FCN 2.28 000 SOFTWARE ERROR [F,T,D,S] Please refer to the description of FCN 1.06 000 FCN 2.28 001 CONDUCTIVITY PUMP TIMER ERROR [F,T,D,S] Transient error. (No safety action) Unconditional error if more than three errors during one treatment.
General description of conditions for occuring: The conductivity pump timer is not working properly. Technical description of conditions for occuring: The timer 9513 has returned one of the following error messages: "Timer load to counter error", "Load register error", "Mode register error" or "Hold register error".
Some possible explanations why the error has occured: * The timer IC 2 on the expansion board (K14716) is broken. Automatical safety actions: Fluid unit stop (Please refer to "Descriptions of safety actions" below)
FCN 2.28 002 CONDUCTIVITY HARDWARE PRESENT WHEN MM90 NOT INSTALLED [F] Unconditional error.
General description of conditions for occuring: The conductivity pump hardware is present even if the MM90 installed preset is set to FALSE. Technical description of conditions for occuring: The preset EL(EEL_MM90$INSTALLED)=FALSE, which means that the MM90 hardware NOT should be installed. The program tried to initialize the conductivity pump timer which should NOT have been present. The attempt to initialize the conductivity pump timer IC 2 on the expansion board (K14716) did NOT fail.
Some possible explanations why the error has occured: * The MM90 hardware is installed, but the preset EL(EEL_MM90$INSTALLED) has incorrectly the value FALSE.
Automatical safety actions: Fluid unit stop (Please refer to "Descriptions of safety actions" below)
Descriptions of safety actions: Blood unit stop The blood pump is stopped. The arterial and venous clamps are closed. Fluid unit stop The heater is turned off. INVA, DIVA and BYVA are closed. The flow pump and suction pump are stopped. The valve and heater drivers are disabled. Safety relay off The 24V supply is disconnected from the blood pump, the heparin pump, the heater relay and the valves. The arterial and venous clamps are closed.
FCN229.PLM Conductivity supervision
'FCN 2.29 000' - Case error conductivity supervision function Unconditional error. No blood pump enable. Internal software error. FCN230.PLM Remaining disinfection time calculation.
'FCN 2.30 000' - Case error remaining didinfection time calculation. Unconditional error. No blood pump enable. Internal software error.
FCN231.PLM BiCart valve function(FLVA).
'FCN 2.31 000' - Case error BiCart valve function Unconditional error. No blood pump enable. Internal software error.
Function group 3 : OPCOM functions.
FCN302.PLM Keyboard
'FCN 3.02 000' - Case error Keyboard function. Unconditional error. No blood pump enable. Internal software error. 'FCN 3.02 001' - Spurreous interrupt. Error caused by keyboard controler IC 87, interrupt controler IC 79 or noise disturbance. Transient error. If more than 3 during one treatment => Unconditional error. Blood pump enabled.
FCNBAR.PLM Bargraphs
'FCN 3.03 000' - Case error Venous pressure bargraph. Unconditional error. No blood pump enable. Internal software error. 'FCN 3.04 000' - Case error Blood flow bargraph. Unconditional error. No blood pump enable. Internal software error. 'FCN 3.05 000' - Case error Conductivity bargraph. Unconditional error. No blood pump enable. Internal software error. 'FCN 3.06 000' - Case error Temperature bargraph. Unconditional error. No blood pump enable. Internal software error. 'FCN 3.07 000' - Case error Extra pressure bargraph. Unconditional error. No blood pump enable. Internal software error. 'FCN 3.08 000' - Case error TMP bargraph. Unconditional error. No blood pump enable. Internal software error. 'FCN 3.09 000' - Case error UF-volume bargraph. Unconditional error. No blood pump enable. Internal software error. 'FCN 3.10 000' - Case error UF-rate bargraph. Unconditional error. No blood pump enable. Internal software error.
FCNLMP.PLM Lamps All lamps broken or shortcircuit errors are minor errors. No stop of treatment.
'FCN BGPH 001' - Backlight lamp broken (Ven.-press & Blood flow bargraph) 'FCN BGPH 002' - Backlight lamp shortage (Ven.-press & Blood flow bargraph) 'FCN BGPH 011' - Backlight lamp broken (Cond. & Temp bargraph) 'FCN BGPH 012' - Backlight lamp shortage (Cond. & Temp bargraph) 'FCN BGPH 021' - Backlight lamp broken (TMP & UF-vol bargraph) 'FCN BGPH 022' - Backlight lamp shortage (TMP & UF-vol bargraph) 'FCN BGPH 031' - Backlight lamp broken (UF-rate & Extra press bargraph) 'FCN BGPH 032' - Backlight lamp shortage (UF-rate & Extra press bargraph)
'FCN 3.11 000' - Case error ON/OFF-lamp. Unconditional error. No blood pump enable. Internal software error. 'FCN 3.11 001' - Lamp broken 'FCN 3.11 002' - Shortage ( High load ).
'FCN 3.12 000' - Case error Disinfection/Rinse lamp. Unconditional error. No blood pump enable. Internal software error. 'FCN 3.12 001' - Lamp broken 'FCN 3.12 002' - Shortage ( High load ).
'FCN 3.13 000' - Case error Reset lamp. Unconditional error. No blood pump enable. Internal software error. 'FCN 3.13 001' - Lamp broken 'FCN 3.13 002' - Shortage ( High load ).
'FCN 3.14 000' - Case error Single needle mode lamp. Unconditional error. No blood pump enable. Internal software error. 'FCN 3.14 001' - Lamp broken 'FCN 3.14 002' - Shortage ( High load ).
'FCN 3.15 000' - Case error Isol-UF mode lamp. Unconditional error. No blood pump enable. Internal software error. 'FCN 3.15 001' - Lamp broken 'FCN 3.15 002' - Shortage ( High load ).
'FCN 3.16 000' - Case error UF mode lamp. Unconditional error. No blood pump enable. Internal software error. 'FCN 3.16 001' - Lamp broken 'FCN 3.16 002' - Shortage ( High load ).
'FCN 3.17 000' - Case error Venous pressure lamp. Unconditional error. No blood pump enable. Internal software error. 'FCN 3.17 001' - Lamp broken 'FCN 3.17 002' - Shortage ( High load ).
'FCN 3.18 000' - Case error Blood flow lamp. Unconditional error. No blood pump enable. Internal software error. 'FCN 3.18 001' - Lamp broken 'FCN 3.18 002' - Shortage ( High load ).
'FCN 3.19 000' - Case error Single needle limits lamp. Unconditional error. No blood pump enable. Internal software error. 'FCN 3.19 001' - Lamp broken 'FCN 3.19 002' - Shortage ( High load ).
'FCN 3.20 000' - Case error Conductivity lamp. Unconditional error. No blood pump enable. Internal software error. 'FCN 3.20 001' - Lamp broken 'FCN 3.20 002' - Shortage ( High load ).
'FCN 3.21 000' - Case error Temperature lamp. Unconditional error. No blood pump enable. Internal software error. 'FCN 3.21 001' - Lamp broken 'FCN 3.21 002' - Shortage ( High load ).
'FCN 3.22 000' - Case error Heparin lamp. Unconditional error. No blood pump enable. Internal software error. 'FCN 3.22 001' - Lamp broken 'FCN 3.22 002' - Shortage ( High load ).
'FCN 3.23 000' - Case error Time lamp. Unconditional error. No blood pump enable. Internal software error. 'FCN 3.23 001' - Lamp broken 'FCN 3.23 002' - Shortage ( High load ).
'FCN 3.24 000' - Case error Extra pressure lamp. Unconditional error. No blood pump enable. Internal software error. 'FCN 3.24 001' - Lamp broken 'FCN 3.24 002' - Shortage ( High load ).
'FCN 3.25 000' - Case error TMP lamp. Unconditional error. No blood pump enable. Internal software error. 'FCN 3.25 001' - Lamp broken 'FCN 3.25 002' - Shortage ( High load ).
'FCN 3.26 000' - Case error UF-volume lamp. Unconditional error. No blood pump enable. Internal software error. 'FCN 3.26 001' - Lamp broken 'FCN 3.26 002' - Shortage ( High load ).
'FCN 3.27 000' - Case error UF-rate lamp. Unconditional error. No blood pump enable. Internal software error. 'FCN 3.27 001' - Lamp broken 'FCN 3.27 002' - Shortage ( High load ).
'FCN 3.28 000' - Case error Air detector lamp. Unconditional error. No blood pump enable. Internal software error. 'FCN 3.28 001' - Lamp broken 'FCN 3.28 002' - Shortage ( High load ).
'FCN 3.29 000' - Case error Arterial pressure lamp. Unconditional error. No blood pump enable. Internal software error. 'FCN 3.29 001' - Lamp broken 'FCN 3.29 002' - Shortage ( High load ).
'FCN 3.30 000' - Case error Attention lamp. Unconditional error. No blood pump enable. Internal software error. 'FCN 3.30 001' - Lamp broken 'FCN 3.30 002' - Shortage ( High load ).
'FCN 3.31 000' - Case error Technical error lamp. Unconditional error. No blood pump enable. Internal software error. 'FCN 3.31 001' - Lamp broken 'FCN 3.31 002' - Shortage ( High load ).
'FCN 3.32 000' - Case error Inlet valve lamp. Unconditional error. No blood pump enable. Internal software error. 'FCN 3.32 001' - Lamp broken 'FCN 3.32 002' - Shortage ( High load ).
'FCN 3.33 000' - Case error Blood leak lamp Unconditional error. No blood pump enable. Internal software error. 'FCN 3.33 001' - Lamp broken 'FCN 3.33 002' - Shortage ( High load ).
'FCN 3.34 000' - Case error Blood pump lamp. Unconditional error. No blood pump enable. Internal software error. 'FCN 3.34 001' - Lamp broken 'FCN 3.34 002' - Shortage ( High load ).
'FCN 3.35 000' - Case error Alarms bypass lamp. Unconditional error. No blood pump enable. Internal software error. 'FCN 3.35 001' - Lamp broken 'FCN 3.35 002' - Shortage ( High load ).
'FCN 3.36 000' - Case error Mute lamp. Unconditional error. No blood pump enable. Internal software error. 'FCN 3.36 001' - Lamp broken 'FCN 3.36 002' - Shortage ( High load ).
'FCN 3.37 000' - Case error Fluid bypass lamp. Unconditional error. No blood pump enable. Internal software error. 'FCN 3.37 001' - Lamp broken 'FCN 3.37 002' - Shortage ( High load ).
'FCN 3.38 000' - Case error Start stop lamp. Unconditional error. No blood pump enable. Internal software error. 'FCN 3.38 001' - Lamp broken 'FCN 3.38 002' - Shortage ( High load ).
'FCN 3.39 000' - Case error Hold lamp. Unconditional error. No blood pump enable. Internal software error. 'FCN 3.39 001' - Lamp broken 'FCN 3.39 002' - Shortage ( High load ).
'FCN 3.40 000' - Case error Select lamp. Unconditional error. No blood pump enable. Internal software error. 'FCN 3.40 001' - Lamp broken 'FCN 3.40 002' - Shortage ( High load ).
'FCN 3.62 000' - Case error Priming lamp. Unconditional error. No blood pump enable. Internal software error. 'FCN 3.62 001' - Lamp broken 'FCN 3.62 002' - Shortage ( High load ).
'FCN 3.41 000' - Case error Blood flowpath lamp Unconditional error. No blood pump enable. Internal software error. 'FCN 3.41 001' - Lamp 46 broken 'FCN 3.41 002' - Lamp 46 shortage ( High load ). 'FCN 3.41 011' - Lamp 47 broken 'FCN 3.41 012' - Lamp 47 shortage ( High load ). 'FCN 3.41 021' - Lamp 48 broken 'FCN 3.41 022' - Lamp 48 shortage ( High load ). 'FCN 3.41 031' - Lamp 49 broken 'FCN 3.41 032' - Lamp 49 shortage ( High load ).
'FCN 3.42 000' - Case error Fluid flowpath lamp. Unconditional error. No blood pump enable. Internal software error. 'FCN 3.42 001' - Lamp 53 broken 'FCN 3.42 002' - Lamp 53 shortage ( High load ). 'FCN 3.42 011' - Lamp 54 broken 'FCN 3.42 012' - Lamp 54 shortage ( High load ). 'FCN 3.42 021' - Lamp 55 broken 'FCN 3.42 022' - Lamp 55 shortage ( High load ). 'FCN 3.42 031' - Lamp 56 broken 'FCN 3.42 032' - Lamp 56 shortage ( High load ). 'FCN 3.42 041' - Lamp 57 broken 'FCN 3.42 042' - Lamp 57 shortage ( High load ). 'FCN 3.42 051' - Lamp 58 broken 'FCN 3.42 052' - Lamp 58 shortage ( High load ). 'FCN 3.42 061' - Lamp 50 broken 'FCN 3.42 062' - Lamp 50 shortage ( High load ). 'FCN 3.42 071' - Lamp 51 broken 'FCN 3.42 072' - Lamp 51 shortage ( High load ). 'FCN 3.42 081' - Lamp 52 broken 'FCN 3.42 082' - Lamp 52 shortage ( High load ).
'FCN 3.67 000' - Case error Auto priming lamp. Unconditional error. No blood pump enable. Internal software error. 'FCN 3.67 001' - Lamp broken 'FCN 3.67 002' - Shortage ( High load ).
'FCN 3.73 000' - Case error Acetate mode lamp. Unconditional error. No blood pump enable. Internal software error. 'FCN 3.73 001' - Lamp broken 'FCN 3.73 002' - Shortage ( High load ).
'FCN 3.74 000' - Case error Bicarbonate mode lamp. Unconditional error. No blood pump enable. Internal software error. 'FCN 3.74 001' - Lamp broken 'FCN 3.74 002' - Shortage ( High load ).
'FCN 3.75 000' - Case error Chemical disinfection lamp. Unconditional error. No blood pump enable. Internal software error. 'FCN 3.75 001' - Lamp broken 'FCN 3.75 002' - Shortage ( High load ).
'FCN 3.76 000' - Case error Heat disinfection lamp. Unconditional error. No blood pump enable. Internal software error. 'FCN 3.76 001' - Lamp broken 'FCN 3.76 002' - Shortage ( High load ).
'FCN 3.77 000' - Case error Rinse back lamp. Unconditional error. No blood pump enable. Internal software error. 'FCN 3.77 001' - Lamp broken 'FCN 3.77 002' - Shortage ( High load ).
'FCN 3.78 000' - Case error pH lamp. Unconditional error. No blood pump enable. Internal software error. 'FCN 3.78 001' - Lamp broken 'FCN 3.78 002' - Shortage ( High load ).
FCN 3.43, INFORMATION DISPLAY FUNCTION Purpose: * Displays text on the information display.
FCN 3.43 000 SOFTWARE ERROR [F,T,D,S] Please refer to the description of FCN 1.06 000 FCN 3.43 0xx SOFTWARE ERROR (SUB STATE ERROR FOR MAIN STATE xx) [F,T,D,S] Unconditional error.
General description of conditions for occuring: Internal software error. Technical description of conditions for occuring: The information display function has received an undefined command from another part of the software. The main state variable O$FCN$STATE(OF_INFO$DISP$MAIN) had the value "xx" as in the error code. The sub state variable O$FCN$STATE(OF_INFO$DISP$SUB) was out of limits for that main state.
Some possible explanations why the error has occured: * Software error. * RAM error.
Some hints of what to do if the error occurs: Please report the error to Gambro Sweden.
Automatical safety actions: Blood unit stop Fluid unit stop Safety relay off (Please refer to "Descriptions of safety actions" below)
FCN 3.43 256 TEXT INDEX OUT OF RANGE [F,T,D,S] Unconditional error.
General description of conditions for occuring: Internal software error. Technical description of conditions for occuring: The information display function was commanded to display an undefined text.
Some possible explanations why the error has occured: * Software error. * RAM error.
Some hints of what to do if the error occurs: Report the error to Gambro Sweden.
Automatical safety actions: Blood unit stop Fluid unit stop Safety relay off (Please refer to "Descriptions of safety actions" below)
FCN 3.43 257 SERVICE TEXT INDEX OUT OF RANGE [F,T,D,S] Unconditional error.
General description of conditions for occuring: Internal software error. The error code can occur also in other modes than service mode if the internal logging is activated. Technical description of conditions for occuring: The information display function was commanded to display an undefined service text.
Some possible explanations why the error has occured: * Software error. * RAM error.
Some hints of what to do if the error occurs: Report the error to Gambro Sweden.
Automatical safety actions: Blood unit stop Fluid unit stop Safety relay off (Please refer to "Descriptions of safety actions" below)
Descriptions of safety actions: Blood unit stop The blood pump is stopped. The arterial and venous clamps are closed. Fluid unit stop The heater is turned off. INVA, DIVA and BYVA are closed. The flow pump and suction pump are stopped. The valve and heater drivers are disabled. Safety relay off The 24V supply is disconnected from the blood pump, the heparin pump, the heater relay and the valves. The arterial and venous clamps are closed.
FCN344.PLM Manual input
'FCN 3.44 000' - Main case error. Manual input function. Unconditional error. No blood pump enable. Internal software error. 'FCN 3.44 xxx' - Sub case error in main case xxx. Unconditional error. No blood pump enable. Internal software error.
FCN345.PLM Bloodpump knob
'FCN 3.45 000' - Case error. Blood pump knob function. Unconditional error. No blood pump enable. Internal software error.
'FCN 3.45 001' - Blood tube diameter out of limits.
FCN346.PLM Venous pressure autolimit
'FCN 3.46 000' - Case error. Venous pressure autolimit function. Unconditional error. No blood pump enable. Internal software error.
FCN347.PLM TMP autolimit
'FCN 3.47 000' - Case error. TMP autolimit function. Unconditional error. No blood pump enable. Internal software error.
FCN348.PLM Extra pressure autolimit
'FCN 3.48 000' - Case error. Extra pressure autolimit function. Unconditional error. No blood pump enable. Internal software error.
FCN349.PLM Passed time
'FCN 3.49 000' - Case error. Passed time function. Unconditional error. No blood pump enable. Internal software error.
FCN350.PLM Remaining time
'FCN 3.50 000' - Case error. Remaining time function. Unconditional error. No blood pump enable. Internal software error.
FCN351.PLM Non diffusion time
'FCN 3.51 000' - Case error. Non diffusion time function. Unconditional error. No blood pump enable. Internal software error.
FCN352.PLM Disrinse time
'FCN 3.52 000' - Case error. Disinfection/rinse time function. Unconditional error. No blood pump enable. Internal software error.
FCN 3.53, BUZZER CONTROL FUNCTION Purpose: * During startup test: Turns on the main buzzer and checks that sound is detected. Turns off the main buzzer. Turns on the extra buzzer and checks that sound is detected. During normal operation: Turns the main buzzer on and off. Turns the main buzzer off when the operator has pressed the mute key. The main buzzer sound can be permanent or intermediate. A third possible buzzer sound is a short beep every two minutes.
FCN 3.53 000 SOFTWARE ERROR [F,T,D,S] Please refer to the description of FCN 1.06 000 FCN 3.53 001 EXTRA BUZZER NOT ON [F] Unconditional error.
General description of conditions for occuring: There was no sound detected when the extra buzzer was turned on. Technical description of conditions for occuring: During the startup test: The main buzzer was turned off (BUZR/=1) The extra buzzer was turned on (BUZE =1) The buzzer verify signal BUZV/=1 after 100ms
Some possible explanations why the error has occured: * Extra buzzer circuits broken. * The buzzer H1 on the driver board (K13014) is broken. * The outport IC 43 on the CPU board (K13022) is broken. * The inverter IC 51 on the CPU board (K13022) is broken. * Sound detection circuits broken. * The microphone M1 on the driver board (K13014) is broken. * IC 54, Q10, V25 or V26 on the driver board (K13014) is broken. * There is a bad connection between the driver board (K13014) P207 and the CPU board (K13019) P109. * The inport IC 74 on the CPU board (K13019) is broken.
Some hints of what to do if the error occurs: If there comes two beeps during the startup test, and both error codes FCN 3.53 001 and FCN 3.53 101 occurs, then refer to the section "Sound detection circuits broken" above. Check the signal BUZE from IC 43 on the CPU board to the buzzer H1 on the driver board. There should come a 100 ms long pulse during the startup test. Check the buzzer H1 on the driver board.
Automatical safety actions: Fluid unit stop (Please refer to "Descriptions of safety actions" below)
FCN 3.53 101 MAIN BUZZER NOT ON [F,T,D,S] Conditional error. Confirm the error by pressing the select key.
General description of conditions for occuring: There was no sound detected when the main buzzer was turned on. Technical description of conditions for occuring: During startup test: The main buzzer was turned on (BUZR/=0) The extra buzzer was turned off (BUZE =0) The buzzer verify signal BUZV/=1 after 100ms At normal operation: The main buzzer was commanded to be on (BUZR/=0) The buzzer verify signal BUZV/=1 after 700ms
Some possible explanations why the error has occured: * Main buzzer circuits broken. * The buzzer H2 on the driver board (K13014) is broken. * The optocoupler IC 39 on the driver board (K13014) is broken. * The outport IC 35 on the driver board (K13013) is broken. * The inverter IC 47 on the driver board (K13014) is broken. * The mute signal MUKY/ from the mute key is always 0V. * The PAL IC 43 on the driver board (K13014) is broken. * IC 44 or V 19 on the driver board (K13014) is broken so that the buzzer H2 does not get any 5V supply. * Sound detection circuits broken. * The microphone M1 on the driver board (K13014) is broken. * IC 54, Q10, V25 or V26 on the driver board (K13014) is broken. * There is a bad connection between the driver board (K13014) P207 and the CPU board (K13019) P109. * The inport IC 74 on the CPU board (K13019) is broken.
Some hints of what to do if the error occurs: If there comes two beeps during the startup test, and both error codes FCN 3.53 001 and FCN 3.53 101 occurs, then refer to the section "Sound detection circuits broken" above. Check the signal BUZR/ from IC 35 on the driver board to the buzzer H2 on the driver board. There should come a 100 ms long pulse during the startup test. Check the buzzer H2 and the opto coupler IC 39 on the driver board.
Automatical safety actions: None (The blood pump is enabled) (Please refer to "Descriptions of safety actions" below)
Descriptions of safety actions: Blood unit stop The blood pump is stopped. The arterial and venous clamps are closed. Fluid unit stop The heater is turned off. INVA, DIVA and BYVA are closed. The flow pump and suction pump are stopped. The valve and heater drivers are disabled. Safety relay off The 24V supply is disconnected from the blood pump, the heparin pump, the heater relay and the valves. The arterial and venous clamps are closed.
FCN354.PLM Time display
'FCN 3.54 000' - Case error. Time display function. Unconditional error. No blood pump enable. Internal software error.
FCN355.PLM Machine off
'FCN 3.55 000' - Case error Machine off function. Unconditional error. No blood pump enable. Internal software error. 'FCN 3.55 001' - Impossible to shut down the AK 90 with the on/off key. Unconditional error. Blood pump enabled. Check the machine off function. 'FCN 3.55 002' - Power PLD unrequested off. No error, used for hardware diagnostic. Pin 12 on IC-1 on power fail extension board has been detected low for more than 5 sedconds. Check the connections and function of the power fail extension board.
FCN356.PLM CDS Disrinse communication
'FCN 3.56 000' - Case error CDS disinfection/rinse communication. Unconditional error. No blood pump enable. Internal software error.
FCN357.PLM CDS chem dis communication
'FCN 3.57 000' - Case error CDS chemical disinfection communication. Unconditional error. No blood pump enable. Internal software error.
FCN358.PLM CDS heat dis communication
'FCN 3.58 000' - Case error CDS heat disinfection communication. Unconditional error. No blood pump enable. Internal software error.
FCN359.PLM CDS external alarm communication
'FCN 3.59 000' - Case error CDS external alarm communication. Unconditional error. No blood pump enable. Internal software error. FCN360.PLM Heparine autostop limit
'FCN 3.60 000' - Case error. Heparin autostop limit function. Unconditional error. No blood pump enable. Internal software error.
FCN 3.61, UF GAIN CONTROL FUNCTION Purpose: * Calculates the needed UF rate to obtain the set UF volume during the remaining treatment time. * Creates an attention if the calculated UF rate is out of limits.
FCN 3.61 000 SOFTWARE ERROR [F,T,D,S] Please refer to the description of FCN 1.06 000 FCN 3.61 001 N O T U S E D A N Y L O N G E R [T] Software error in isolated UF mode. Please refer to the description of FCN 1.06 000
FCN 3.61 002 N O T U S E D A N Y L O N G E R [T] Software error in normal diffusion mode. Please refer to the description of FCN 1.06 000
FCN363.PLM Access control
'FCN 3.63 000' - Case error. Access control function. Unconditional error. No blood pump enable. Internal software error.
FCN364.PLM CDS rinse communication
'FCN 3.64 000' - Case error CDS rinse communication. Unconditional error. No blood pump enable. Internal software error. FCN365.PLM Calibration function
'FCN 3.65 000' - Case error Calibration function. Unconditional error. No blood pump enable. Internal software error. FCN366.PLM Rinse time calculation function
'FCN 3.66 000' - Case error Rinse time calculation function. Unconditional error. No blood pump enable. Internal software error. FCN368.PLM UF autocalibration function
'FCN 3.68 000' - Case error UF auto calibration function. Unconditional error. No blood pump enable. Internal software error. FCN369.PLM Reset set values and limits function
'FCN 3.69 000' - Case error Reset set values and limits function. Unconditional error. No blood pump enable. Internal software error. FCN370.PLM Disinfection timer function
'FCN 3.70 000' - Case error Disinfection timer function. Unconditional error. No blood pump enable. Internal software error. FCN371.PLM Cleaning timer function
'FCN 3.71 000' - Case error Cleaning timer function. Unconditional error. No blood pump enable. Internal software error. FCN 3.79, CONCENTRATE BANK SELECT FUNCTION Purpose: * In acetate mode: Reads the variable ORB(OB_AC$CONC$BANK$SEL) and gives one of the variables ORL(O_CONC$BANK$0), ORL(O_CONC$BANK$1) or ORL(O_CONC$BANK$2) the value TRUE. * In bicarbonate mode: Reads the variable ORB(OB_BC$CONC$BANK$SEL) and gives one of the variables ORL(O_CONC$BANK$0), ORL(O_CONC$BANK$1) or ORL(O_CONC$BANK$2) the value TRUE.
FCN 3.79 000 SOFTWARE ERROR [F,T,D,S] Please refer to the description of FCN 1.06 000 FCN 3.79 001 ACETATE CONCENTRATE BANK SELECT OUT OF LIMIT [F,T,D,S] Unconditional error.
General description of conditions for occuring: Internal software error. Technical description of conditions for occuring: The variable ORB(OB_AC$CONC$BANK$SEL) has a value which is not 0, 1 or 2. The variable gets a value when the operator selects one of three available acetate concentrate banks.
Some possible explanations why the error has occured: * RAM error.
Some hints of what to do if the error occurs: Log the variable ORB(OB_AC$CONC$BANK$SEL). Try to select another acetate concentrate bank. Check if the RAM is broken. If the RAM is OK then report the error to Gambro.
Automatical safety actions: Fluid unit stop (Please refer to "Descriptions of safety actions" below)
FCN 3.79 002 BICARBONATE CONCENTRATE BANK SELECT OUT OF LIMIT [F,T,D,S] Unconditional error.
General description of conditions for occuring: Internal software error. Technical description of conditions for occuring: The variable ORB(OB_BC$CONC$BANK$SEL) has a value which is not 0, 1 or 2. The variable gets a value when the operator selects one of three available bicarbonate concentrate banks.
Some possible explanations why the error has occured: * RAM error.
Some hints of what to do if the error occurs: Log the variable ORB(OB_BC$CONC$BANK$SEL). Try to select another bicarbonate concentrate bank. Check if the RAM is broken. If the RAM is OK then report the error to Gambro.
Automatical safety actions: Fluid unit stop (Please refer to "Descriptions of safety actions" below)
Descriptions of safety actions: Blood unit stop The blood pump is stopped. The arterial and venous clamps are closed. Fluid unit stop The heater is turned off. INVA, DIVA and BYVA are closed. The flow pump and suction pump are stopped. The valve and heater drivers are disabled. Safety relay off The 24V supply is disconnected from the blood pump, the heparin pump, the heater relay and the valves. The arterial and venous clamps are closed.
FCN380.PLM Attention function
'FCN 3.80 000' - Case error attention display function. Unconditional error. No blood pump enable. Internal software error. 'FCN 3.80 001' - Case error attention evaluation function. Unconditional error. No blood pump enable. Internal software error. 'FCN 3.80 002' - Attention evaluation function. Attention state is corrupt. Unconditional error. No blood pump enable. Internal software error.
'FCN 3.80 003' - Attention evaluation function. Attention queue is full, no more space. Unconditional error. No blood pump enable. Internal software error.
FCN381.PLM Preset concentrate calculation fcn
'FCN 3.81 000' - Case error Preset concentrate calculation fcn. Unconditional error. No blood pump enable. Internal software error.
FCN382.PLM Preset alternative & mode fcn
'FCN 3.82 000' - Case error Preset alternative & mode fcn. Unconditional error. No blood pump enable. Internal software error.
FCN383.PLM Preset disinfection alternative, type and mode. [F,S]
'FCN 3.83 000' - Case error, Preset disinfection alternative, type and mode fcn. Unconditional error. No blood pump enable. Internal software error.
FCN384.PLM Chem disinfection bank. [F,T]
'FCN 3.84 000' - Case error, Chem disinfection bank fcn. Unconditional error. No blood pump enable. Internal software error.
Function group 4 : A/D, E2PROM, WATCHDOG functions.
FCN 4.04, MM 90 A/D CONVERTER FUNCTION Purpose: This function does only execute if the preset EL(EEL_MM90$INSTALLED) has the value TRUE. A voltage to frequency converter and a timer is used for the A/D conversion. * Reads A/D channel and A/D value from a buffer. * Uses the offset voltage and the reference voltage to compensate the raw data. * Filters the raw data. * Scales the raw data.
FCN 4.04 000 SOFTWARE ERROR [F,T,D,S] Please refer to the description of FCN 1.06 000 FCN 4.04 001 MM 90 A/D CONVERTER BUFFER POINTER ERROR [F,T,D,S] Transient error. (No safety action) Unconditional error if more than three errors during one treatment.
General description of conditions for occuring: The MM 90 A/D converter interrupt function puts data in a buffer and incrementes an input pointer. The MM 90 A/D converter function fetches data from that buffer and increments an output pointer. One of these pointers had an incorrect value. Technical description of conditions for occuring: ALTERNATIVE 1: The MM 90 A/D converter interrupt function should have stored A/D converter data in a buffer and should have changed the value of ADMM_IN_PTR to a value that was not the same as ADMM_OUT_PTR. The MM 90 A/D converter interrupt routine had not executed. The MM 90 A/D converter function FCN 404 detected that ADMM_OUT_PTR = ADMM_IN_PTR for 10s. ALTERNATIVE 2: The MM 90 A/D converter function FCN 404 should all the time give the variable ADMM_OUT_PTR even values (0,2,4,....) FCN 404 detected that ADMM_OUT_PTR was not even. ALTERNATIVE 3: The MM 90 A/D converter interrupt routine detected that the MM 90 A/D converter function had fetched too much data from the A/D converter buffer. ADMM_OUT_PTR = ADMM_IN_PTR+4 ALTERNATIVE 4: The MM 90 A/D converter interrupt routine should all the time give the variable ADMM_IN_PTR values dividable by 4 (0,4,8,....) The MM 90 A/D converter interrupt routine detected that ADMM_IN_PTR had a value which was NOT 0,4,8,... Some possible explanations why the error has occured: ALTERNATIVE 1: * The timer IC 2 on the expansion board (K14716) is broken. * There is a bad connection between the expansion board (K14716) P2 and the CPU board (K13020) P107. * The CPU IC 89 on the CPU board (K13020) is broken. ALTERNATIVE 2, 3, 4: * RAM error.
Automatical safety actions: Fluid unit stop (Please refer to "Descriptions of safety actions" below) |