C CFM 020 007 VFC CHANNEL 2 REFERENCE MEASUREMENT ERROR Unconditional error, blood pump available General description of conditions for occurring: - The VFC channel 2 has been connected to the reference voltage. The pulse counter (IC23 counter 4, control driver board) has measured a frequency that differs too much from the frequency that usually is measured when the reference voltage is connected. Technical description of conditions for occurring: VFC2.VALUE > 0.86241*VFC2.OFFSET for 2s OR VFC2.VALUE < 0.82859*VFC2.OFFSET for 2s
VFC2.OFFSET = VFC value when 0V is connected. Some possible explanations why the error has occurred: - The VFC, (voltage to frequency) IC 15 on control driver board, or some component around it is broken - IC 23 broken (control driver board) C CFM 020 008 VFC CHANNEL 1 OFFSET MEASUREMENT ERROR Unconditional error, blood pump available General description of conditions for occurring: - The VFC channel 1 has been connected to 0V. The pulse counter has measured a frequency that differs too much from the frequency that usually is measured when 0V is connected. Technical description of conditions for occurring: VFC1.VALUE has been outside the allowed interval 6907-10907 for 2s Some possible explanations why the error has occurred: - The VFC, (voltage to frequency) IC 14 on control driver board, or some component around it is broken - IC 23 broken (control driver board) C CFM 020 009 VFC CHANNEL 2 OFFSET MEASUREMENT ERROR Unconditional error, blood pump available General description of conditions for occurring: - The VFC channel 2 has been connected to 0V. The pulse counter has measured a frequency that differs too much from the frequency that usually is measured when 0V is connected. Technical description of conditions for occurring: VFC2.VALUE has been outside the allowed interval 6907-10907 for 2s Some possible explanations why the error has occurred: - The VFC, (voltage to frequency) IC 15 on control driver board, or some component around it is broken - IC 23 broken (control driver board) C CFM 020 010 MEASURED MAGNETIZATION CURRENT INCORRECT Unconditional error, blood pump available General description of conditions for occurring: - Measured magnetization current differs to much from typical value. Technical description of conditions for occurring: - UF coil current is outside the limits 0.18 A - 0.28 A for 5 seconds. Some possible explanations why the error has occurred: - UF cell broken - Bad connection between UF cell and control driver board - Broken control driver board C CFM 020 011 UF CELL TARATION NOT APPROVED Unconditional error, blood pump available General description of conditions for occurring: - A self calibration (taration) was run. It failed as the values of the variables FRR(FR_UFRTOLERANCE) and FRR(FR_UFROFFSET) had changed too much since the last self calibration. A new self calibration was run after 5 minutes. It did also fail for the same reason. Technical description of conditions for occurring: - CalibrDiffFlowNew differed more than 1 ml/min from CalibrDiffFlowOld A new self calibration was done after 5 minutes. CalibrDiffFlow again differed more than 1.0 ml/min from CalibrDiffFlowNew AND CalibrDiffFlow again differed more than 1.0 ml/min from CalibrDiffFlowOld Some possible explanations why the error has occurred: - UF cell broken or dirty - There is air through the UF cell - Leakage through: RIVA, ZEVA, DIVA or TAVA C CFM 020 012 UF CELL TARATION NOT APPROVED Unconditional error, blood pump available General description of conditions for occurring: - A self calibration (taration) was run. It failed as the value of the variable FRR(FR_UFRTOLERANCE) differed too much from the value that was stored in the EEPROM or estimated at the last performed self calibration. Technical description of conditions for occurring: CALIBR DIFF FLOWnew differed more than 7.5 ml/min from CALIBR DIFF FLOWold Some possible explanations why the error has occurred: - UF cell broken or dirty - There is air through the UF cell - Leakage through: RIVA, ZEVA, DIVA or TAVA C CFM 020 014 TOO LOW SELF CALIBRATION FLOW Unconditional error, blood pump available General description of conditions for occurring: - The flow measuring unit measured a flow through channel 1 that was less than 100 ml/min during self calibration (taration). Technical description of conditions for occurring: - FRR(FR_CH1RATE) < 100.0 ml/min. Some possible explanations why the error has occurred: - Uncalibrated or broken. pressure regulator - Dirty flow restrictor. - Dirty outlet restrictor - Dirty or broken flow pump:. - Dirty or broken suction pump. - Broken or uncalibrated degassing pressure transducer. - ZEVA: leakage - RIVA: leakage C CFM 020 015 UNSTABLE FLOW SIGNALS DURING SELF CALIBRATION Unconditional error, blood pump available General description of conditions for occurring: - The self calibration (taration) failed because at least one of the measured flow signals varied too much. A new self calibration that was done 5 minutes later did also fail because at least one of the measured flow signals varied too much. Technical description of conditions for occurring: - Variance of the flow signals were calculated The differential flow during either the offset calibration or during the differential flow varied too much Some possible explanations why the error has occurred: - Air is flowing through flow channel 1. There is too much air in the inlet water - Valves: RIVA, ZEVA, diva or TAVA are leaking - Flow cell: the flow cell is dirty. - Flow cell: the flow cell is broken - Flow restrictor: the restrictor is dirty - Flow pump: the pump is not working properly - Degassing chamber: the o-ring at the top of the chamber is not in its correct position. fluid is leaking. - Degassing pressure transducer: not working properly. C CFM 020 016 SELF CALIBRATION TIME OUT Unconditional error, blood pump available General description of conditions for occurring: - The self calibration (taration) executed for a too long time. Technical description of conditions for occurring: - Self calibration time > max. self calibration time (max. self calibration time = 15s - 68s) Some possible explanations why the error has occurred: - Secondary error to the error C CFM 020 006 - Secondary error to the error C CFM 020 007 - Secondary error to the error C CFM 020 008 - Secondary error to the error C CFM 020 009. C CFM 020 017 VFC CALIBRATION TIME OUT (VFC1 OFFSET). Unconditional error, blood pump available General description of conditions for occurring: - Please refer to the description of C CFM 020 008 C CFM 020 018 VFC CALIBRATION TIME OUT (VFC2 OFFSET). Unconditional error, blood pump available General description of conditions for occurring: - Please refer to the description of C CFM 020 009 C CFM 020 019 VFC CALIBRATION TIME OUT (VFC1 REFERENCE). Unconditional error, blood pump available General description of conditions for occurring: - Please refer to the description of C CFM 020 006 C CFM 020 020 VFC CALIBRATION TIME OUT (VFC2 REFERENCE). Unconditional error, blood pump available General description of conditions for occurring: - Please refer to the description of C CFM 020 007 C CFM 020 021 MEASURED MAGNETIZATION CURRENT INCORRECT Unconditional error, blood pump available General description of conditions for occurring: - Please refer to the description of C CFM 020 010 C CFM 020 024 UNSTABLE SIGNAL DURING UF CELL CALIBRATION General description of conditions for occurring: - The UF cell calibration failed because at least one of the measured flow signals varied too much Some possible explanations why the error has occurred: - Air is flowing through flow channel 1. There is too much air in the inlet water - Valves: RIVA, ZEVA, diva or TAVA are leaking - Flow cell: the flow cell is dirty. - Flow cell: the flow cell is broken - Flow restrictor: the restrictor is dirty - Flow pump: the pump is not working properly - Degassing chamber: the o-ring at the top of the chamber is not in its correct position. fluid is leaking. - Degassing pressure transducer: not working properly. C CFM 020 025 UNSTABLE SIGNAL DURING UF OFFSET CALIBRATION Unconditional error, blood pump available General description of conditions for occurring: - The channel 1 flow signal and the channel 2 flow signal varied too much during the main flow offset calibration in service mode. There should be no flow through the measuring cells. Technical description of conditions for occurring: - The variance of the channel 1 flow and the variance of the channel 2 flow were out of limit Some possible explanations why the error has occurred: - The flow cell is dirty - The flow cell is broken - BYVA/DIVA/TAVA leakage C CFM 020 029 UF CALIBRATION REFERENCE VOLUME OUT OF LIMIT Unconditional error, blood pump available General description of conditions for occurring: - The reference volume was outside its specified limits during the UF calibration in service mode. Technical description of conditions for occurring: ORI(OI_UFCalRefVol) < PI(EPI_UFCalRefVolLl) or ORI(OI_UFCalRefVol) > PI(EPI_UFCalRefVolHl) Some possible explanations why the error has occurred: - Memory RAM corrupt C CFM 020 030 CASE ERROR Unconditional error, blood pump available General description of conditions for occurring: - Internal software error C CFM 020 031 CASE ERROR Unconditional error, blood pump available General description of conditions for occurring: - Internal software error C CFM 020 032 CASE ERROR Unconditional error, blood pump available General description of conditions for occurring: - Internal software error C CFM 020 033 UF PULS TIMER ERROR Unconditional error, blood pump available General description of conditions for occurring: - The UF pulse timer is not working properly. Technical description of conditions for occurring: - The timer 9513 has returned error message(s). Some possible explanations why the error has occurred: - The timer IC 34 on the control driver board is broken. C CFM 020 034 UF INTERRUPT TIMER ERROR Unconditional error, blood pump available General description of conditions for occurring: - The UF interrupt timer is not working properly. Technical description of conditions for occurring: - The timer 9513 has returned error message(s). Some possible explanations why the error has occurred: - The timer IC 23 on the control driver board is broken. C CFM 020 035 VFC BUFFER FULL ERROR Unconditional error, blood pump available General description of conditions for occurring: - The UF interrupt procedure has filled the VFC data buffer. Some possible explanations why the error has occurred: - Software error - The timer IC 23 on the control driver board is broken. - The interrupt controller is broken C CFM 020 036 VFC BUFFER OVERFLOW ERROR Unconditional error, blood pump available General description of conditions for occurring: - Software error C CFM 020 037 MUX MISMATCH OCCURRED TOO MANY TIMES Unconditional error, blood pump available General description of conditions for occurring: - Software error C CFM 020 038 MUX ERROR Unconditional error, blood pump available General description of conditions for occurring: - Software error
C CFM 020 039 UNSTABLE FLOW SIGNALS DURING SELF CALIBRATION REMOVED FROM PROGRAM VERSION 8.00 Unconditional error, blood pump available General description of conditions for occurring: - The self calibration (taration) failed because at least one of the measured flow signals varied too much. A new self calibration that was done 5 minutes later did also fail because at least one of the measured flow signals varied too much. Technical description of conditions for occurring: - Variance of the flow signals were calculated The flow through one or both of the UF channels during the differential flow calibration stage varied too much. Some possible explanations why the error has occurred: - Air is flowing through flow channel 1. There is too much air in the inlet water - Valves: RIVA, ZEVA, diva or TAVA are leaking - Flow cell: the flow cell is dirty. - Flow cell: the flow cell is broken - Flow restrictor: the restrictor is dirty - Flow pump: the pump is not working properly - Degassing chamber: the o-ring at the top of the chamber is not in its correct position. fluid is leaking. - Degassing pressure transducer: not working properly. - This error code occurs if an internal UF-calibration is made in P5.00, P6.00 or P7.00 C CFT 020, Flow Meter, Treatment Mode
C CFT 020 000 CASE ERROR Transient error General description of conditions for occurring: - Internal software error
C CF 021, UF Rate Filter
C CF 021 000 CASE ERROR Transient error General description of conditions for occurring: - Internal software error
C CF 022, UF Rate Regulator
C CF 022 000 CASE ERROR Transient error General description of conditions for occurring: - Internal software error
C CFS 023, Dia Pressure, Service Mode
C CFS 023 000 CASE ERROR Transient error General description of conditions for occurring: - Internal software error
C CFT 026, UF Volume Accumulation, Treatment Mode
C CFT 026 000 CASE ERROR Transient error General description of conditions for occurring: - Internal software error
C CF 027, Disinfection Valves
C CF 027 000 CASE ERROR Transient error General description of conditions for occurring: - Internal software error
C CFD 028, MM95, Cond. Control, Disinfection Mode
C CFD 028 000 CASE ERROR Transient error General description of conditions for occurring: - Internal software error
C CFF 028, MM95, Cond. Control, Function Check
C CFF 028 000 CASE ERROR Transient error General description of conditions for occurring: - Internal software error
C CFM 028, MM95, Cond. Control
C CFM 028 000 CASE ERROR Transient error General description of conditions for occurring: - Internal software error C CFM 028 001 CONCENTRATE PUMP TIMER ERROR Unconditional error, blood pump available General description of conditions for occurring: - The concentrate pump timer is not working properly. Technical description of conditions for occurring: - The timer 9513 has returned error message(s). Some possible explanations why the error has occurred: - The timer IC 21 on the control CPU board is broken. C CFM 028 002 CONFIGURATION ERROR Unconditional error, blood pump available General description of conditions for occurring: - Internal software error
C CFS 028, MM95, Cond. Control, Service Mode
C CFS 028 000 CASE ERROR Transient error General description of conditions for occurring: - Internal software error
C CFT 028, MM95, Cond. Control Treatment Mode
C CFT 028 000 CASE ERROR Transient error General description of conditions for occurring: - Internal software error
C CF 029, Fluid composition supervision
C CF 029 000 CASE ERROR Transient error General description of conditions for occurring: - Internal software error
C CF 030, Remaining Disinfection Time Calculation
C CF 030 000 CASE ERROR Transient error General description of conditions for occurring: - Internal software error
C CF 031, Bicart Valve
C CF 031 000 CASE ERROR Transient error General description of conditions for occurring: - Internal software error
C CF 032, AK95, UF Logging
C CF 032 000 CASE ERROR Transient error General description of conditions for occurring: - Internal software error
C CFT 033, UF Gain Control
C CFT 033 000 CASE ERROR Transient error General description of conditions for occurring: - Internal software error
C CFS 050, Communication
C CFS 050 000 CASE ERROR Transient error General description of conditions for occurring: - Internal software error
C CFS 068, UF calibration box mode function
C CFS 068 000 CASE ERROR Transient error General description of conditions for occurring: - Internal software error C CFS 068 001 SET DISCRETE PORT ERROR Unconditional error, blood pump available General description of conditions for occurring: - Internal software error C CFS 068 002 SET DISCRETE PORT ERROR Unconditional error, blood pump available General description of conditions for occurring: - Internal software error C CFS 068 003 SET DISCRETE PORT ERROR Unconditional error, blood pump available General description of conditions for occurring: - Internal software error C CFS 068 004 SET DISCRETE PORT ERROR Unconditional error, blood pump available General description of conditions for occurring: - Internal software error C CFS 068 005 SET DISCRETE PORT ERROR Unconditional error, blood pump available General description of conditions for occurring: - Internal software error C CFT 069, Disinfection control
C CFT 069 000 CASE ERROR Transient error General description of conditions for occurring: - Internal software error
C CO 001, E2PROM SCAN
C CO 001 000 CASE ERROR Transient error General description of conditions for occurring: - Internal software error
C CO 002, Key Board
C CO 002 000 CASE ERROR Transient error General description of conditions for occurring: - Internal software error C CO 002 001 DISC. I/O ERROR Transient error General description of conditions for occurring: - Internal software error
C CO 003, Venous Pressure Bargraph
C CO 003 000 CASE ERROR Transient error General description of conditions for occurring: - Internal software error
C CO 004, Blood Flow Bargraph
C CO 004 000 CASE ERROR Transient error General description of conditions for occurring: - Internal software error
C CO 005, Conductivity Bargraph
C CO 005 000 CASE ERROR Transient error General description of conditions for occurring: - Internal software error
C CO 006, Temperature Bargraph
C CO 006 000 CASE ERROR Transient error General description of conditions for occurring: - Internal software error
C CO 009, UF Volume Bargraph
C CO 009 000 CASE ERROR Transient error General description of conditions for occurring: - Internal software error
C CO 010, UF Rate Bargraph
C CO 010 000 CASE ERROR Transient error General description of conditions for occurring: - Internal software error C CO 010 001 DISC. I/O ERROR Transient error General description of conditions for occurring: - Internal software error
C CO 011, ON/OFF Lamp
C CO 011 000 CASE ERROR Transient error General description of conditions for occurring: - Internal software error
C CO 012, Rinse Lamp
C CO 012 000 CASE ERROR Transient error General description of conditions for occurring: - Internal software error
C CO 013, Reset Lamp
C CO 013 000 CASE ERROR Transient error General description of conditions for occurring: - Internal software error
C CO 015, Isol UF Mode Lamp
C CO 015 000 CASE ERROR Transient error General description of conditions for occurring: - Internal software error
C CO 016, Volume Control Mode Lamp
C CO 016 000 CASE ERROR Transient error General description of conditions for occurring: - Internal software error
C CO 019, Single Needle Limits Lamp
C CO 019 000 CASE ERROR Transient error General description of conditions for occurring: - Internal software error
C CO 022, Heparin Lamp
C CO 022 000 CASE ERROR Transient error General description of conditions for occurring: - Internal software error
C CO 023, Time Lamp
C CO 023 000 CASE ERROR Transient error General description of conditions for occurring: - Internal software error
C CO 026, UF Volume Lamp
C CO 026 000 CASE ERROR Transient error General description of conditions for occurring: - Internal software error
C CO 027, UF Rate Lamp
C CO 027 000 CASE ERROR Transient error General description of conditions for occurring: - Internal software error
C CO 030, Attention Lamp
C CO 030 000 CASE ERROR Transient error General description of conditions for occurring: - Internal software error
C CO 031, Technical Error Lamp
C CO 031 000 CASE ERROR Transient error General description of conditions for occurring: - Internal software error
C CO 032, Auto Heat Lamp
C CO 032 000 CASE ERROR Transient error General description of conditions for occurring: - Internal software error
C CO 034, Blood Pump Lamp
C CO 034 000 CASE ERROR Transient error General description of conditions for occurring: - Internal software error
C CO 035, Auto Rinse Lamp
C CO 035 000 CASE ERROR Transient error General description of conditions for occurring: - Internal software error
C CO 037, Fluid Bypass Lamp
C CO 037 000 CASE ERROR Transient error General description of conditions for occurring: - Internal software error
C CO 038, UF Start/Stop Lamp
C CO 038 000 CASE ERROR Transient error General description of conditions for occurring: - Internal software error
C CO 041, Blood Flow Path Lamp
C CO 041 000 CASE ERROR Transient error General description of conditions for occurring: - Internal software error
C CO 042, Fluid Flow Path Lamp
C CO 042 000 CASE ERROR Transient error General description of conditions for occurring: - Internal software error C CO 042 001 DISC. I/O ERROR Transient error General description of conditions for occurring: - Internal software error C CO 042 002 FLUID BYPASS ERROR Transient error General description of conditions for occurring: - Internal software error
C CO 043, Information Display
C CO 043 000 CASE ERROR Transient error General description of conditions for occurring: - Internal software error C CO 043 256 INFORMATION DISPLAY TABLE ERROR (1) General description of conditions for occurring: - Internal software error C CO 043 257 INFORMATION DISPLAY TABLE ERROR (2) General description of conditions for occurring: - Internal software error C CO 043 260 LANGUAGE SELECT ERROR General description of conditions for occurring: - Internal software error C CO 043 261 DISC. I/O ERROR General description of conditions for occurring: - Internal software error C CO 043 262 DISC. I/O ERROR General description of conditions for occurring: - Internal software error C CO 043 263 DISC. I/O ERROR General description of conditions for occurring: - Internal software error
C COS 043, Information Display, Service Mode
C COS 043 016 INFORMATION DISPLAY TABLE ERROR, SERVICE TEXT Transient error General description of conditions for occurring: - Internal software error C COS 043 017 INFORMATION DISPLAY TABLE ERROR, CALIBRATION TEXTS Transient error General description of conditions for occurring: - Internal software error C COS 043 018 INFORMATION DISPLAY TABLE ERROR, TRANSDUCER NUMBER Transient error General description of conditions for occurring: - Internal software error C COS 043 019 VARIABLE SCALING ERROR Transient error General description of conditions for occurring: - Internal software error
C COM 044, Manual Input
C COM 044 000 CASE ERROR Transient error General description of conditions for occurring: - Internal software error
C COP 044, Manual Input, Preset Mode
C COP 044 000 CASE ERROR Transient error General description of conditions for occurring: - Internal software error
C COS 044, Manual Input, Service Mode
C COS 044 xxx CASE ERROR Transient error General description of conditions for occurring: - Internal software error
C COT 044, Manual Input, Treatment Mode
C COT 044 xxx CASE ERROR Transient error General description of conditions for occurring: - Internal software error
C CO 045, Blood Pump Knob
C CO 045 000 CASE ERROR Transient error General description of conditions for occurring: - Internal software error C CO 045 001 INCORRECT BLOOD TUBE DIAMETER (PRESET VALUE) Transient error General description of conditions for occurring: - The parameter for the set value of the blood tube diameter is out of limit. Technical description of conditions for occurring: - ORI(OI_BlTubeDiam) < PI(EPI_BlTubeDiamLl) or ORI(OI_BlTubeDiam) > PI(EPI_BlTubeDiamHl) Some possible explanations why the error has occurred: - Internal software error - Read/write memory error (re-preset and restart AK 95) C CO 045 002 DISC. I/O ERROR Transient error General description of conditions for occurring: - Internal software error
C CO 049, Time Calculation
C CO 049 000 CASE ERROR Transient error General description of conditions for occurring: - Internal software error
C CO 050, Remaining Time Calculation
C CO 050 000 CASE ERROR Transient error General description of conditions for occurring: - Internal software error
C CO 051, Non Diffusion Time Calculation
C CO 051 000 CASE ERROR Transient error General description of conditions for occurring: - Internal software error
C CO 053, Buzzer Control
C CO 053 000 CASE ERROR Transient error General description of conditions for occurring: - Internal software error
C CO 054, Time Display
C CO 054 000 CASE ERROR Transient error General description of conditions for occurring: - Internal software error
C CO 060, Heparin Auto Stop Limit
C COS 060 000 CASE ERROR Transient error General description of conditions for occurring: - Internal software error
C CO 062, Priming Lamp
C CO 062 000 CASE ERROR Transient error General description of conditions for occurring: - Internal software error
C CO 067, Auto Priming Lamp
C CO 067 000 CASE ERROR Transient error General description of conditions for occurring: - Internal software error
C CO 069, BSM, Reset Limits And Set Value
C CO 069 000 CASE ERROR Transient error General description of conditions for occurring: - Internal software error
C CO 072, Conductivity
C CO 072 000 CASE ERROR Transient error General description of conditions for occurring: - Internal software error
C CO 075, Chem Lamp
C CO 075 000 CASE ERROR Transient error General description of conditions for occurring: - Internal software error
C CO 076, Heat Lamp
C CO 076 000 CASE ERROR Transient error General description of conditions for occurring: - Internal software error
C CO 077, Rinse Back Lamp
C CO 077 000 CASE ERROR Transient error General description of conditions for occurring: - Internal software error
C CO 079, Concentrate Bank
C CO 079 000 CASE ERROR Transient error General description of conditions for occurring: - Internal software error C CO 079 001 CASE ERROR Transient error General description of conditions for occurring: - Internal software error C CO 079 002 CASE ERROR Transient error General description of conditions for occurring: - Internal software error
C CO 080, Attention
C CO 080 000 CASE ERROR Transient error General description of conditions for occurring: - Internal software error C CO 080 001 CASE ERROR Transient error General description of conditions for occurring: - Internal software error C CO 080 002 EVALUATION STATE ERROR Transient error General description of conditions for occurring: - Internal software error C CO 080 003 ATTENTION QUEUE ERROR Transient error General description of conditions for occurring: - Internal software error - Corrupt RAM memory
C CO 084, Disinfection Bank Selection
C CO 084 000 CASE ERROR Transient error General description of conditions for occurring: - Internal software error C CO 084 001 CASE ERROR Transient error General description of conditions for occurring: - Internal software error
C CO 087, Protective Set
C CO 087 000 CASE ERROR Transient error General description of conditions for occurring: - Internal software error
C COxx 088, Functional Check
C CO 088 000 CASE ERROR Transient error General description of conditions for occurring: - Internal software error C COFB 088 001 AIR DETECTOR TEST ERROR, CPU P Unconditional error, blood pump not available General description of conditions for occurring: - CPU P air detector does not detect "air" when the transmitted intensity is lowered during Fch. Technical description of conditions for occurring: - PRI(PI_AirDetAirPErrCnt) > 10 Some possible explanations why the error has occurred: - Air detector broken - Broken BM I/O board (IC5, IC14 and components, IC12, IC17, Q6, Q7, IC19 and components) - Broken protective CPU board (IC43 and components) (N.E. IC48) - Broken protective driver board (IC6 and components, V6 and components, IC4 and components, IC55) C COFB 088 002 AIR DETECTOR TEST ERROR, CPU W Unconditional error, blood pump not available General description of conditions for occurring: - CPU W air detector does not detect "air" when the transmitted intensity is lowered during Fch. Technical description of conditions for occurring: - PRI(PI_AirDetAirWErrCnt) > 10 Some possible explanations why the error has occurred: - Air detector broken - Broken BM I/O board (IC5, IC14 and components, IC12, IC17, Q6, Q7, IC19 and components) - Broken protective CPU board (IC3 and components) - Broken protective driver board (IC2 and components, V3 and components, IC1 and components, IC55) C COFB 088 003 AIR DETECTOR TEST ERROR, CPU P AND W Unconditional error, blood pump not available General description of conditions for occurring: - One or both of CPU P and CPU W's air detectors does not sense alarm free status when the transmitted intensity is increased to normal level. This error code should only occure if anything is done with the air detector during the low intensity test phase, since this condition (alarm free status) was already fulfilled at the start of the test. Technical description of conditions for occurring: - PRI(PI_AirDetNoAirErrCnt) > 10 Some possible explanations why the error has occurred: - Low level in venous drip chamber. Check the level and restart the monitor. C COFF 088 004 BLOOD LEAK DETECTOR TEST ERROR, VALUE TOO HIGH Unconditional error, blood pump available General description of conditions for occurring: - The blood leak detector value is too high when the LED is off. Technical description of conditions for occurring: - PRI(PI_BlLeakOffLevelErrCnt) > 4 - Blood leak detector value > 2mV measured by ADC (CPU P) Some possible explanations why the error has occurred: - Blood leak detector uncalibrated - Blood leak detector unit broken - Too much external light in blood leak detector ( AK 95 open for service) - Broken protective driver board (R71 and components, IC29, IC25) - Broken/uncalibrated A/D system, protective driver board (IC12, IC13, IC15, and components) C COFF 088 005 BLOOD LEAK DETECTOR TEST ERROR, VALUE OUT OF LIMIT Unconditional error, blood pump available General description of conditions for occurring: - The blood leak detector value is out of the limits when the LED is lit Technical description of conditions for occurring: - PRI(PI_BlLeakOnLevelErrCnt) > 4 - Blood leak detector value < -10 or blood leak detector value > +60 (AK95P.CPU_P.SUP.MEAS.AD_INPUT.blLeakValue) Some possible explanations why the error has occurred: - Blood leak detector uncalibrated - Blood leak detector dirty - Blood leak detector broken - Broken protective driver board (R71 and components, IC29, IC25) - Broken/uncalibrated A/D system, protective driver board (IC12, IC13, IC15, and components) C COFF 088 006 BLOOD LEAK DETECTOR TEST ERROR, UNSTABLE OFF Unconditional error, blood pump available General description of conditions for occurring: The blood leak detector value is unstable when the LED is turned off. Technical description of conditions for occurring: - PRI(PI_BlLeakOffStableErrCnt) > 4 - Blood leak detector variance >5 Some possible explanations why the error has occurred: - Blood leak detector broken - Broken protective driver board (R71 and components, IC29, IC25) - Broken/uncalibrated A/D system, protective driver board (IC12, IC13, IC15, and components) C COFF 088 007 BLOOD LEAK DETECTOR TEST ERROR, UNSTABLE ON Unconditional error, blood pump available General description of conditions for occurring: The blood leak detector value is unstable when the LED is turned on. Technical description of conditions for occurring: - PRI(PI_BlLeakOnStableErrCnt) > 4 - Blood leak detector variance >5 Some possible explanations why the error has occurred: - Blood leak detector broken - Broken protective driver board (R71 and components, IC29, IC25) - Broken/uncalibrated A/D system, protective driver board (IC12, IC13, IC15, and components) C COFB 088 008 BLOOD PUMP TEST ERROR, PUMP NOT RUNNING (FIRST TEST) Unconditional error, blood pump not available General description of conditions for occurring: - The blood pump was commanded to run. The blood pump tacho value is too low AND the rotation indicator is not detecting movement (first check). Technical description of conditions for occurring: - PRI(PI_BlPumpEdge1ErrCnt) > 10 Some possible explanations why the error has occurred: - Blood pump not connected to BM I/O board, P602 - Pump rotation indicator (P605) AND tacho (P604) not connected to BM I/O - Blood pump electronics broken (BM I/O IC7, IC3, IC11, IC8, Q1-4 and components). C COFB 088 009 BLOOD PUMP TEST ERROR, PUMP NOT RUNNING (SECOND TEST) Unconditional error, blood pump not available General description of conditions for occurring: - The blood pump was commanded to run. The blood pump tacho value is too low AND the rotation indicator is not detecting movement (second check). Technical description of conditions for occurring: - PRI(PI_BlPumpEdge2ErrCnt) > 4 Some possible explanations why the error has occurred: - Blood pump not connected to BM I/O board, P602 - Pump rotation indicator (P605) AND tacho (P604) not connected to BM I/O - Blood pump electronics broken (BM I/O IC7, IC3, IC11, IC8, Q1-4 and components). |