FCN 1.18, BLOOD LEAK FUNCTION (CPU C) Purpose: * The blood leak transmitter sends light to the blood leak receiver. If the receiver detects too little light during the function checks, then an attention is given. * If the receiver detects too little light during treatment, then a blood leak alarm is given. * Resets the blood leak alarm when the operator presses the blood leak key. * Checks that a blood leak can be detected by reducing the light for 7s during the function checks and for 7s every 5 minutes during treatment.
FCN 1.18 000 SOFTWARE ERROR [F,T,D,S] Please refer to the description of FCN 1.00 000. FCN 1.18 001 BLOOD LEAK SENSOR TEST "REDUCED LIGHT" FAILED [F,T,D] General description of conditions for occurring: The blood leak receiver detected too much light when the blood leak transmitter was reduced.
Technical description of conditions for occurring: The blood leak transmitter was reduced to approximately 75%. Wait 7s. QRI(QI_BLOOD$LEAK) < 95. (QRI(QI_BLOOD$LEAK) = 0 when the fluid is completely transparent and it is about 120 when the fluid is completely dark.)
Some possible explanations why the error has occurred: * Incorrect blood leak detector calibration. * The blood leak receiver is broken. * The A/D system is broken. * The blood leak transmitter was never reduced.
Some hints of what to do if the error occurs: Check the blood leak detector calibration.
FCN 1.18 002 BLOOD LEAK AD OVERFLOW General description of conditions for occurring: The error code has been removed. Some possible explanations why the error has occurred: * Signal to AD system to high. ( > 60 mV.) * Resistor R5 on blood leak detector board to big, this will cause a signal that is to high for the AD converter. Re calibrate. * Blood leak sensor partially filled. The light beam was reflected in the fluid surface. * Strong infra red ambient light. * AD converter is not working properly.
FCN 1.18 003 CPU A IS NOT RESPONDING ON BLOOD LEAK DETECTOR ALARM General description of conditions for occurring: The error code has been removed.
FCN 1.18 004 BLOOD LEAK TEST ERROR DURING FUNCTION CHECK General description of conditions for occurring: The error code has been removed. The detector was not able to see that the infra red beam was lit. Technical description of conditions for occurring: QRI(QI_BLOOD$LEAK) > PW(EPW_BLOOD$LEAK$ON) during test.
Some possible explanations why the error has occurred: * Dirty blood leak detector. * The blood leak detector was not filled with fluid when this test was done. * Infra red diode not working properly. * Aborted heat disinfection might cause this message, due to a slight thermal change in the photo transistor.
HDF scale supervision function.
FCN 1.25 000 = Case error. The program is not working properly.
Profiling function.(Na profiling in Acetate mode and Na, HCO3 profiling in Bicarbonate mode)
FCN 1.26 000 = Case error. The program is not working properly. FCN 1.26 001 = Undefined Na type in Bicarbonate mode neither of (Off,Linear,Degressive,Progressive). FCN 1.26 002 = Undefined HCO3 type in Bicarbonate mode neither of (Off,Linear,Degressive,Progressive). FCN 1.26 003 = Undefined Na type in Acetate mode neither of (Off,Linear,Degressive,Progressive). FCN 1.26 004 = Division by zero or overflow error.
Profiling parameter set function.
FCN 1.27 000 = Case error. The program is not working properly. FCN 1.27 001 = Removed. Previously: No correct safe memory area for profiling settings. FCN 1.27 002 = Irrelevant profiling model selected. FCN 1.27 003 = Division by zero or overflow error.
HDF infusion pump rate supervision function.
FCN 1.28 000 = Case error. The program is not working properly.
HDF infusion pump rate calculation function.
FCN 1.29 000 = Case error. The program is not working properly. FCN 1.29 001 = The HDF infusion pump is running to fast. FCN 1.29 002 = A pump error has occurred . General description of conditions for occurring: Infusion pump running when blood pump is stopped (Not in EXTRA mode). Infusion pump stopped in EXTRA mode. Main Flow supervision function. (CPU B).
FCN 1.33 00x Error codes removed in Prom 2.00.
Main Flow supervision FCH function. (CPU B).
FCN 1.34 00x Error codes removed in Prom 2.00.
BiCart Select Flow Multiplexer Function (CPU C)
FCN 1.35 000 DIVISION BY ZERO OR OVER FLOW ERROR
FRN node controller Function (CPU C)
FCN 1.36 000 = Case error. The program is not working properly. [F,T,D,S]
FCN 1.36 001 FRN high flow measurement failed too many times [F] General description of conditions for occurring: The main flow measured by the FRN measurement step 1 differed more than +/- 13 ml/min from the flow measured by the UF-cell channel 1, high flow set value approximately 700 ml/min.
Some possible explanations why the error has occurred: * The FRN is not calibrated. * The FRN or the UF-cell is badly calibrated. * Program in the FRN node is exchanged.
Some hints of what to do if the error occurs: Calibrate the FRN. Calibrate the UF-cell.
FCN 1.36 002 FRN low flow measurement failed too many times [F] General description of conditions for occurring: The main flow measured by the FRN measurement step 2 differed more than +/- 13 ml/min from the flow measured by the UF-cell channel 1, low flow set value = high flow set value - 50 ml/min.
Some possible explanations why the error has occurred: * The FRN or the UF-cell is badly calibrated.
Some hints of what to do if the error occurs: Calibrate the FRN. Calibrate the UF-cell.
FCN 1.36 003 FRN diff flow measurement failed too many times [F] General description of conditions for occurring: The FRN main flow measurement differance between step 1 and step 2 differs more than +/- 4 ml/min from the differance set value 50 ml/min.
Some possible explanations why the error has occurred: * The FRN or the UF-cell is badly calibrated.
Some hints of what to do if the error occurs: Calibrate the FRN. Calibrate the UF-cell.
FRN node supervision Function (CPU B)
FCN 1.37 001 = Communication error between CPU C and FRN [F,T,D,S] node detected.
Some possible explanations why the error has occurred: * The cables between CPU C and the Echelon nodes are broken. * Programs in the Echelon nodes are not working properly. * Programs in the Echelon nodes are not installed.
Arterial pressure guard supervision. (CPU A).
FCN 2.00 000 = Case error. The program is not working properly.
Extra pressure supervision. (CPU A).
FCN 2.01 000 = Case error. The program is not working properly. FCN 2.01 001 = Test error. Extra pressure was not approx. 0 mm Hg during functional test, not even after attention.
Venous pressure supervision. (CPU A).
FCN 2.02 000 = Case error. The program is not working properly. FCN 2.02 001 = Test error. Venous pressure was not approx. 0 mm Hg during functional test, not even after attention.
Level detector function (CPU A).
FCN 2.03 000 = Case error. The program is not working properly. FCN 2.03 001 = Receiver trigged by short kick pulse. FCN 2.03 002 = PRS has not answered the synchronizing pulse. FCN 2.03 003 = CPU A can not synchronize with CPU B. FCN 2.03 004 = CPU B has not reset receiver latch.
Priming detector function. (CPU A).
FCN 2.04 000 = Case error. The program is not working properly. FCN 2.04 001 = PRS has not detected the missing "LED ON" pulse. FCN 2.04 002 = Can not synchronize with CPU B. FCN 2.04 003 = PRL(P_PRIM$DET$EMPTY),-TUBE,-BLOOD combination error.
Arterial tidal volume supervision. (CPU A).
FCN 2.05 000 = Case error. The program is not working properly.
Arterial pump supervision function. CPU A).
FCN 2.07 000 = Case error. The program is not working properly. FCN 2.07 001 = Technical error in the arterial pump. FCN 2.07 002 = Arterial pump head does not move when it should. FCN 2.07 003 = Arterial pump head move when it should not. FCN 2.07 004 = Arterial pump head does not move with right speed.
Venous pump supervision function. CPU A).
FCN 2.08 000 = Case error. The program is not working properly. FCN 2.08 001 = Technical error in the venous pump. FCN 2.08 002 = Venous pump head does not move when it should. FCN 2.08 003 = Venous pump head move when it should not. FCN 2.08 004 = Venous pump head does not move with right speed.
Preset conversion function. CPU A).
FCN 2.09 000 = Case error. The program is not working properly.
Optional pressure function. (CPU A).
FCN 2.10 000 = Case error. The program is not working properly. FCN 2.10 001 = Test error. Optional pressure was not approx. 0 mm Hg during functional test, not even after attention.
Pressure transducer connection supervision. (CPU A).
FCN 2.11 000 = Case error. The program is not working properly.
HDF scale supervisory function. (CPU A). FCN 2.12 000 = Case error. The program is not working properly.
Bargraphs. (CPU A).
FCN 3.00 000 = Case error venous pressure. FCN 3.00 003 = Function state conflict ( Not 0 during LED test ). FCN 3.01 000 = Case error blood flow. FCN 3.01 003 = Function state conflict ( Not 0 during LED test ). FCN 3.02 000 = Case error conductivity. FCN 3.02 003 = Function state conflict ( Not 0 during LED test ). FCN 3.03 000 = Case error temperature. FCN 3.03 003 = Function state conflict ( Not 0 during LED test ). FCN 3.04 000 = Case error extra pressure. FCN 3.04 003 = Function state conflict ( Not 0 during LED test ). FCN 3.05 000 = Case error TMP. FCN 3.05 003 = Function state conflict ( Not 0 during LED test ). FCN 3.06 000 = Case error UF volume. FCN 3.06 003 = Function state conflict ( Not 0 during LED test ). FCN 3.07 000 = Case error UF rate. FCN 3.07 003 = Function state conflict ( Not 0 during LED test ).
FCN BGPH 001 = LED broken : Venous press & Blood flow BGPH FCN BGPH 004 = Shortage : Venous press & Blood flow BGPH FCN BGPH 011 = LED broken : Cond & Temp BGPH FCN BGPH 014 = Shortage : Cond & Temp BGPH FCN BGPH 021 = LED broken : Extra press & TMP BGPH FCN BGPH 024 = Shortage : Extra press & TMP BGPH FCN BGPH 031 = LED broken : UF-vol & UF-rate BGPH FCN BGPH 034 = Shortage : UF-vol & UF-rate BGPH
UF gain control (CPU A)
FCN 3.08 000 = Case error. The program is not working properly.
Information display (CPU A)
FCN 3.09 xxx = Case error in main state xxx. FCN 3.09 256 = SVC error. FCN 3.09 258 = RIO error status=??. FCN 3.09 259 = RIO error status=1. FCN 3.09 260 = RIO error status=0. FCN 3.09 261 = RIO error status=-1. FCN 3.09 262 = RIO error status=-2. FCN 3.09 263 = RIO error status=-3. FCN 3.09 270 = reference to non-existant treatment text FCN 3.09 271 = reference to non-existant service text FCN 3.09 272 = reference to non-existant treatment language FCN 3.09 280 = nonexistant chem type in EEB_CH$TYPE$0 FCN 3.09 281 = nonexistant chem type in EEB_CH$TYPE$1 FCN 3.09 282 = nonexistant chem type in EEB_CH$TYPE$2 FCN 3.09 283 = nonexistant preset type in OB_PRESET$TEXT FCN 3.09 284 = nonexistant heating type in EEB_HEAT$TYPE$0 FCN 3.09 285 = nonexistant heating type in EEB_HEAT$TYPE$1 FCN 3.09 286 = nonexistant heating type in EEB_HEAT$TYPE$2 FCN 3.09 287 = nonexistant rinse type in EEB_RINSE$TYPE$0 FCN 3.09 288 = nonexistant rinse type in EEB_RINSE$TYPE$1 FCN 3.09 289 = nonexistant rinse type in EEB_RINSE$TYPE$2 FCN 3.09 981 = LLLA , Low limit LED. Nr : 43 broken FCN 3.09 982 = LLLA , Low limit LED. Nr : 43 shortage FCN 3.09 983 = SPLA , Set parameter LED. Nr : 44 broken FCN 3.09 984 = SPLA , Set parameter LED. Nr : 44 shortage FCN 3.09 985 = HLLA , High limit LED. Nr : 45 broken FCN 3.09 986 = HLLA , High limit LED. Nr : 45 shortage FCN 3.09 992 = LED test: two LEDs broken. FCN 3.09 993 = LED test: three LEDs broken. FCN 3.09 994 = LED test: one LED broken. FCN 3.09 995 = LED test: short circuit. FCN 3.09 996 = LED test: short circuit. FCN 3.09 997 = LED test: one or more LED(s) broken. FCN 3.09 998 = LED test: bad reply from CPU B. FCN 3.09 999 = Main Case error. The program is not working properly.
Access code control. (CPU A).
FCN 3.10 000 = Case error. The program is not working properly.
LED functions. (CPU A).
Function 3.11 to 3.44, 3.75, 3.76, 3.78 : Key LEDs.
FCN 3.nn 000 = Case error. The program is not working properly. FCN 3.nn 001 = LED broken or bad connection. FCN 3.nn 002 = Removed. FCN 3.nn 003 = Function state conflict ( Not 0 during LED test ). FCN 3.nn 004 = Shortage error.
FCN 3.12 nnn = L 3 Auto heat LED test. FCN 3.13 nnn = L 9 Auto rinse LED test. FCN 3.14 nnn = L 4 Isolated UF LED test. FCN 3.15 nnn = L 5 TMP mode LED test. FCN 3.16 nnn = L 6 UF mode LED test. FCN 3.17 nnn = L 10 Single needle mode LED test. FCN 3.18 nnn = L 11 Acetate mode LED test. FCN 3.19 nnn = L 12 Bicarbonate mode LED test. FCN 3.20 nnn = L 7 Chemical disinfection LED test. FCN 3.21 nnn = L 2 Heat disinfection LED test. FCN 3.22 nnn = L 22 Heparin pump LED test. FCN 3.23 nnn = L 14 Blood flow LED test. FCN 3.24 nnn = L 15 Single needle parameters LED test. FCN 3.25 nnn = L 17 Conductivity LED test. FCN 3.26 nnn = L 16 Temperature LED test. FCN 3.27 nnn = L 28 UF volume LED test. FCN 3.28 nnn = L 29 UF rate LED test. FCN 3.29 nnn = L 30 Level detector LED test. FCN 3.30 nnn = L 32 Attention LED test. FCN 3.31 nnn = L 18 Priming detector LED test. FCN 3.32 nnn = L 34 pH LED test. FCN 3.33 nnn = L 35 Blood leak LED test. FCN 3.34 nnn = L 21 Time LED test. FCN 3.35 nnn = L 31 Arterial guard LED test. FCN 3.36 nnn = L 8 Rinse / Drain LED test. FCN 3.37 nnn = L 13 Venous pressure LED test. FCN 3.38 nnn = L 25 Extra pressure LED test. FCN 3.39 nnn = L 26 TMP LED test. FCN 3.40 nnn = L 36 Blood pump LED test. FCN 3.41 nnn = L 38 Mute LED test. FCN 3.42 nnn = L 39 Fluid bypass LED test. FCN 3.43 nnn = L 41 Start / Stop LED test. FCN 3.44 nnn = L 42 Hold LED test. FCN 3.75 nnn = L 19 Profiling LED test. FCN 3.76 nnn = L 23 HDF LED test. FCN 3.78 nnn = L 24 HF LED test.
Remaining time calculation. (CPU A).
FCN 3.45 000 = Case error. The program is not working properly.
Passed time calculation. (CPU A).
FCN 3.46 000 = Case error. The program is not working properly.
Non-diffusion time (bypass time) calculation. (CPU A).
FCN 3.47 000 = Case error. The program is not working properly.
UF volume predict. (CPU A).
FCN 3.48 000 = Case error. The program is not working properly.
TMP - UF transition. (CPU A).
FCN 3.49 000 = Case error. The program is not working properly.
Blood flow path LEDs. (CPU A).
FCN 3.50 000 = Case error. The program is not working properly. FCN 3.50 003 = Function state conflict ( Not 0 during LED test ).
FCN 3.50 001 = LED broken : AFL1 , Arterial LED 1 Nr : 46 FCN 3.50 004 = Shortage : AFL1 , Arterial LED 1 Nr : 46 FCN 3.50 011 = LED broken : AFL2 , Arterial LED 2 Nr : 47 FCN 3.50 014 = Shortage : AFL2 , Arterial LED 2 Nr : 47 FCN 3.50 021 = LED broken : VFL1 , Venous LED 1 Nr : 48 FCN 3.50 024 = Shortage : VFL1 , Venous LED 1 Nr : 48 FCN 3.50 031 = LED broken : VFL2 , Venous LED 2 Nr : 49 FCN 3.50 034 = Shortage : VFL2 , Venous LED 2 Nr : 49
Fluid flowpath LEDs. (CPU A).
FCN 3.51 000 = Case error FCN 3.51 003 = Function state conflict ( Not 0 during LED test ).
FCN 3.51 001 = LED broken : FLOL, Fluid Left Ok LED Nr : 53 FCN 3.51 004 = Shortage : FLOL, Fluid Left Ok LED Nr : 53 FCN 3.51 011 = LED broken : FLAL, Fluid Left Alarm LED Nr : 54 FCN 3.51 014 = Shortage : FLAL, Fluid Left Alarm LED Nr : 54 FCN 3.51 021 = LED broken : BYOL, Bypass Ok LED Nr : 55 FCN 3.51 024 = Shortage : BYOL, Bypass Ok LED Nr : 55 FCN 3.51 031 = LED broken : BYAL, Bypass Alarm LED Nr : 56 FCN 3.51 034 = Shortage : BYAL, Bypass Alarm LED Nr : 56
FCN 3.51 041 = LED broken : FROL, Fluid Right Ok LED Nr : 57 FCN 3.51 044 = Shortage : FROL, Fluid Right Ok LED Nr : 57 FCN 3.51 051 = LED broken : FRAL, Fluid Right Alarm LED Nr : 58 FCN 3.51 054 = Shortage : FRAL, Fluid Right Alarm LED Nr : 58
FCN 3.51 061 = LED broken : DLOL, Dialyzer Left Ok LED Nr : 50 FCN 3.51 064 = Shortage : DLOL, Dialyzer Left Ok LED Nr : 50 FCN 3.51 071 = LED broken : DROL, Dialyzer Right Ok LED Nr : 51 FCN 3.51 074 = Shortage : DROL, Dialyzer Right Ok LED Nr : 51
FCN 3.51 081 = LED broken : DRAL, Dialyzer Right Al. LED Nr : 52 FCN 3.51 084 = Shortage : DRAL, Dialyzer Right Al. LED Nr : 52
Fluid bypass (DIVA bus). (CPU A).
FCN 3.52 000 = Case error. The program is not working properly.
Venous pressure auto alarm limits adjust. (CPU A).
FCN 3.53 000 = Case error. The program is not working properly.
Extra pressure auto alarm limits adjust. (CPU A).
FCN 3.54 000 = Case error. The program is not working properly.
TMP auto alarm limits adjust. (CPU A).
FCN 3.55 000 = Case error. The program is not working properly.
Time display. (CPU A).
FCN 3.56 000 = Case error. The program is not working properly.
Keyboard input. (CPU A).
FCN 3.57 000 = Case error. The program is not working properly. FCN 3.57 001 = Hardware error. The keyboard controller requests an interrupt but the CPU cannot determine which key was pressed. FCN 3.57 002 = Error in START/UF/STOP KEY handshaking with CPU B General description of conditions for occurring: START/UF/STOP key is stuck or does not respond. Manual data input. (CPU A).
FCN 3.58 000 = Main Case error. The program is not working properly. FCN 3.58 xxx = Case error in main state 0 - 255. FCN 3.58 256 = OI_UF$RATE$SET has become negative
Machine off. (CPU A).
FCN 3.59 000 = Case error. The program is not working properly. FCN 3.59 001 = Estimated heat time too large. FCN 3.59 002 = Estimated rinse time too large. FCN 3.59 003 = Machine still running after machine-off sequence. FCN 3.59 004 = clock alarm write error. FCN 3.59 005 = clock read error. FCN 3.59 006 = dummy alarm write error. FCN 3.59 007 = clock NODA write error. FCN 3.59 008 = clock COMP write error. FCN 3.59 009 = Auto on (heat/rinse) safe memory area checksum error. This area is set to default values when error. FCN 3.59 010 = Auto on (heat/rinse) safe memory week scheduler area checksum error. This area is set to default values when error. To correct error change time in week scheduler.
Main buzzer control. (CPU A).
FCN 3.60 000 = BM case error. The program is not working properly. FCN 3.60 001 = FM case error. The program is not working properly.
Blood potentiometer data input. (CPU A).
FCN 3.61 000 = Case error. The program is not working properly.
Heparin auto stop limit. (CPU A).
FCN 3.62 000 = Case error. The program is not working properly.
Linear transducer calibration calculation. (CPU A).
FCN 3.63 000 = Case error. The program is not working properly.
Real time calculation. (CPU A).
FCN 3.64 000 = Case error. The program is not working properly. FCN 3.64 002 = Error when writing YEAR in EEPROM. FCN 3.64 003 = CMOS realtime clock was not initialized. Automatic initialization follows, but previous settings are lost. FCN 3.64 004 = The time saved in battery powered RAM is corrupt. Previous settings are lost. FCN 3.64 1xy = Error accessing CMOS realtime clock over the I2C bus: x = type of access x = 1 reading time once every minute x = 2 reseting alarm time at startup x = 3 reseting NODA flag at startup x = 4 reseting COMP flag at startup x = 5 reading time at startup x = 6 writing valid time at startup y = I2C access error y = 1 I2C acknowledge error y = 2 I2C byte address error y = 9 I2C bus address error (bus 3 is not valid)
Select chemical mode. (CPU A).
FCN 3.65 000 = Case error. The program is not working properly.
Parallel E2prom function. (CPU A).
FCN 3.66 000 = Case error. The program is not working properly. FCN 3.66 001 = CRC error. FCN 3.66 002 = Preset variable out of range. FCN 3.66 003 = Too few variables in EEPROM according to program. Initiate the new variables in preset handling. FCN 3.66 004 = EEPROM read error. FCN 3.66 005 = EEPROM write error. FCN 3.66 006 = PC preset calculation not performed. FCN 3.66 007 = PC preset not complete. FCN 3.66 008 = Default preset error in EEL array. FCN 3.66 009 = Default preset error in EEB array.. FCN 3.66 010 = Default preset error in EEW array.. FCN 3.66 011 = Default preset error in EEI array.. FCN 3.66 012 = Default preset error in EER array.. FCN 3.66 013 = PC preset to CPU D not performed. FCN 3.66 014 = PC preset to CPU B not performed.
Internal support function. (CPU A).
FCN 3.67 000 = Case error. The program is not working properly.
Battery test function. (CPU A).
FCN 3.68 000 = Case error. The program is not working properly. FCN 3.68 001 = The 12V winding in the AC/DC unit is broken. FCN 3.68 002 = The "BATTERY FAILURE" attention has been confirmed more than 3 times. Change or recharge battery.
LED test function. (CPU A).
FCN 3.69 000 = Case error LED test function. FCN 3.69 001 = LED port out of range [0..7] LED test
Preset calculation function. (CPU A). FCN 3.70 000 = Case error. The program is not working properly.
Conductivity function. (CPU A).
FCN 3.71 000 = Case error. The program is not working properly. FCN 3.71 001 = Bicarbonate constant comparation error. Protective system and CPU A differ in constants.
Concentrate preset mode/alternative function. (CPU A). FCN 3.72 000 = Case error. The program is not working properly.
Attention function. (CPU A).
FCN 3.73 000 = Case error. The program is not working properly.
Bicarbonate mode time measurement function. (CPU A).
FCN 3.74 000 = Case error. The program is not working properly.
More LEDs. (CPU A).
FCN 3.75 nnn = Profiling LED. (See 3.11 to 3.44.) FCN 3.76 nnn = HDF LED. (See 3.11 to 3.44.) FCN 3.78 nnn = HF LED. (See 3.11 to 3.44.)
Profiling parameter set function.
FCN 3.79 000 = Case error. The program is not working properly. FCN 3.79 001 = No correct safe memory area for profiling settings. FCN 3.79 002 = Irrelevant profiling model selected.
Conductivity profiling function.
FCN 3.80 000 = Case error. The program is not working properly. FCN 3.80 001 = NA undefined profile type. FCN 3.80 002 = HCO3 undefined profile type. FCN 3.80 003 = NA AC undefined profile type.
Convert function. Convert preset variables.
FCN 3.81 000 = Case error. The program is not working properly.
UF equation and UF gain utilities (UF profiling).
FCN 3.82 000 = Case error. The program is not working properly. FCN 3.82 001 = UF undefined profile type.
System pressure alarm limit controller.
FCN 3.83 000 = Case error. The program is not working properly.
Treatment time handling function in protective system.(CPU B)
FCN 3.84 000 = Case error. The program is not working properly.
GSS LED. (CPU A).
FCN 3.86 nnn = GSS LED. (See 3.11 to 3.44.)
Negative UF supervision. (CPUA).
FCN 3.89 000 = Case error. The program is not working properly. BM AD function. (CPU A).
ADC INTR 000 = AD-buffer full error
FCN 4.00 000 = Undefined AD case error FCN 4.00 001 = Undefined sequence selection FCN 4.00 002 = Multiple AD-restarts FCN 4.00 003 = Ref voltage out of range FCN 4.00 004 = Offset voltage out of range FCN 4.00 005 = Illegal sensitivity coeff FCN 4.00 006 = Ref voltage limits start up check
FM AD function. (CPU D).
ADC INTR 001 = A/D timer load counter error call from interrupt state. ADC INTR 002 = A/D timer mode load hold register test error call from interrupt state. ADC 4.01 0xx = More then 20% overrange on channel xx Channel xx Explanation 0 MAIN PC BOARD TEMPERATURE 1 MOTOR CURRENT 2 VALVE CURRENT 3 COND. CONTROL ACETATE/TOTAL 4 DEGASS PRESSURE 5 HEATING VESSEL TEMPERATURE 6 COND. CONTROL BCM OPTION 7 TEMP BCM COND. TRANSDUCER 8 UF TRANSDUCER COIL CURRENT (POSITIVE SYNC) 9 HIGH PRESSURE GUARD (HPGS) 10 DIALYSIS PRESSURE (PD) 11 AUX INPUT 12 AUX INPUT 13 AUX INPUT 14 OFFSET VOLTAGE AD-SYSTEM 15 REFERENCE VOLTAGE AD-SYSTEM
FCA 4.01 001 = A/D timer load counter error call from ad-task. FCA 4.01 002 = A/D timer mode load hold register test error call from ad-task. FCB 4.01 xxx = AD conversion stop for xxx sec. FCN 4.01 000 = Case error. The program is not working properly. FCN 4.01 001 = Timer error. FCN 4.01 002 = AD converter offset out of limits. -10 mV < offset (= 0.0 mV) < + 10 mV FCN 4.01 003 = AD converter reference out of limits. 45.45 mV < reference ( = 49.50 mV) < 54.89 mV FCN 4.01 004 = AD-system timeout 1 sec. FCN 4.01 005 = AD buffer full. FCN 4.01 006 = AD buffer full and empty.
pH measuring function. (CPU D).
FCN 4.02 000 = Case error. The program is not working properly. FCN 4.02 001 = Timer problem detected. FCN 4.02 F00 = init timer error. FCN 4.02 F01 = overflow error. (Currently not used after prom 3.40) FCN 4.02 F02 = read timer error.
Clamp controller. (CPU A).
FCN 4.03 000 = Case error. The program is not working properly. FCN 4.03 001 = Clamp verify error. Venous clamp was not opened. FCN 4.03 002 = Clamp verify error. Arterial clamp was not opened. FCN 4.03 003 = Clamp verify error. Arterial clamp was not closed. FCN 4.03 004 = Clamp verify error. Venous clamp was not closed. FCN 4.03 005 = Venous clamp does not close during functional test. FCN 4.03 006 = Arterial clamp does not open during functional test. FCN 4.03 007 = Venous clamp does not open during functional test. FCN 4.03 008 = Arterial clamp does not close during functional test.
Flow generator transducer calibration function. (CPU D).
FCN 4.04 000 = Case error. The program is not working properly.
FCN 4.05, FLOW MEASURING FUNCTION (CPU D) Purpose: * NORMAL OPERATION: The signal TMO0 is used for the synchronization of the UF flow measurement. TMO0 is a 401.44 Hz square wave (50% duty cycle) with TTL levels. It switches the magnetic field in the UF measuring unit and controls the sampling of the received flowsignals. The two measured flow signals are amplified at the UF preamplifier board. The flow signals then passes multiplexers, are amplified and sampled on the FM CPU board. The sampled voltages are converted to frequen- cies that are measured by two counters that both are contained in IC94 on the FM CPU board. FCN 4.05 calibrates the counter values and creates the variables FRR(FR_CH1$RATE) and FRR(FR_CH2$RATE). FRR(FR_CH1$RATE) is the flow through channel 1 and it has the unit ml/min. FRR(FR_CH2$RATE) is the flow through channel 2. FRR(FR_UF$RATE) is calculated as the difference between the flow through channel 2 and the flow through channel 1. FRR(FR_UF$RATE) has the unit 1 l/h. The UF volume FRR(FR_UF$VOL) is calculated. It's unit is 1 l. * EVERY 30 MINUTES: A self calibration is done twice an hour. The normal operation of the self calibration is as follows: 1 The measuring system is calibrated by first grounding the input and then applying a reference voltage to the input. 2 ZEVA is opened and DIVA/BYVA are closed so that no flow passes the flow cell. The zero flow signals are measured for 15 seconds and the program checks that the signals are stable. Flow transducer offsets are calculated. If necessary the measurement is prolonged for 15s. 3 BYVA is opened and DIVA/ZEVA are closed so that the same flow passes through flow channel 1 and flow channel 2. The flow signals are measured after 15s. The measurement continues for 15s and the program checks that the signals are stable. If necessary the measurement is prolonged for 15s. The program checks that the parameters FRR(FR_UFR$TOLERANCE) and FRR(FR_UFR$OFFSET) have not changed too much since the last self calibration. If a self calibration fails, then it is repeated after 5 minutes. * VALVE LEAKAGE TEST : The direct valve, the bypass valve and the zeroing valve are leakage tested during the function checks. DIVA, BYVA and ZEVA are closed. The flow output pump tries to force fluid through the closed valves, and the flow measuring function FCN 4.05 checks that no flow is leaking through them.
* CALIBRATION: The calibration is done in service mode, and it consists of the following steps:
1 Tempered fluid flows through both the flow channels for 10 minutes. 2 The linearity of the UF measuring unit is checked by letting three different flows flow through both the channels. 3 The offsets are calculated by not having any flow through the channels. 4 The channel 1 coefficient is calculated by letting the flow through channel 1 fill a reference volume. 5 The channel 2 coefficient is calculated by having the same flows through both the channels. 6 The calibration parameters are verified.
If "Internal" mode is selected to update offset's and channel 2 coefficient then only steps 1,3 and 5 is performed.
NOTE! The channel 1 flow signal is forwarded through the VFC channel 2 and vice versa. NOTE! The most probable error explanations are described the first.
FCN 4.05 000 SOFTWARE ERROR [F,T,D,S] Please refer to the description of FCN 1.00 000.
FCN 4.05 001 INVALID CHANGE OF MUX ADDRESS [F,T,D,S] General description of conditions for occurring: The UF measurement multiplexers are not selecting the correct data. Technical description of conditions for occurring: The address signals UIS0 and UIS1 for the multiplexer IC3 on the FM CPU board have incorrect values OR The address signals UIS2 and UIS3 for the multiplexer IC9 have incorrect values.
Some possible explanations why the error has occurred: * FM CPU BOARD: The Multiplexer, IC3,IC9,IC11 is broken. * FM CPU BOARD: The IO-PORT, IC102 is broken. * FM CPU BOARD: The RAM, IC98 is broken. * SOFTWARE: The program is not executing correctly. * SOFTWARE: The software code is incorrect.
FCN 4.05 002 TIMER ERROR [F,T,D,S] Please refer to the description of FCN 5.00 001.
Some possible explanations why the error has occurred: * FM CPU BOARD: The timer IC94 is broken.
FCN 4.05 003 VFC CHANNEL 1 PULSE COUNTER OVERFLOW [F,T,D,S] General description of conditions for occurring: The pulse counter for VFC channel 1 has counted up to its maximum value. Note! VFC channel 1 takes care of flow measurement signals CH2P and CH2N! Technical description of conditions for occurring: The pulse counter for VFC channel 1 has counted up to its maximum value at 5 following attempts OR The pulse counter has counted up to a value greater than 32768 at 5 following attempts. The value of the pulse counter = SRR(6) and has the unit 0.08 ml/min. The counter was set to 0 after each time the error occurred.
Some possible explanations why the error has occurred: * FM CPU BOARD: The VFC, IC36 or some of the components around it are broken.
FCN 4.05 004 VFC CHANNEL 2 PULSE COUNTER OVERFLOW [F,T,D,S] Please refer to the description of FCN 4.05 003 The value of the pulse counter = SRR(11) and has the unit 0.08 ml/min. Note! VFC channel 2 takes care of flow measurement signals CH1P and CH1N.
FCN 4.05 005 TIMER ERROR [F,T,D,S] Please refer to the description of FCN 5.00 001.
Some possible explanations why the error has occurred: * FM CPU BOARD: The timer IC94 on the FM CPU board is broken. |