RxL/XL 出错号,出错信息及相应故障排除方法表格的相互查找一览表
为帮助查找如何排除故障的信息,编制了出错号,出错信息及相应的解决办法参照表。
表中的第三栏给出了本章中对应于出错号的列出故障排除方法的相应表格。
表2-3. 出错号,出错信息及相应故障排除方法表格的相互查找一览表
¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾
出错号 出错信息 相应故障排除方法表格
¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾
001 样本针(上下)传感器问题或采样针不能动 2-4
Sample probe sensor problem or probe did not move(sop) ¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾
002 样本针(上下)传感器不能开通 2-5
Sample probe sensor did not turn on(cfh)
¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾
003 样本针非法出错号 2-7
Sample probe illegal error code
¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾
004 样本针(上下)传感器问题或采样针不能动 2-4
Sample probe sensor problem or probe did not move(soh)
¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾
005 样本针(上下)马达丢步 2-6
Sample probe lost steps
¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾
006 样本针(上下)马达丢步 2-6
Sample probe lost steps
¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾
007-012 这些出错号在XL/RxL内不用
¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾
013 100mL样本注射器泵传感器问题或注射器不动 2-4
Sample 100mL syringe pump sensor problem or syringe did not move(sop)
¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾
014 100mL样本注射器泵传感器不开通 2-5
Sample 100mL pump sensor did not turned on(cfh)
¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾
015 样本测量泵非法出错号 2-7
Sample small pump illegal error code
¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾
016 100mL样本注射器泵传感器问题或注射器不动 2-4
Sample 100mL syringe pump sensor problem or syringe did not move(soh)
¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾
017 100mL样本注射器泵马达丢步 2-6
Sample 100mL syringe lost steps
¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾
018 100mL样本注射器泵马达丢步 2-6
Sample 100mL syringe lost steps
¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾
019 R2 500mL注射器泵传感器问题或注射器不动 2-4
R2 500mL syringe pump sensor problem or syringe did not move(sop)
¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾
020 R2 500mL注射器泵传感器不开通 2-5
R2 500mL syringe pump sensor did not turned on(cfh)
¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾
¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾
出错号 出错信息 相应故障排除方法表格
¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾
021 R2 测量泵非法出错号 2-7
R2 small pump illegal error code
¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾
022 R2 500mL注射器泵传感器问题或注射器不动 2-4
R2 500mL syringe pump sensor problem or syringe did not move(soh)
¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾
023 R2 500mL注射器泵马达丢步 2-6
R2 500mL syringe lost steps
¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾
024 R2 500mL注射器泵马达丢步 2-6
Sample 100mL syringe lost steps
¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾
025 样本臂(转动)传感器问题或样本臂不动 2-4
¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾
026 样本臂(转动)传感器不开通 2-5
Sample arm sensor did not turned on(cfh)
¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾
027 样本臂非法出错号 2-7
Sample arm illegal error code
¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾
028 样本臂(转动)传感器问题或样本臂不动 2-4
Sample arm sensor problem or arm did not move(soh)
¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾
029 样本臂(转动)马达丢步 2-6
Sample arm lost steps
¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾
030 样本臂(转动)马达丢步 2-6
Sample arm lost steps
¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾
031 光电表传感器问题或光电表不动 2-4
Photometer sensor problem or photometer did not move(sop)
¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾
032 光电表传感器不开通 2-5
Photometer sensor did not turned on(cfh)
¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾
033 光电表非法出错号 2-7
Photometer illegal error code
¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾
034 光电表传感器问题或光电表不动 2-4
Photometer sensor problem or photometer did not move(soh)
¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾
035 光电表马达丢步 2-6
Photometer lost steps
¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾
036 光电表马达丢步 2-6
Photometer lost steps
¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾
037 2500mL样本注射器泵传感器问题或注射器不动 2-4
Sample 2500mL syringe pump sensor problem or syringe did not move(sop)
¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾
038 2500mL样本注射器泵传感器不开通 2-5
Sample 100mL pump sensor did not turned on(cfh)
¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾
¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾
出错号 出错信息 相应故障排除方法表格
¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾
039 样本清洗泵非法出错号 2-7
Sample large pump illegal error code
¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾
040 2500mL样本注射器泵传感器问题或注射器不动 2-4
Sample 2500mL syringe pump sensor problem or syringe did not move(soh)
¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾
041 2500mL样本注射器泵马达丢步 2-6
Sample 2500mL syringe lost steps
¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾
042 2500mL样本注射器泵马达丢步 2-6
Sample 2500mL syringe lost steps
¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾
043 R2 2500mL注射器泵传感器问题或注射器不动 2-4
R2 2500mL syringe pump sensor problem or syringe did not move(sop)
¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾
044 R2 2500mL注射器泵传感器不开通 2-5
R2 2500mL syringe pump sensor did not turned on(cfh)
¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾
045 R2 清洗泵非法出错号 2-7
R2 large pump illegal error code
¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾
046 R2 2500mL注射器泵传感器问题或注射器不动 2-4
R2 2500mL syringe pump sensor problem or syringe did not move(soh)
¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾
047 R2 2500mL注射器泵马达丢步 2-6
R2 2500mL syringe pump lost steps
¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾
048 R2 2500mL注射器泵马达丢步 2-6
R2 2500mL syringe pump lost steps
¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾
049 R2 臂(转动)传感器问题或R2 臂不动 2-4
R2 arm sensor problem or R2 arm did not move(sop)
¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾
050 R2 臂(转动)传感器不开通 2-5
R2 arm sensor did not turned on(cfh)
¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾
051 R2 臂非法出错号 2-7
R2 arm illegal error code
¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾
052 R2 臂(转动)传感器问题或R2 臂不动 2-4
R2 arm sensor problem or R2 arm did not move(soh)
¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾
053 R2 臂(转动)马达丢步 2-6
R2 arm lost steps
¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾
054 R2 臂(转动)马达丢步 2-6
R2 arm lost steps
¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾¾
055 样本盘传感器问题 2-4
Segment wheel sensor problem(sop)