Sample probe of AU680 can detect liquid level but after aspirated plasma the Red Blood Cell income to mix with plasma.
we checked ground pass, check liquid level detect ok, no any bubble in the sample and dispenser syringe. exchange with new sample probe, exchange MV3982 for sample probe and S Wash.
No any error display on the screen.
Pls help!!!!
Thank you, I have checked the sequence of S dispenser. After aspirate plasma to needle, Sample Dispenser will push out a little plasma to sample tube. That is the cause of the above phenomenon. We changed the needle, changed the sample tube. Everything was ok after the change.