找回密码
 立即注册
搜索
yeec近年来原创帖合集 本站基础知识下载汇总 yeec网站学习币充值链接 学习中心正式上线

[分享] BeckmanCoulter DXC500AU介绍

[复制链接]
郑振寰 发表于 2024-3-11 13:11 | 显示全部楼层 |阅读模式
这是BeckmanCoulter最新推出的生化分析仪,与DXC 700AU改自AU680一样,DXC 500AU改自AU480。
主要更改有软件界面与DXC 700AU相同,台面盖板大幅度更改,以便在线添加试剂。封闭试剂通道。
这样400速、800速和2000速的全系列生化都改为封闭试剂。至于AU5800是否也改为DXC-AU系列不得而知。不过AU5800已经有封闭试剂版本了。
在厂家官网上,生化分析仪产品介绍中,只有DXC500AU、DXC700AU和AU5800三种机型。而AU480列到了更多栏目,AU680彻底消失。

DxC 500 AU Chemistry Analyzer
System Simplicity for Expedient and Efficient Performance
Meet the DxC 500 AU—an easy-to-use chemistry analyzer that helps you optimize your operations, protect your productivity, and trust your results. Discover how we’re taking our clinical chemistry portfolio’s tradition of excellence to the next level.






Explore DxC 500 AU Features





1. Reliability
We’ve designed reliability into every corner of the DxC 500 AU so you can count on:
•    Exceptional uptime with a high-performing analyzer
•    Consistently accurate results with the highest percentage of high-Sigma-performing assays — including the most Six-Sigma assays across major vendors
•    A solid supply chain that keeps you prepared to run the tests you need
•    Strong regulatory performance with minimal recalls
•    Dependable service with high first-time-fix rates

2. A Menu You Can Count On
Our menu of 120+ assays is made up of high-quality Six Sigma performance assays. We designed this exceptional menu and high-performing analyzer to deliver consistently accurate results with minimal variability, and help detect errors/defects — all with minimal quality control and troubleshooting required.

3. ISE Electrodes
Our ISE electrodes are built to last up to six months or 40,000 samples — longer than any others on the market. That’s long-lasting, stable performance your busy lab can rely on.



4. On-the-Fly Loading
Count on continuous operation with on-the-fly reagent and consumable loading at any point during sample processing.


5. High-Performing Reagents
Expect more from each kit. Our reagent portfolio offers:
•    Long-lasting stability and more tests per kit, meaning less time spent on reagent management and less unused reagent waste in your laboratory
•    Standardized reagents and assays for consistently accurate results across your network.


6. Intuitive Interface
The intuitive interface gives even your newest users the support they need to work with confidence and drive value for your organization with:

• Proactive task indicators
• Step-by-step instructions
• Simplified staff onboarding and training



7. Sample Management
The DxC 500 AU’s state-of-the-art sample handler takes tedious prep work off your plate — just load it and go. With 12 lanes and the flexible DxLAB sample rack (which accommodates various tube and cup sizes), you can continuously load all samples in any open slot — including STAT. The analyzer will automatically prioritize testing appropriately.


Overview
Standardize on Excellence in Clinical Chemistry
Have confidence in accurate results you can trust—from anywhere in your network—with:
  • Standardized user experience for reduced operator variation, training, and troubleshooting
  • Standardized test menus, assay protocols, and reference ranges for equivalent results across platforms
  • Standardized quality control for confident comparisons of assay performance within your lab network and beyond
With this level of standardization, you can count on consistent and accurate results, informed patient care, and strong traceability to common standards.
Reduce Waste and Simplify Your Workflows with Flexible Reagent Storage
The DxC 500 AU’s reagent refrigerator offers 76 positions (and up to 60 individual analytes) to accommodate the various sizes you need—including 15, 30, and 60 mL. With ready-to-use reagents, you can choose the right kit size for your lab’s workload—no prep work required.
Don’t Waste Another Minute Finding the Right Sample Rack for Each Tube
Our unique DxLAB sample racks simplify sample loading of tube and nesting cups for easy processing of calibrators, quality control, and patient testing in a single rack. They accommodate various tube and cup sizes, so you can continuously load them as they arrive in the lab—no need to sort by size.
Busy Shift? Bring It On.
With throughput of up to 800 photometric tests per hour with ISEs, demanding workloads are no match for you. The DxC 500 AU clinical chemistry analyzer empowers low-volume labs to handle testing surges with ease, so you’re always ready to deliver the right results, on time—every time.


Specifications

Main Specifications

Analytical method:
Fully automated, random-access clinical chemistry system with STAT capability

Analytical principles:
Spectrophotometry and potentiometry Assay types

Assay types:
Endpoint, rate, fixed point and indirect ISE

Analytical method:
Colorimetry, turbidimetry, latex agglutination and indirect ISE

Chemistry menu capacity:
>135 preprogrammed, photometrics currently available (including LIH + HbA1c (Thb, HbA1c + HbA1c%) and ISE

Throughput:
400 photometric tests per hour, up to 800 with ISE
ISE maximum test per hour: 200 if ISE only
HbA1c maximum tests per hour: 65 if HbA1c only

Sample types (assay dependent):
Serum, Plasma, Urine, CSF, Whole blood

Sample management/capacity:
168 sample tubes (24 racks of 7 tubes), unique continuous loading feature
Refrigerated STAT carousel: 22 samples can be run simultaneously (cal, QC and routine samples)

Sample container sizes:
Primary and secondary tubes: diameter between 11.5 mm and 16 mm; height between 55 mm and 102 mm; nested micro cups and sample cups

Sample volume:
1.0 μL-25 μL in 0.1 μL increments (assay dependent)

Sample Barcode symbology:
NW7, EAN13, CODE 39, CODE 128, ISBT-128, 2 of 5 standard, 2 of 5 interleaved mixed readable (max. six types at the same time, except if using ISBT-128)

Compartment temperatures:
STAT wheel: 39.2°F to 50°F (4.5°C to 12°C)
Reagent compartment: 39.2°F to 50°F (4.5°C to 12°C)

Reagent storage capacity:
76 positions in a single reagent carousel (refrigerated between 39.2°F (4°C) and 53.6°F (12°C); bottle sizes: 15 mL, 30 mL, 60 mL

Barcoded reagents:
Automatic tracking of: Available tests, Expiration date, Lot number

Reagent volume:
10 μL-250 μL (1 μL increments)

Total reaction volume:
90 μL-350 μL

Reaction cuvette:
Permanent glass cuvettes

Reaction time:
Up to 8 minutes, 37.5 seconds

Reaction temperature:
98.6°F ± 0.5°F (37°C ±0.3°C)

Reaction incubation method:
Dry bath

Photometric range:
0 OD-3.0 OD

Wavelength:
13 different wavelengths between 340 nm and 800 nm

Automated sample pre-dilution:
Repeat run with increased or decreased sample volume or sample pre-dilution (1, 3, 5, 10, 15, 20, 25, 50, 75, up to 100 times)

Sample quality analysis:
Lipemia, hemolysis and icterus indices
Clot detection and probe crash protection

Calibration:
Calibration curve stability up to 90 days (assay dependent)
Calibrator value assignment by 2D barcode
Calibration curves and parameters displayed on screen

Quality control:
Shewhart day-to-day management (Levey-Jennings method)
Multi-rule control (Westgard method)

Reflex testing:
User-defined

Calculated result:
Pre-defined for the commonly used calculations
User-defined

Installation Requirements:

Analyzer with monitor
Weight: Max 992 pounds (450 kg)
Length: 6.9 ft (2100 mm)
Height: 5.3 ft (1602 mm)
Depth: 2.53 ft (770 mm)

Dimensions and weight
Clearance Requirements:
Clearance around the analyzer for ventilation
Rear: Minimum of 20 in (500 mm)
Clearance around the analyzer for service
Rear: Minimum of 20 in (500mm)
Left: 20 in (500 mm)
Right: Minimum of 25 in (500 mm)
Top: Minimum of 0 in (0 mm)
Front: Minimum of 24.7 in (635 mm)

Drain requirements

Drain Location
Maximum distance 33 ft (10 m) from the system
Maximum height 5 ft (1.5 m) from the floor

Drain Volume
Concentrated waste: 5.5 L hour (normal), 7.5 L hour HbA1c
Diluted waste: 12L/hour

Water supply requirements:
Maximum water consumption: 20 L hour
Water type: Deionized CAP type II or better, bacteria free
Continuous flow supply
Resistivity: > 0.5 MΩ∙cm filtered with a 0.5 μm filter

Power and environmental requirements:
Power Supply: 200V, 208V, 220V, 230V, 240V, 60 Hz

Computer
CPU Intel Core i7-6700 processor
Memory: 32GB DDR4
Storage: 512GB SSD
Operating system: Windows 10
External communication types: RS232 serial interface
Ethernet: LAN using TCP/IP

Monitor
Touchscreen display: 21.5 in diagonal TFT LCD with projected capacitive touch

Interface
Pixels: 1920 x 1080 RGB

Operating environment:
Ambient temperature: 64°F to 90°F (18°C to 32°C)
Humidity: 20% to 80% RH (without condensation)
Altitude: Maximum 6,561 ft (2,000 m) above sea level
Generated Noise Level: <60 dB

Communication modes:
Unidirectional, bi-directional, bi-directional with true host query
LAN connection to LIS per ASTM protocol


本帖子中包含更多资源

您需要 登录 才可以下载或查看,没有账号?立即注册

×
看贴要回是本分,有问必答是人才,解决问题回贴是公德.
医疗设备维修.维修咨询(请尽可能在论坛提问),协助维修,上门服务.
电话:13991827712

yeec维修网视频培训资料购买链接
BeckmanCoulter DXA系列培训资料
Ortho VITROS 系列培训资料
Ortho enGen_ThermoFisher TCA 实验室自动化系统培训资料
Roche Cobas 实验室自动化系统培训资料
Roche Cobas modular系列分析仪培训资料
Horiba-ABX Yumizen系列培训资料
DiaSorin Liaison系列培训资料
Advia2120培训资料
Inpeco-Aptio系列培训资料
Atellica Solution系列培训资料
Siemens Immunoassay系列培训资料 西门子化学发光系列
SIEMENS Advia系列培训资料 西门子生化系列
Toshiba/Abbott系列培训资料 东芝雅培生化系列
Abbott Architect 系列培训资料 雅培生化化学发光系列
ACL TOP 系列培训资料 沃芬TOP血凝系列
BeckmanCoulter Immunoassay系列培训资料 贝克曼化学发光系列
BeckmanCoulter DXH 系列培训资料 贝克曼DXH血球系列
BeckmanCoulter自动样品处理系统介绍性培训资料 贝克曼前后处理流水线系列
BeckmanCoulter AU系列培训资料 贝克曼AU生化系列
BeckmanCoulter DXC系列培训资料 贝克曼DXC生化系列
LaboSpect003/008/AS 7100/7180分析仪培训资料
Horiba-ABX系列培训资料 Horiba-ABX血球系列
Sysmex 血凝系列培训(CA/CS)
Sysmex 尿液分析系列培训(UF1000/5000/UC3500)
Sysmex 血球系列培训(KX21/POCH/XS/XT/XE)
Sysmex XN系列培训(XN-L/XN1000/XN2000/XN3000/XN9000)
Sysmex HISCL系列培训
可直接淘宝店铺购买https://yeec.taobao.com,或咨询手机/微信:13991827712,QQ:67708237
 

秦时明月 发表于 2024-3-13 16:49 | 显示全部楼层
多谢郑工分享
陶望三 发表于 2024-3-25 13:50 | 显示全部楼层
多谢分享
蔡锦涛 发表于 2024-7-7 10:08 | 显示全部楼层
多谢分享
您需要登录后才可以回帖 登录 | 立即注册

本版积分规则

QQ|申请友链|手机版|小黑屋|加入QQ群|yeec维修网

GMT+8, 2025-3-31 17:57 , Processed in 6.702597 second(s), 35 queries .

Powered by Discuz! X3.5

© 2001-2025 Discuz! Team.

快速回复 返回顶部 返回列表