500

-1.500

-1.500

-1.500

Rate High Limit

-1.500

-1.500

2.200

2.200

-1.500

2.200

2.200

2.200

2.200

Reaction 2 ABS Low Limit

2.200

-1.500

-1.500

-1.500

-1.500

-1.500

-1.500

-1.500

-1.500

High Limit

-1.500

2.200

2.200

2.200

2.200

2.200

2.200

2.200

2.200

Rate Low Limit

2.200

2.200

-1.500

-1.500

2.000

-1.500

-1.500

-1.500

-1.500

Rate High Limit

-1.500

-1.500

2.200

2.200

-1.500

2.200

2.200

2.200

2.200

Mean Deviation

2.200

2.200

2.200

2.200

2.200

2.200

2.200

2.200

2.200

Substrate Depletion Initial Rate

-99.999

-99.999

99.999

99.999

-99.999

99.999

99.999

99.999

99.999

Delta ABS

1.5

1.5

1.5

1.5

1.5

1.5

1.5

1.5

1.5

版号:20020926

上海科华临床生化诊断试剂在美国贝克曼(BECKMAN)LX20全自动分析仪上测定参数(二)上

Test Name

KTC

KTG

KGLU

KCREA()

KBUN

KUA

KCAAR3

KIP

KTP

KALB

KTB()

KDB(重)

Chemistry Parameters

Reaction Type

ENDPOINT2

ENDPOINT2

ENDPOINT2

ENDPOINT2

RATE-2

ENDPOINT2

ENDPOINT2

ENDPOINT2

ENDPOINT2

ENDPOINT2

ENDPOINT2

ENDPOINT2

Units

mmol/L

mmol/L

mmol/L

μmol/L

mmol/L

μmol/L

mmol/L

mmol/L

g/L

g/L

μmol/L

μmol/L

Decimal Precision

X.XX

X.XX

X.X

X.X

X.X

X

X.XX

X.XX

X

X

X.X

X.X

Reaction Direction

POSITIVE

POSITIVE

POSITIVE

POSITIVE

NEGATIVE

POSITIVE

POSITIVE

POSITIVE

POSITIVE

POSITIVE

POSITIVE

POSITIVE

Math Model

LINEAR

LINEAR

LINEAR

LINEAR

LINEAR

LINEAR

LINEAR

LINEAR

LINEAR

LINEAR

LINEAR

LINEAR

Primary Wavelength

560

560

520

560

340

560

340

340

560

600

560

560

Secondary Wavelength

700

700

700

700

410

700

700

410

700

700

700

700

Calculation Factor

*

*

*

*

*

*

*

*

*

*

*

*

No. of Calibrators

1

1

1

1

1

1

1

1

1

1

1

1

Set Points 1

#

#

#

#

#

#

#

#

#

#

#

#

Cal Time Limit (hr)

336

336

336

336

336

336

336

336

336

336

336

336

Processing Parameters

First Inject Component

A

A

A

A

A

A

A

A

A

A

A

A

Dispense Volume (μl)

150

150

150

210

180

150

150

200

300

327

210

210

Inject Time (s)

-

-

-

-

-

-

-

-

-

-

-

-

Second Inject Component

B

B

B

NONE

B

B

B

NONE

NONE

NONE

NONE

NONE

Dispense Volume (μl)

75

75

75

0

60

75

75

0

0

0

0

0

Add Time (s)

-180

-180

-180

0

-180

-180

-180

-180

-180

-180

0

0

Third Inject Component

NONE

NONE

NONE

B

NONE

NONE

NONE

NONE

NONE

NONE

B

B

Dispense Volume (μl)

-

-

-

70

-

-

-

-

-

-

70

70

Add Time (s)

-

-

-

336

-

-

-

-

-

-

336

336

Sample Volume (μl)

3

3

3

7

3

6

3

4

4

3

10

10

注# : 标准品浓度,应按每批标示值输入。 版号:20020926

上海科华临床生化诊断试剂在美国贝克曼(BECKMAN)LX20全自动分析仪上测定参数(二)下

Test Name

KTC

KTG

KGLU

KCREA()

KBUN

KUA

KCAAR3

KIP

KTP

KALB

KTB(重)

KDB(重)

Reagent Blank Start Read

-100

-100

-100

272

-100

-100

-100

-100

-100

-100

272

272

End Read

-10

-10

-10

304

-10

-10

-10

-10

-10

-10

304

304

Reaction 1 Start Read

300

300

300

592

60

300

300

580

580

50

600

592

End Read

360

360

600

624

180

360

360

600

600

70

660

624

Usable Result Range Lower Limit

0

0

0

0

0

0

0

0

0

0

0

0

Upper Limit

25.85

11.30

27.75

8840

36

4720

3.75

4.0

100

60

342

513

Error Detection Limits

Blank ABS Low Limit

-1.500

-1.500

-1.500

-1.500

-1.500

-1.500

-1.500

-1.500

-1.500

-1.500

-1.500

-1.500

ABS High Limit

2.200

2.200

2.200

2.200

2.200

2.200

2.200

2.200

2.200

2.200

2.200

2.200

Rate Low Limit

-1.500

-1.500

-1.500

-1.500

-1.500

-1.500

-1.500

-1.500

-1.500

-1.500

-1.500

-1.500

Rate High Limit

2.200

2.200

2.200

2.200

2.200

2.200

2.200

2.200

2.200

2.200

2.200

2.200

Mean Deviation

2.200

2.200

2.200

2.200

2.200

2.200

2.200

2.200

2.200

2.200

2.200

2.200

Reaction 1 ABS Low Limit

-1.500

-1.500

-1.500

-1.500

-1.500

-1.500

-1.500

-1.500

-1.500

-1.500

-1.500

-1.500

High Limit

2.200

2.200

2.200

2.200

2.200

2.200

2.200

2.200

2.200

2.200

2.200

2.200

Reaction 2 ABS Low Limit

-1.500

-1.500

-1.500

-1.500

-1.500

-1.500

-1.500

-1.500

-1.500

-1.500

-1.500

-1.500

High Limit

2.200

2.200

2.200

2.200

2.200

2.200

2.200

2.200

2.200

2.200

2.200

2.200

Mean Deviation

2.200

2.200

2.200

2.200

2.200

2.200

2.200

2.200

2.200

2.200

2.200

2.200

Substrate Depletion Initial Rate

99.999

99.999

99.999

99.999

-99.999

99.999

99.999

99.999

99.999

99.999

99.999

99.999

Delta ABS

1.5

1.5

1.5

1.5

1.5

1.5

1.5

1.5

1.5

1.5

1.5

1.5

版号:20020926

上海科华临床生化诊断试剂在美国贝克曼(BECKMAN)LX20全自动分析仪上测定参数(三)上

Test Name

apoA1()

apoB()

Lp(a)

IgG(?)

IgM

IgA

C3

C4

CRP

PA

HDL-C

LDL-C

Chemistry Parameters

Reaction Type

ENDPOINT2

ENDPOINT2

ENDPOINT2

ENDPOINT2

ENDPOINT2

ENDPOINT2

ENDPOINT2

ENDPOINT2

ENDPOINT2

ENDPOINT2

RATE-1

ENDPOINT2

Units

g/L

g/L

mg/L

g/L

g/L

g/L

g/L

g/L

mg/L

mg/L

mmol/L

mmol/L

Decimal Precision

X.XX

X.XX

X

X.XX

X.XX

X.XX

X.XX

X.XX

X

X

X.XX

X.XX

Reaction Direction

POSITIVE

POSITIVE

POSITIVE

POSITIVE

POSITIVE

POSITIVE

POSITIVE

POSITIVE

POSITIVE

POSITIVE

POSITIVE

POSITIVE

Math Model

1

1

1

1

1

1

1

1

1

1

LINEAR

LINEAR

Primary Wavelength

340

340

340

700

340

340

340

340

340

340

600

600

Secondary Wavelength

700

700

700

700

700

700

700

700

700

700

700

Calculation Factor

*

*

*

*

*

*

*

*

*

*

*

*

No. of Calibrators

5

5

5

5

5

5

5

5

5

5

2

2

Set Points 1,2,3,4,5

#

#

#

#

#

#

#

#

#

#

#

#

Cal Time Limit (hr)

336

336

336

336

336

336

336

336

336

336

336

336

Processing Parameters

First Inject Component

A

A

A

A

A

A

A

A

A

A

A

A

Dispense Volume (μl)

225

225

250

200

200

200

200

200

200

200

200

200

Inject Time (s)

-

-

-

-

-

-

-

-

-

-

-

-

Second Inject Component

NONE

NONE

NONE

NONE

NONE

NONE

NONE

NONE

NONE

NONE

NONE

NONE

Dispense (μl)

0

0

0

0

0

0

0

0

0

0

0

0

Add Time (s)

0

0

0

0

0

0

0

0

0

0

0

0

Third Inject Component

B

B

B

B

B

B

B

B

B

B

B

B

Dispense (μl)

75

75

25

100

100

50

100

75

50

75

67

67

Add Time (s)

336

336

336

336

336

336

336

336

336

336

336

336

Sample Volume (μl)

3

3

15

3

3

3

3

3

17

7

3

3

版号:20020926

上海科华临床生化诊断试剂在美国贝克曼(BECKMAN)LX-20全自动分析仪上测定参数(三)下

apoA1

(双)

apo B

(双)

Lp(a)

IgG

IgM

IgA

C3

C4

CRP

PA

HDL-C

LDL-C

Reagent Blank Start Read

272

272

272

272

272

272

272

272

272

272

272

272

End Read

304

304

304

304

304

304

304

304

304

304

304

304

Reaction 1 Start Read

592

592

592

592

592

592

592

592

592

592

360

592

End Read

624

624

624

624

624

624

624

624

624

624

520

624

Usable Result Range Lower Limit

0.35

0.25

0

3

0.30

0.50

0.10

0.025

0

30

0.08

0.26

Upper Limit

2.50

2.40

1000

50

2.39

8.24

3.30

0.7

200

800

3.40

22.44

Error Detection Limits

Blank ABS Low Limit

-1.500

-1.500

-1.500

-1.500

-1.500

-1.500

-1.500

-1.500

-1.500

-1.500

-1.500

-1.500

ABS High Limit

2.200

2.200

2.200

2.200

2.200

2.200

2.200

2.200

2.200

2.200

2.200

2.200

Rate Low Limit

-1.500

-1.500

-1.500

-1.500

-1.500

-1.500

-1.500

-1.500

-1.500

-1.500

-1.500

-1.500

Rate High Limit

2.200

2.200

2.200

2.200

2.200

2.200

2.200

2.200

2.200

2.200

2.200

2.200

Mean Deviation

2.200

2.200

2.200

2.200

2.200

2.200

2.200

2.200

2.200

2.200

2.200

2.200

Reaction 1 ABS Low Limit

-1.500

-1.500

-1.500

-1.500

-1.500

-1.500

-1.500

-1.500

-1.500

-1.500

-1.500

-1.500

High Limit

2.200

2.200

2.200

2.200

2.200

2.200

2.200

2.200

2.200

2.200

2.200

2.200

Reaction 2 ABS Low Limit

-1.500

-1.500

-1.500

-1.500

-1.500

-1.500

-1.500

-1.500

-1.500

-1.500

-1.500

-1.500

High Limit

2.200

2.200

2.200

2.200

2.200

2.200

2.200

2.200

2.200

2.200

2.200

2.200

Mean Deviation

2.200

2.200

2.200

2.200

2.200

2.200

2.200

2.200

2.200

2.200

2.200

2.200

Substrate Depletion Initial Rate

99.999

99.999

99.999

99.999

99.999

99.999

99.999

99.999

99.999

99.999

99.999

99.999

Delta ABS

2.200

2.200

2.200

2.200

2.200

2.200

2.200

2.200

2.200

2.200

2.200

2.200

版号:20020926

看贴要回是本分,有问必答是人才,解决问题回贴是公德.
医疗设备维修.维修咨询(请尽可能在论坛提问),协助维修,上门服务.
电话:13991827712
站站站 发表于 2011-8-1 01:50 | 显示全部楼层
谢谢分享!
您需要登录后才可以回帖 登录 | 立即注册

本版积分规则

QQ|申请友链|手机版|小黑屋|加入QQ群|yeec维修网

GMT+8, 2024-12-22 18:28 , Processed in 0.752234 second(s), 34 queries .

Powered by Discuz! X3.5

© 2001-2024 Discuz! Team.

快速回复 返回顶部 返回列表