找回密码
 立即注册
搜索
yeec近年来原创帖合集 本站基础知识下载汇总 yeec网站学习币充值链接 学习中心正式上线

Abacus血球V1.7 & 2.x通讯格式协议

[复制链接]
赵光 发表于 2007-12-13 19:29 | 显示全部楼层 |阅读模式

注册登录才能更好的浏览或提问。

您需要 登录 才可以下载或查看,没有账号?立即注册

×

    Serial Communication Protocol (V1.7 & 2.x)
General Description
The Abacus/Arcus software is able to make serial connection link to a receiver device (e.g.
computer) which is connected to the serial port. If the Baud Rate setting in the Service Menu
is set to a valid value (other than Not Connected), the instrument will try to initiate a
communication sequence.
The serial communication is always initiated by Abacus/Arcus, and the receiver must
respond to this.
The hardware protocol: 8 data bit, 1 stop bit, no parity. If the other side is not responding in
1 second, the instrument will repeat the transmission twice, but if still no response, the other
side will be supposed to be not ready to receive data, and thus Abacus/Arcus will not try to
communicate any more.
From this state it can wake up by receiving an <ENQ> (ASCII code 5) character from the
other side, and the instrument will immediately respond an <ACK> (ASCII code 6). From this
point the instrument will send data if it is ready for transmission.
The communication is based on packages. There are 5 package types:
• INIT package: Device identification, software version, current date and time.
• DATA package: Sample and patient information, measured parameters, and markers.
• RBC package: Sample information, RBC histogram.
• WBC package: Sample information, WBC histogram.
• PLT package: Sample information, PLT histogram.
The communication sequence is always started with an INIT package. If the link is successful,
the instrument will send DATA package, and the receiver can request RBC, WBC and PLT
packages at acknowledge.
Special characters used in the communication:
Character ASCII code
<SOH> 1
<STX> 2
<ETX> 3
<EOT> 4
<ENQ> 5
<ACK> 6
<HT> 9
<LF> 10
<NAK> 21
<SPACE> 32
Abacus/Arcus Service Manual 59
Format of Packages Sent by Abacus/Arcus
The packages sent by Abacus/Arcus are always between a <SOH> and <EOT> character,
and they consist of header, data and tail. The header consists of a package identifier and a
package type descriptor. The tail includes the checksum. Typical format:
<SOH>MID CMD<STX>MESSAGE<ETX>CHKSUM<EOT>
MID: Message ID, one capital letter between ’A’ and ’Z’
CMD: Command, one capital letter, package type descriptor
• ’I’ for INIT;
• ’D’ for DATA;
• ’R’ for RBC;
• ’W’ for WBC;
• ’P’ for PLT.
MESSAGE: the message consists of ASCII characters (between 32..128)
CHKSUM: two hexadecimal digits of a one-byte checksum which contains
the lower byte of the sum of characters between the <SOH>
and <ETX> (incl. those, too).
Format of Acknowledge of the Receiver
The receiver must acknowledge transmission by the following message:
<ACK>CMD MID
CMD: the type of the next package to send – this makes possible to a
request for histograms. If no more packages are required,
<SPACE> should be sent.
MID: the identifier of the package that acknowledged by this
message.
If receiving of the message was not successful, <NAK> should be sent, and Abacus/Arcus
will repeat the last package.
The receiver has approx. 1 second to reply, otherwise Abacus/Arcus will repeat the last
transmission, twice automatically. But if still no response after 3 trials, Abacus/Arcus will not
start to communicate any more, even if there is data to send (see General Description).
Detailed Description of Packages
The packages sent by Abacus/Arcus are placed between a header and a tail. There are 3
main types of them: INIT, DATA and histogram packages. The histogram package can be:
RBC, WBC, and PLT. Their format is the same, the type descriptor differs only.
INIT package
Abacus/Arcus sends the general identifiers by this package during initiation of a
communication:
DEVICE<HT>VERSION<HT>DATE<HT>TIME
DEVICE: device identifier
VERSION: version of the software, e.g. „1.7”
DATE: date in YYYYMMDD format
TIME: time in HHMMSS format
Abacus/Arcus Service Manual 60
DATA package
This package contains sample and patient data, the measured parameters and the markers
of histograms. One data element consists of two items: name and value, which are
separated by a <HT> character, and closed by <LF>.
Important! SNO, DATE, TIME, SID and PID fields will always be present in this order,
but other fields may be omitted.
SNO<HT>152<LF> internal identifier
DATE<HT>19980715<LF> date of measurement (YYYYMMDD)
TIME<HT>114500<LF> time of measurement (HHMMSS)
SID<HT>1AXX435<LF> sample identifier (alphanumeric)
PID<HT>B72D44<LF> patient identifier (alphanumeric)
NAME<HT>JOE SMITH<LF> patient name (alphanumeric)
BIRTH<HT>19650208<LF> birthdate of patient (YYYYMMDD)
SEX<HT>0<LF> patient sex (0-male, 1-female)
DOC<HT>Dr John Gold<LF> doctor's name
OPID<HT>1172<LF> operator identifier (alphanumeric)
MODE<HT>0<LF> patient type
WRN<HT>0<LF> warning bits1 (32-bit hexadecimal)
PM1<HT>12<LF> PLT lower marker2
PM2<HT>204<LF> PLT upper marker2
RM1<HT>51<LF> RBC lower marker2
WM1<HT>23<LF> WBC upper marker2
WM2<HT>57<LF> WBC LYM-MID marker2
WM3<HT>92<LF> WBC MID-GRA marker2
PARN<HT>22<LF> number of parameters to send
P01<HT> 6.6<HT>0<LF> first parameter ID3, value and flag4
P02<HT>4.29<HT>0<LF> second parameter ID, value and flag
...
P22<HT> 8.2<HT>0<LF> last parameter ID, value and flag
Remarks:
1 Position and meaning of the warning flags:
Bit Mask Letter Meaning
0 0x00001 c RBC/PLT clogging.
1 0x00002 s RBC/PLT time error.
2 0x00004 d RBC/PLT data package errors.
3 0x00008 m RBC/PLT coincidence is too high. Linearity error.
4 0x00010 k RBC peek warning.
5 0x00020 l RBC/PLT limit warning.
6 0x00040 b RBC blank is high, or no RBC blank.
7 0x00080 p PLT blank is high, or no PLT blank.
8 0x00100 C WBC clogging.
9 0x00200 S WBC time error.
10 0x00400 D WBC data package errors.
11 0x00800 M WBC coincidence is too high. Linearity error.
Abacus/Arcus Service Manual 61
12 0x01000 R Too many RBC cut from WBC.
13 0x02000 L WBC/RBC limit warning.
14 0x04000 B WBC blank is high, or no WBC blank.
15 0x08000 H HGB blank is high, or no HGB blank.
16 0x10000 E No WBC three part.
17 0x20000 W WBC three part warning or WBC three part diff.
unsuccessful.
2 The markers are given in histogram channel between 0 and 255, where 0 means
that the marker could not be found.
3 The parameter IDs in Abacus/Arcus are: P01: WBC (109/l); P02: RBC (1012/l); P03:
HGB (g/l); P04: HCT (%); P05: MCV (fl); P06: MCH (pg); P07: MCHC (g/l); P08:
PLT (109/l); P09: PCT (%); P10: MPV (fl); P11: PDWsd (fl); P12: PDWcv (%); P13:
RDWsd (fl); P14: RDWcv (%); P15: LYM (109/l); P16: MID (109/l); P17: GRA (109/l);
P18: LYM% (%); P19: MID% (%); P20: GRA% (%); P21: RBCtime (sec); P22:
WBCtime (sec).
4The parameter value is always 4 character wide, spaces from left added if
necessary. It can be 9999, if the value could not be displayed in 4 digits, or ---- if
the value could not be calculated because of an error.
The meaning of the parameter flags:
Flag Displayed Meaning
0 Value correct.
1 + Value high (more than upper limit)
2 − Value low (less than lower limit)
3 * Value is unreliable.
4 E Value not given because of error. Value is ----.
5 Value cannot be calculated. There is no value!
RBC, WBC and PLT package
The format of the histogram packages are the same, they differ in type descriptor only. The
package contains the sample data (to identify the package) and the histogram.
SNO<HT>152<LF> internal identifier
DATE<HT>19980715<LF> date of measurement (YYYYMMDD)
TIME<HT>114500<LF> time of measurement (HHMMSS)
SID<HT>1AXX435<LF> sample identifier (alphanumeric)
PID<HT>B72D44<LF> patient identifier (alphanumeric)
CHN<HT>256<LF> number of histogram channels (256)
9<HT> value of first histogram channel
...
1<HT> value of 255th histogram channel
0 value of the last histogram channel
Abacus/Arcus Service Manual 62
Example
Let us suppose that the receiver was not ready to receive when Abacus/Arcus was switched
on, therefore the instrument is not starting a communication.
Receiver sends that it is ready to receive now, Abacus/Arcus sends acknowledge.
<ENQ>
<ACK>
Abacus/Arcus sends INIT next time to initiate link (there is data to send):
V1.7 software:
<SOH>AI<STX>DiaTron Abacus<HT>1.7g<HT>20010405<HT>163902<ETX>88<EOT>
V2.x software:
<SOH>AI<STX>Abacus<HT>2.15a<HT>20010405<HT>160734<ETX>E4<EOT>
Receiver sends acknowledge:
<ACK><SPACE>A
Abacus/Arcus sends DATA package:
<SOH>BD<STX>SNO<HT>1<LF>DATE<HT>20010405<LF>TIME<HT>163800<LF>SID<HT>1<LF>
PID<HT><LF>NAME<HT><LF>BIRTH<HT>00000000<LF>SEX<HT>0<LF>DOC<HT><LF>OPID<H
T>0<LF>MODE<HT>0<LF>WRN<HT>0<LF>PM1<HT>10<LF>PM2<HT>135<LF>RM1<HT>34<LF>
WM1<HT>21<LF>WM2<HT>55<LF>WM3<HT>91<LF>PARN
<HT>22<LF>P01<HT>6.4<HT>0<LF>P02<HT>4.36<HT>0<LF>P03<HT>131<HT>0<LF>
P04<HT>39.0<HT>0<LF>P05<HT>90<HT>0<LF>P06<HT>30.0<HT>0<LF>P07<HT>335
<HT>0<LF>P08<HT>262<HT>0<LF>P09<HT>0.24<HT>0<LF>P10<HT>9.2<HT>0<LF>
P11<HT>12.0<HT>0<LF>P12<HT>36.9<HT>0<LF>P13<HT>52.3<HT>1<LF>P14<HT>
16.3<HT>0<LF>P15<HT>2.1<HT>0<LF>P16<HT>0.5<HT>0<LF>P17<HT>3.7<HT>0<LF>P18<HT>
33.4<HT>0<LF>P19<HT>7.8<HT>1<LF>P20<HT>58.8<HT>0<LF>P21<HT>
8.4<HT>0<LF>P22<HT> 5.0<HT>0<LF><ETX>89<EOT>
Receiver acknowledges DATA transmission, and requests for RBC package:
<ACK>RB
Abacus/Arcus sends RBC package:
<SOH>CR<STX>SNO<HT>1<LF>DATE<HT>20010405<LF>TIME<HT>163800<LF>SID<HT>1<LF>
PID<HT>57290<LF>CHN<HT>256<LF>4<HT>9<HT>17<HT>27<HT>37<HT>48<HT>58<HT>66<H
T>70<HT>72<HT>70<HT>66<HT>60<HT>54<HT>47<HT>39<HT>33<HT>28<HT>22<HT>18<HT>1
5<HT>11<HT>9<HT>8<HT>7<HT>5<HT>4<HT>4<HT>3<HT>3<HT>2<HT>2<HT>1<HT>1<HT>0<
HT>0<HT>0<HT>0<HT>0<HT>0<HT>0<HT>0<HT>0<HT>0<HT>0<HT>0<HT>0<HT>0<HT>0<HT>
0<HT>0<HT>0<HT>0<HT>0<HT>0<HT>0<HT>1<HT>1<HT>1<HT>2<HT>2<HT>3<HT>4<HT>5<H
T>6<HT>7<HT>8<HT>9<HT>10<HT>11<HT>12<HT>14<HT>16<HT>18<HT>21<HT>24<HT>28<H
T>31<HT>35<HT>38<HT>43<HT>47<HT>53<HT>58<HT>65<HT>71<HT>78<HT>86<HT>94<HT>1
03<HT>113<HT>121<HT>130<HT>139<HT>149<HT>160<HT>171<HT>184<HT>196<HT>207<HT
>217<HT>225<HT>231<HT>237<HT>242<HT>245<HT>248<HT>251<HT>252<HT>253<HT>253<
HT>253<HT>253<HT>253<HT>254<HT>255<HT>254<HT>252<HT>248<HT>243<HT>236<HT>22
9<HT>221<HT>214<HT>207<HT>199<HT>192<HT>185<HT>178<HT>170<HT>162<HT>154<HT>
146<HT>137<HT>129<HT>121<HT>113<HT>104<HT>96<HT>88<HT>82<HT>75<HT>71<HT>66<
HT>62<HT>58<HT>55<HT>51<HT>47<HT>43<HT>38<HT>35<HT>31<HT>28<HT>25<HT>22<HT
>19<HT>18<HT>16<HT>14<HT>12<HT>11<HT>10<HT>9<HT>9<HT>9<HT>9<HT>8<HT>8<HT>7
<HT>7<HT>6<HT>6<HT>5<HT>5<HT>5<HT>4<HT>4<HT>3<HT>3<HT>3<HT>2<HT>2<HT>2<HT
>1<HT>1<HT>1<HT>1<HT>1<HT>1<HT>1<HT>2<HT>2<HT>2<HT>2<HT>2<HT>2<HT>2<HT>1<
HT>1<HT>1<HT>1<HT>0<HT>0<HT>0<HT>0<HT>0<HT>0<HT>0<HT>0<HT>0<HT>0<HT>0<HT>
0<HT>0<HT>0<HT>0<HT>0<HT>0<HT>0<HT>0<HT>0<HT>0<HT>0<HT>0<HT>0<HT>0<HT>0<H
T>0<HT>0<HT>0<HT>0<HT>0<HT>0<HT>0<HT>0<HT>0<HT>0<HT>0<HT>0<HT>0<HT>0<HT>0<
HT>0<HT>0<HT>0<HT>0<HT>0<HT>0<HT>0<HT>0<HT>0<HT>0<HT>0<HT>0<HT>0<ETX>60<E
OT>
Receiver acknowledges RBC transmission, and requests for WBC package:
<ACK>WC
Abacus/Arcus Service Manual 63
Abacus/Arcus sends WBC package:
<SOH>DW<STX>SNO<HT>1<LF>DATE<HT>20010405<LF>TIME<HT>163800<LF>SID<HT>1<LF
>PID<HT>57290<LF>CHN<HT>256<LF>0<HT>0<HT>0<HT>0<HT>0<HT>0<HT>0<HT>0<HT>0<H
T>0<HT>255<HT>255<HT>255<HT>251<HT>209<HT>163<HT>118<HT>82<HT>56<HT>40<HT>3
2<HT>29<HT>29<HT>30<HT>34<HT>40<HT>47<HT>55<HT>64<HT>76<HT>88<HT>103<HT>119
<HT>136<HT>155<HT>174<HT>195<HT>215<HT>233<HT>246<HT>254<HT>255<HT>249<HT>2
40<HT>227<HT>213<HT>200<HT>187<HT>176<HT>164<HT>153<HT>141<HT>129<HT>118<HT
>108<HT>99<HT>92<HT>86<HT>81<HT>76<HT>71<HT>67<HT>64<HT>62<HT>61<HT>60<HT>6
0<HT>60<HT>57<HT>55<HT>51<HT>47<HT>44<HT>43<HT>42<HT>42<HT>43<HT>44<HT>45<
HT>47<HT>47<HT>46<HT>45<HT>43<HT>40<HT>37<HT>34<HT>32<HT>30<HT>30<HT>30<HT
>30<HT>29<HT>28<HT>26<HT>24<HT>21<HT>19<HT>17<HT>15<HT>14<HT>13<HT>12<HT>13
<HT>13<HT>13<HT>13<HT>13<HT>12<HT>11<HT>11<HT>11<HT>12<HT>12<HT>14<HT>15<H
T>16<HT>17<HT>17<HT>17<HT>17<HT>17<HT>17<HT>17<HT>17<HT>17<HT>18<HT>19<HT>2
1<HT>22<HT>23<HT>24<HT>25<HT>25<HT>25<HT>26<HT>29<HT>31<HT>34<HT>35<HT>36<
HT>36<HT>35<HT>35<HT>35<HT>37<HT>38<HT>41<HT>43<HT>46<HT>47<HT>48<HT>50<HT
>52<HT>56<HT>60<HT>64<HT>70<HT>74<HT>79<HT>82<HT>86<HT>89<HT>92<HT>95<HT>99
<HT>102<HT>103<HT>104<HT>105<HT>105<HT>106<HT>108<HT>110<HT>112<HT>115<HT>1
17<HT>119<HT>120<HT>121<HT>120<HT>120<HT>120<HT>121<HT>122<HT>125<HT>129<HT
>132<HT>136<HT>139<HT>141<HT>142<HT>142<HT>140<HT>138<HT>135<HT>132<HT>129<
HT>127<HT>125<HT>123<HT>122<HT>120<HT>117<HT>115<HT>112<HT>109<HT>108<HT>10
8<HT>109<HT>110<HT>110<HT>108<HT>104<HT>99<HT>94<HT>89<HT>85<HT>82<HT>80<HT
>79<HT>78<HT>77<HT>75<HT>72<HT>68<HT>63<HT>58<HT>54<HT>50<HT>47<HT>44<HT>43
<HT>41<HT>41<HT>42<HT>42<HT>42<HT>42<HT>39<HT>37<HT>34<HT>30<HT>28<HT>25<H
T>24<HT>23<HT>22<HT>22<HT>21<HT>21<HT>19<HT>17<HT>13<HT>9<HT>5<ETX>F7<EOT>
Receiver acknowledges WBC transmission, and closes the link. It is possible to request for
PLT histogram if the acknowledge contains a “P” instead of <SPACE>, but it is not shown in
this example.
<ACK><SPACE>D
***


yeec维修网视频培训资料购买链接
BeckmanCoulter DXA系列培训资料
Ortho VITROS 系列培训资料
Ortho enGen_ThermoFisher TCA 实验室自动化系统培训资料
Roche Cobas 实验室自动化系统培训资料
Roche Cobas modular系列分析仪培训资料
Horiba-ABX Yumizen系列培训资料
DiaSorin Liaison系列培训资料
Advia2120培训资料
Inpeco-Aptio系列培训资料
Atellica Solution系列培训资料
Siemens Immunoassay系列培训资料 西门子化学发光系列
SIEMENS Advia系列培训资料 西门子生化系列
Toshiba/Abbott系列培训资料 东芝雅培生化系列
Abbott Architect 系列培训资料 雅培生化化学发光系列
ACL TOP 系列培训资料 沃芬TOP血凝系列
BeckmanCoulter Immunoassay系列培训资料 贝克曼化学发光系列
BeckmanCoulter DXH 系列培训资料 贝克曼DXH血球系列
BeckmanCoulter自动样品处理系统介绍性培训资料 贝克曼前后处理流水线系列
BeckmanCoulter AU系列培训资料 贝克曼AU生化系列
BeckmanCoulter DXC系列培训资料 贝克曼DXC生化系列
LaboSpect003/008/AS 7100/7180分析仪培训资料
Horiba-ABX系列培训资料 Horiba-ABX血球系列
Sysmex 血凝系列培训(CA/CS)
Sysmex 尿液分析系列培训(UF1000/5000/UC3500)
Sysmex 血球系列培训(KX21/POCH/XS/XT/XE)
Sysmex XN系列培训(XN-L/XN1000/XN2000/XN3000/XN9000)
Sysmex HISCL系列培训
可直接淘宝店铺购买https://yeec.taobao.com,或咨询手机/微信:13991827712,QQ:67708237
 

您需要登录后才可以回帖 登录 | 立即注册

本版积分规则

QQ|申请友链|手机版|小黑屋|加入QQ群|注销账号|yeec维修网

GMT+8, 2024-11-24 08:08 , Processed in 0.653589 second(s), 33 queries .

Powered by Discuz! X3.5

© 2001-2024 Discuz! Team.

快速回复 返回顶部 返回列表