找回密码
 立即注册
搜索
yeec近年来原创帖合集 本站基础知识下载汇总 yeec网站学习币充值链接 学习中心正式上线

bayer化学发光故障代码

[复制链接]
郑振寰 发表于 2005-1-21 09:14 | 显示全部楼层 |阅读模式

注册登录才能更好的浏览或提问。

您需要 登录 才可以下载或查看,没有账号?立即注册

×
Error Codes index
CMD NO. Error code Subsystem and Command Name Error Name or Message
09 0000 All Subsystems Conversion failed
04 0006 All Subsystems Motor over current circuit tripped
10 0006 Luminomefcer — Advance Luminometer Motor over current circuit tripped
10 0006 Probe - Home Syringe Motor over current circuit tripped
10 0006 Track — Advance Track Motor over current circuit tripped
10 0006 Transport — Home Sample Transport Motor over current circuit tripped
11 0006 Luminonieter -~ Move Waste Probe Motor over current circuit tripped
11 0006 Probe — Aspirate Motor over current circuit tripped
11 0006 Transport — Move Sample Transport Absolute Motor over current circuit tripped
12 0006 Probe — Dispense Motor over current circuit tripped
12 0006 Track - Move Lift Motor over current circuit tripped
13 0006 Transport - Home Reagent Transport Motor over current circuit tripped
13 0006 Transport - Move Sample Transport Barcode Motor over current circuit tripped
14 0006 Transport — Move Reagent Transport Absolute Motor over current circuit tripped
15 0006 Probe - Syringe to Stop Motor over current circuit tripped
15 0006 Transport — Move Reagent Transport Barcode Motor over current circuit tripped
17 0006 Probe — Move Carriage Motor over current circuit tripped
11 000B Fluid — Move Aspirate Probes Move aspirate probe timeout error
12 000B Fluid — Home Water Pump Home water pump timeout error
12 000B Luminometer - Home Base Pump Base pump home timeout error
13 000B Fluid — Aspirate and Dispense Water Pump Aspirate and dispense water pump timeout error
13 000B Luminometer — Home Acid Pump Acid pump home timeout error
14 000B Fluid - Aspirate Water Pump Aspirate water pump timeout error
14 000B Luminometer — Aspirate Reagent 2 (Base) Base pump home timeout error
15 000B Luminometer — Dispense Base with Read Base pump home timeout en-or
17 000B Fluid — Dispense Water Pump Dispense water pump timeout error
17 000B Luminometer — Dispense Base without Read Base pump home timeout error
1D 000B Luminometer — Aspirate Acid Acid pump home timeout error
1E 000B Luminometer — Dispense Acid Acid pump home timeout error
11 000C Fluid — Move Aspirate Probes Motor over-current circuit tripped
12 000C Fluid — Home Water Pump Motor over current circuit tripped
12 000C Luminometer — Home Base Pump Motor over current circuit tripped
13 000C Fluid — Aspirate and Dispense Water Pump Motor over current circuit tripped
13 000C Luminometer — Home Acid Pump Motor over current circuit tripped
14 000C Fluid — Aspirate Water Pump Motor over current circuit tripped
14 000C Luminometer — Aspirate Base Motor over current circuit tripped
15 000C Luminometer — Dispense Base with Read Motor over current circuit tripped
17 000C Fluid — Dispense Water Pump Motor over current circuit tripped
17 000C Luminometer — Dispense Base without. Read Motor over current circuit tripped
1D 000C Luminometer - Aspirate Acid Motor over current circuit tripped
1E 000C Luminometer — Dispense Acid Motor over current circuit tripped
11 000D Aspirate/Wash - Move Aspirate Probes Move aspirate probes map error
12 000D Aspirate/Wash - Home Water Pump Home water pump map error
12 000D Luminometer — Home Base Pump Base pump home map error
13 000D Aspirate/Wash - Aspirate and Dispense Water Pump Aspirate and dispense water pump map error
13 000D Luminometer — Home Acid Pump Acid pump home map error
14 000D Aspirate/Wash — Aspirate Water Pump Aspirate water pump map en-or
14 000D Luminometer - Aspirate Base Aspirate base map error
15 000D Luminometer - Dispense Base with Read Dispense base map error
17 000D Aspirate/wash — Dispense Water Pump Dispense water pump map error
17 000D Luminometer - Dispense Base without Read Dispense base map error
ID 000D Luminometer — Aspirate Acid Aspirate acid map error
1E 000D Luminometer - Dispense Acid Dispense acid map error
11 000E Fluid — Move Aspirate Probes Move aspirate probe EOT error
12 000E Fluid — Home Water Pump Home water pump end-of-travel error
12 000E Luminometer — Home Base Pump Unexpected home sensor transition
13 000E Fluid — Aspirate and-Dispense Water Pump Aspirate and dispense water pump EOT error
13 000E Luminometer - Home Acid Pump Unexpected home sensor transition
14 000E Fluid - Aspirate Water Pump Aspirate water pump end-of-travel error
14 000E Luminometer — Aspirate Base Unexpected home sensor transition
15 000E Luminometer - Dispense Base with Read Unexpected home sensor transition
17 000E Fluid — Dispense Water Pump Dispense water pump end-of-travel error
17 000E Luminometer — Dispense Base without Read Unexpected home sensor transition
1D 000E Luminometer — Aspirate Acid Unexpected home sensor transition
1E 000E Luminometer — Dispense Acid Unexpected home sensor transition
12 000F Fluid — Home Water Pump Home water pump slippage error
12 000F Luminometer — Home Base Pump Excessive home half-step error
13 000F Fluid — Aspirate and Dispense Water Pump Aspirate and dispense water pump slippage error
13 000F Luminometer — Home Acid Pump Excessive home half-step error
14 000F Fluid — Aspirate Water Pump Aspirate water pump slippage error
14 000F Luminometer — Aspirate Base Wrong half-step count
15 000F Luminometer — Dispense Base with Read Wrong half-step count
17 000F Fluid — Dispense Water Pump Dispense water pump slippage error
17 000F Luminometer — Dispense Base without Read Wrong half-step count
ID 000F Luminometer — Aspirate Acid Wrong half-step count
IE 000F Luminometer — Dispense Acid Wrong half-step count
11 0010 Transport — Move Sample Transport Absolute Sample transport not homed
12 0010 Luminometer — Home Base Pump Base pump home back-up error
12 0010 Transport — Move Sample Transport Absolute Sample transport not homed
13 0010 Luminometer — Home Acid Pump Acid pump home back-up error
14 0010 Transport — Move Reagent Transport Absolute Reagent transport not homed
15 0010 Luminometer — Dispense Base with Read Base pump home backup error
15 0010 Transport — Move Reagent Transport Barcode Reagent transport not homed
17 0010 Luminometer - Dispense Base without Read Base pump home back-up error
1E 0010 Luminometer - Dispense Acid Acid pump home back-up error
11 0011 Probe — Aspirate Syringe not homed
12 0011 Luminometer - Home Base Pump Base pump home back-up map error
12 0011 Probe — Dispense Syringe not homed
12 0011 Track - Move Lift Move lift timeout error
13 0011 Luminometer — Home Acid Pump Acid pump home back-up map error
15 0011 Luminometer - Dispense Base with Read Base pump home back-up map error
17 0011 Luminometer — Dispense Base without Read Base pump home back-up map en-or
1E 0011 Luminometer — Dispense Acid Acid pump home back-up map error
11 0012 Probe - Aspirate Probe not homed
12 0012 Probe — Dispense Probe not homed
12 0012 Track - Move Lift Unexpected lift EOT transition error
14 0012 Probe — Move Probe Probe not homed
12 0014 Track - Move Lift Lift move map error
11 0015 Luminometer — Move Waste Probe Waste probe timeout error
12 0015 Track - Move Lift Lift move state error
11 0016 Luminometer — Move Waste Probe Unexpected waste probe end-of-travel transition
11 0017 Transport - Move Sample Transport Absolute Outer sample ring moved unexpectedly
12 0017 Transport — Move Sample Transport Barcode Outer sample ring moved unexpectedly
11 0018 Luminometer — Move Waste Probe Waste probe move map error
11 0018 Transport — Move Sample Transport Absolute Inner sample ring moved unexpectedly
12 0018 Transport - Move Sample Transport Barcode Inner sample ring moved unexpectedly
11 0019 Luminometer — Move Waste Probe Waste probe state error
10 0022 Probe — Home Syringe Syringe home map error
10 0022 Track - Advance Track Track advance setup error
15 0022 Probe — Syringe to Stop Syringe home map error
10 0024 Luminometer — Advance Luminometer Luminometer advance timeout error
10 0024 Probe — Home Syringe Syringe EOT blocked timeout error
10 0024 Track - Advance Track Track advance timeout error
15 0024 Probe — Syringe to Stop Syringe EOT blocked timeout error
10 0025 Luminometer — Advance Luminometer Luminometer advance end-of-travel error
10 0025 Probe — Home Syringe Syringe EOT unblocked error 1
10 0025 Track - Advance Track Track advance end-of-travel error
15 0025 Probe - Syringe to Stop Syringe EOT unblocked error 1
10 0026 Probe — Home Syringe Syringe EOT unblocked error 2
10 0027 Luminometer — Advance Luminometer Luminometer advance map error
10 0027 Probe — Home Syringe Syringe EOT unblocked timeout error
10 0027 Track - Advance Track Track advance map error
10 0028 Probe — Home Syringe Syringe EOT blocked error
10 0029 Track — Advance Track Misloaded cuvette
10 002A Track — Advance Track Pusher home error
10 002B Track - Advance Track Invalid pusher position
10 002C Track - Advance Track Pusher outbound stall error
10 002D Track — Advance Track Pusher inbound stall error
10 002E Track - Advance Track Pusher off home error
10 002F Track - Advance Track Pusher map error
04 002X All Subsystems Timeout error
10 0030 Probe - Aspirate Aspirate setup error
10 0031 Probe - Aspirate Invalid aspirate fluid position
11 0032 Probe - Aspirate Aspirate move error
11 0033 Probe - Aspirate Aspirate fluid map error
11 0034 Probe - Aspirate Aspirate fluid level sense error
11 0040 Probe - Aspirate No syringe selected
12 0040 Probe - Dispense No syringe selected
11 0041 Probe - Aspirate Insufficient syringe volume
12 0041 Probe - Dispense Insufficient syringe volume
11 0042 Probe - Aspirate Insufficient container volume
11 0043 Probe - Aspirate Syringe end-of-travel error
12 0043 Probe - Dispense Syringe end-of-travel error
11 0045 Probe - Aspirate Aspirate map error
12 0045 Probe - Dispense Dispense map error
12 0046 Probe - Dispense Insufficient air gap volume
12 0050 Probe - Dispense Invalid dispense position
12 0051 Probe - Dispense Dispense setup error
12 0052 Probe - Dispense Dispense move error
12 0053 Probe - Dispense Dispensevertical end-of-travel error
12 0054 Probe - Dispense Dispense vertical map error
12 0055 Probe - Dispense Dispense vertical error
12 0056 Probe - Dispense Dispense to zero error
13 0060 Probe - Home Probe Bad vertical sense condition
13 0061 Probe - Home Probe Bad horizontal sense condition
13 0062 Probe - Home Probe No vertical end-of-travel dear
13 0063 Probe - Home Probe Vertical end-of-travel dear error
13 0064 Probe - Home Probe Dry error
13 0065 Probe - Home Probe Dry seek vertical end-of-travel error
13 0066 Probe - Home Probe No vertical EOT blocked
13 0067 Probe - Home Probe Vertical end-of-travel blocked error
13 0068 Probe - Home Probe No horizontal end-of-travel clear
13 0069 Probe - Home Probe No horizontal end-of-travel blocked
13 006A Probe - Home Probe No horizontal position clear
13 006B Probe - Home Probe Horizontal move error
13 006C Probe - Home Probe Excessive horizontal hysteresis
13 006D Probe - Home Probe Horizontal position error
13 006E Probe - Home Probe Horizontal window error
13 0070 Probe - Home Probe Probe homing map error
11 0071 Transport - Move Sample Transport Absolute Move sample transport inner ring map error
12 0071 Transport - Move Sample Transport Absolute Move sample transport inner ring map error
11 0072 Transport - Move Sample Transport Absolute Move sample transport outer ring map error
12 0072 Transport - Move Sample Transport Absolute Move sample transport outer ring map error
13 0072 Probe — Home Probe Horizontal position measure error
11 0073 Transport — Move Sample Transport Absolute Move sample transport inner ring timeout error
12 0073 Transport — Move Sample Transport Barcode Move sample transport inner ring timeout error
13 0073 Probe — Home Probe Horizontal position stall error
11 0074 Transport — Move Sample Transport Absolute Move sample transport inner ring EOT error
12 0074 Transport - Move Sample Transport Barcode Move sample transport inner ring EOT error
13 0074 Probe — Home Probe Horizontal position stall inner ring error
11 0075 Transport — Move Sample Transport Absolute Move sample transport outer ring timeout error
12 0075 Transport — Move Sample Transport Barcode Move sample transport outer ring timeout error
13 0075 Probe — Home Probe Wash station position en-or
11 0076 Transport - Move Sample Transport Absolute Move sample transport outer ring EOT error
12 0076 Transport - Move Sample Transport Barcode Move sample transport outer ring EOT error
13 0076 Probe — Home Probe Outer ring position error
11 0077 Transport — Move Sample Transport Absolute Move sample transport wrong ring moved error
12 0077 Transport - Move Sample Transport Barcode Move sample transport wrong ring moved error
13 0077 Probe — Home Probe Inner ring position error
14 0080 Probe — Move Probe Bad fluid sense condition
14 0081 Probe — Move Probe Horizontal sense error
10 0082 Transport — Home Sample Transport Home sample transport inner ring map error
14 0082 Probe — Move Probe Vertical sense condition error
看贴要回是本分,有问必答是人才,解决问题回贴是公德.
医疗设备维修.维修咨询(请尽可能在论坛提问),协助维修,上门服务.
电话:13991827712

yeec维修网视频培训资料购买链接
BeckmanCoulter DXA系列培训资料
Ortho VITROS 系列培训资料
Ortho enGen_ThermoFisher TCA 实验室自动化系统培训资料
Roche Cobas 实验室自动化系统培训资料
Roche Cobas modular系列分析仪培训资料
Horiba-ABX Yumizen系列培训资料
DiaSorin Liaison系列培训资料
Advia2120培训资料
Inpeco-Aptio系列培训资料
Atellica Solution系列培训资料
Siemens Immunoassay系列培训资料 西门子化学发光系列
SIEMENS Advia系列培训资料 西门子生化系列
Toshiba/Abbott系列培训资料 东芝雅培生化系列
Abbott Architect 系列培训资料 雅培生化化学发光系列
ACL TOP 系列培训资料 沃芬TOP血凝系列
BeckmanCoulter Immunoassay系列培训资料 贝克曼化学发光系列
BeckmanCoulter DXH 系列培训资料 贝克曼DXH血球系列
BeckmanCoulter自动样品处理系统介绍性培训资料 贝克曼前后处理流水线系列
BeckmanCoulter AU系列培训资料 贝克曼AU生化系列
BeckmanCoulter DXC系列培训资料 贝克曼DXC生化系列
LaboSpect003/008/AS 7100/7180分析仪培训资料
Horiba-ABX系列培训资料 Horiba-ABX血球系列
Sysmex 血凝系列培训(CA/CS)
Sysmex 尿液分析系列培训(UF1000/5000/UC3500)
Sysmex 血球系列培训(KX21/POCH/XS/XT/XE)
Sysmex XN系列培训(XN-L/XN1000/XN2000/XN3000/XN9000)
Sysmex HISCL系列培训
可直接淘宝店铺购买https://yeec.taobao.com,或咨询手机/微信:13991827712,QQ:67708237
 

 楼主| 郑振寰 发表于 2005-1-21 09:17 | 显示全部楼层
10 0083 Transport — Home Sample Transport Home sample transport outer ring map error
14 0083 Probe — Move Probe Vertical state error
10 0084 Transport - Home Sample Transport Inner sample transport ring home sensor failure
14 0084 Probe — Move Probe Horizontal state error
10 0085 Transport - Home Sample Transport Outer sample transport ring home sensor failure
14 0085 Probe — Move Probe Fluid seek timeout error
10 0086 Transport — Home Sample Transport Home sample transport inner ring timeout error
10 0086 Transport - Home Sample Transport Home sample transport outer ring timeout error
10 0087 Transport — Home Sample Transport Home sample transport inner ring EOT error
14 0087 Probe — Move Probe Horizontal map error
14 0088 Probe — Move Probe Vertical map error
10 0089 Transport — Home Sample Transport Home sample transport outer ring EOT error
14 0089 Probe — Move Probe Vertical down firmware error
14 008A Probe — Move Probe Post dry error
14 008B Probe — Move Probe Vertical end-of-travel blocked timeout error
14 008C Probe — Move Probe Vertical end-of-travel blocked error
14 0090 Probe — Move Probe Move probe firmware error
1A 0090 Probe — Probe Diagnostics Vertical home failed
13 0092 Transport — Home Reagent Transport Home reagent transport map error
14 0092 Probe — Move Probe Invalid fluid sense range
13 0093 Transport — Home Reagent Transport Home reagent transport timeout error
13 0094 Transport — Home Reagent Transport Home reagent transport end-of-travel en-or
13 0095 Transport — Home Reagent Transport Reagent transport home sensor timeout error
13 0096 Transport — Home Reagent Transport Reagent transport home sensor failure
17 00A1 Probe — Move Carriage Carriage move timeout error
14 00A2 Transport - Move Reagent Transport Absolute Move reagent transport map error
15 00A2 Transport - Move Reagent Transport Barcode Move reagent transport map error
17 00A2 Probe — Move Carriage Carriage move end-of-travel error
14 00A3 Transport - Move Reagent Transport Absolute Move reagent transport timeout error
15 00A3 Transport - Move Reagent Transport Barcode Move reagent transport timeout error
17 00A3 Probe — Move Carriage Carriage move map error
14 00A4 Transport — Move Reagent Transport Absolute Move reagent transport end-of-travel error
15 00A4 Transport — Move Reagent Transport Barcode Move reagent transport end-of-travel error
1F 00A4 Luminometer — Perform Dark Read Reading over-range error
12 00AB Track - Move Lift Invalid lift sensors
14 00B2 Transport - Move Reagent Transport Absolute Mix reagent map error
15 00B2 Transport - Move Reagent Transport Barcode Mix reagent map error
13 00B3 Transport — Home Reagent Transport Reagent transport mix failure
14 00B3 Transport - Move Reagent Transport Absolute Reagent transport mix failure
15 00B3 Transport - Move Reagent Transport Barcode Reagent transport mix failure
10 00BB Aspirate/Wash — Vacuum Control Bad or missing Valve Driver board
15 00BB Aspirate/Wash — Set Valve Outputs Bad or missing Valve Driver board
15 00BB Luminometer - Dispense Base with Read Reading over-range error
1F 00BB Luminometer — Perform Dark Read Firmware read error
20 00BB Luminometer — Perform Light Read Reading over-range error
11 00BD Fluid — Move Aspirate Probes Bad aspirate probe position
14 00BD Fluid — Aspirate Water Pump Bad pump position
17 00BD Fluid — Dispense Water Pump Bad pump position.
15 00CC Luminometer — Dispense Base with Read Missed PMT interrupt error
20 00CC Luminometer — Perform Light Read Missed PMT interrupt error
11 00CD Aspirate/Wash — Move Aspirate Probes Probe not homed
14 00CD Aspirate/Wash — Aspirate Water Pump Pump is not homed
17 00CD Aspirate/Wash — Dispense Water Pump Pump is not homed
15 000D Luminometer — Dispense Base with Read Invalid PMT read interval error
1F 00DA Luminometer — Perform Dark Read Invalid PMT read interval error
20 00DA Luminometer — Perform Light Read Invalid PMT read interval error
15 00DD Luminometer — Dispense Base with Read Firmware read error
20 00DD Luminometer — Perform Light Read Firmware read error
1F 00EE Luminometer — Perform Dark Read Missed PMT interrupt error
00 0101 System event The system ignores the PAUSE key when the system is not running.
00 0102 System event Communication is lost between sequencer and main processor.
00 0104 System event Cuvette did not load into track.
00 0105 System event Cuvette waste bin is full. Empty waste bin when run is completed.
00 0106 System event LED was not found.
00 0107 System event Raw data was not received when expected.
00 0108 System event Received light data when expecting dark data.
00 0109 System event Received dark data when expecting light data.
00 010A System event Raw data was received when not ready.
00 010B System event Probes could not be retracted.
00 010C System event Synchronization is lost between sequencer and main processor.
00 010D System event Synchronization is lost between sequencer and main processor.
00 010F System event Synchronization is lost between sequencer and main processor.
00 0110 System event Cuvette sensor remained blocked after a home track command.
00 0111 System event No response from barcode scanner
00 0112 System event Barcode checksum error
00 0113 System event Home timeout failure
00 0114 System event Home timeout failure
00 0115 System event Home timeout failure
00 0116 System event Home timeout failure
00 0117 System event Home timeout failure
00 0118 System event Home timeout failure
00 0119 System event Home timeout failure
00 011A System event Home timeout failure
00 011B System event Home timeout failure
00 011C System event START of RUN. Tests run = ********
00 011C System event Start from PAUSE. Tests run = ********
00 011C System event Start cleaning procedure: dean cycle. Tests run =********
00 011C System event Start cleaning procedure: rinse cycle. Tests run =********
00 011C System event Start priming procedure. Tests run =********
00 011C System event Start manual operation: home subsystems. Tests run = ********
00 011C System event Start manual operation: clear track. Tests run = ********
00 011C System event Start manual operation: retract probes. Tests run =********
00 011C System event Start manual operation: extend probes. Tests run =********
00 011C System event Start Service Diagnostics. Tests run =********
00 011C System event Start EOM cleaning procedure. Tests run =********
00 011C System event Start EOM purge procedure. Tests run =********
00 011C System event Start EOM prime procedure. Tests run = ********
00 011C System event Start EOM rinse only procedure. Tests run =********
00 011E System event Unable to start due to thermal equilibration failure
00 011F System event Test ********has been canceled for this run.
00 0120 System event Track home timeout due to aspirate/wash probes in unknown position
00 0122 System event Invalid fluid sensed on reagent probe #, position **
00 0127 System event Sample probe cannot retract.
00 0128 System event Vacuum cannot reach required pressure.
00 0129 System event Warm-up failed due to track equilibration error
00 012A System event Warm-up failed due to track equilibration error
00 012B System event Warm-up failed due to track equilibration error
00 012C System event Warm-up failed due to reagent heater equilibration error
00 012D System event Warm-up failed due to a reagent probe equilibration error
00 012E System event Warm-up failed due to a reagent probe equilibration error
00 012F System event Warm-up failed due to a reagent probe equilibration error
00 0130 System event Unable to determine error type
00 0131 System event Unable to determine error type
00 0132 System event Unable to determine error type
00 0133 System event Unable to determine error type
00 0134 System event Unable to determine error type
00 0135 System event Unable to determine error type
00 0136 System event Unable to determine error type
看贴要回是本分,有问必答是人才,解决问题回贴是公德.
医疗设备维修.维修咨询(请尽可能在论坛提问),协助维修,上门服务.
电话:13991827712
 楼主| 郑振寰 发表于 2005-1-21 09:17 | 显示全部楼层
000137System eventUnable to determine error type
000138System eventUnable to determine error type
000139System eventUnable to determine error type
00013ASystem eventUnable to determine error type
00013BSystem eventUnable to determine error type
00013CSystem eventUnable to determine error type
00013DSystem eventUnable to determine error type
00013ESystem eventUnable to determine error type
00013FSystem eventUnable to determine error type
000140System eventUnable to determine error type
000141System eventUnable to determine error type
000142System eventValve error on sample probe subsystem
000143System eventValve error on reagent probe # subsystem
000144System eventValve error on aspirate/wash subsystem
000145System eventValve error on track subsystem
000146System eventBarcode decoder communication error
000147System eventBarcode decoder communication error
000148System eventBarcode decoder communication error
000149System eventBarcode decoder communication error
00014ASystem eventBarcode decoder communication error
00014BSystem eventBarcode decoder communication error
00014CSystem eventBarcode decoder communication error
00014DSystem eventCycle slip error
00014ESystem eventCycle slip error
00014FSystem eventCycle slip error
000150System eventCycle slip error
000151System eventCycle slip error
000152System eventCycle slip error
000153System eventSSP and sequencer substate error
000154System eventSSP and sequencer substate error
000155System eventSSP and sequencer substate error
000156System eventSSP and sequencer substate error
000157System eventSSP and sequencer substate error
000158System eventSSP and sequencer substate error
000159System eventThe run stopped because the sample probe cover was removed.
00015ASystem eventThe run stopped because the reagent probes cover was removed.
00015BSystem eventThe run stopped because the front cover was lifted.
00015CSystem eventAspirate error
00015DSystem eventDark count beyond limit: region *, cup ***,******
00015ESystem eventDark count beyond limit: region *
00015FSystem eventRerun calibrator cup ***, test******. Trailing samples are rescheduled.
000160System eventCalibration failed for test. ****** Samples in track will be canceled.
000161System eventRaw data received was corrupt.
000162System eventDue to excessive errors, the sample probe has been taken off-line.
000163System eventDue to excessive errors, the sample transport has been taken off-line.,
000164System eventDue to excessive errors, reagent probe # has been taken off-line.
000165System eventDue to excessive errors, the reagent transport has been taken off-line.
000166System eventDue to excessive errors, cuvettes cannot be dispensed.
000167System eventReagent vial •••""""fluid level is too low.
000168System eventReagent vial "•"•"•"fluid level is too high.
000169System eventVacuum remained too low to continue priming.
00016ASystem eventThe system cannot process communication requests. The queue is full.
00016BSystem eventDark count beyond limit: region *, cup ***, ******
00016CSystem eventCheck luminometer performance using System Performance Tests.
00016DSystem eventPoor light to pre-light ratio, cup ***, test ****
00016ESystem eventNoise in light signal, cup ***, test ****
00016FSystem eventNo light signal detected, cup ***, test ****
000170System eventReagent transport timeout error
000171System eventThe results database is full. Delete results before next run.
000172System eventExtended Operation Module pump command error
000173System eventExtended. Operation Module pump command en-or
000174System eventExtended Operation Module valve command error
000175System eventExtended Operation Module valve command error
000176System eventExtended Operation Module error: ACS water bottle not filling.
000177System eventExtended Operation Module error: ACS waste bottle not emptying.
000178System eventExtended Operation Module error: ACS water bottle too full.
000179System eventThe Extended Operation Module is not installed or is inactive.
000180System eventManual operations supply errors
000181System eventClot detected while aspirating fluid
000183System eventClot detected while dispensing fluid
000184System eventClot alert component error
000185System eventClot alert threshold reached
000186System eventClot cleared
000187System eventClot alert circuit error cleared
000188System eventSample cup *** cancelled by clot alert
000201SRQ - GeneralExternal RAM checksum error detected
000202SRQ - GeneralStack overflow detected
000203SRQ - GeneralMotor over current error, MC board *
000204SRQ - GeneralA/D converter error
000205SRQ - GeneralOver temperature detected on MC board *
000210SRQ - ProbeUninitiated move detected for sample or reagent probe
000211SRQ - ProbeUninitiated move detected for syringe
000212SRQ - ProbeUnrecoverable uninitiated move for probe carriage
000213SRQ - ProbeSample probe guard removed
000214SRQ - ProbeSample probe guard installed
000215SRQ - ProbeReagent probe guard removed
000216SRQ - ProbeReagent probe guard installed
000217SRQ - ProbeBackplane fuse is open
000221SRQ - TrackTrack thermal system is out of tolerance
000223SRQ - TrackTrack thermal system is out of tolerance
000225SRQ - TrackTrack thermal system is out of tolerance
000227SRQ - TrackTrack thermal system is out of tolerance
00022BSRQ - TrackReagent probe I thermal system is out of tolerance
00022DSRQ - TrackReagent probe 2 thermal system is out of tolerance
00022FSRQ - TrackReagent probe 3 thermal system is out of tolerance
000230SRQ - TrackTrack thermal system is out of tolerance
000231SRQ - TrackUnrecoverable uninitiated move for track
000232SRQ - TrackUnrecoverable uninitiated move for elevator
000233SRQ - TrackUnrecoverable uninitiated move for cuvette pusher
000234SRQ - TrackCuvette pusher error
000235SRQ - TrackCuvette feeder error
000236SRQ - TrackTrack advance too short
000237SRQ - TrackTrack advance too long
000238SRQ - TrackTrack advance to load too short
000239SRQ - TrackTrack advance to load too long
000240SRQ - LuminometerUnrecoverable uninitiated move for luminometer
000241SRQ - LuminometerUnrecoverable uninitiated move for waste probe
000242SRQ - LuminometerUnrecoverable uninitiated move for acid pump
000243SRQ - LuminometerUnrecoverable uninitiated move for base pump
000246SRQ - LuminometerReagents 1 and 2 door lifted
000247SRQ - LuminometerReagents 1 and 2 door open
000248SRQ - LuminometerReagent 1 is low
000249SRQ - LuminometerReagent 2 is low
00024ASRQ - LuminometerAcid level high
00024BSRQ - LuminometerBase level high
000260SRQ - TransportsUninitiated move detected for sample transport
000261SRQ - TransportsUninitiated move detected for sample transport
000262SRQ - TransportsUninitiated move detected for reagent transport
000263SRQ - TransportsSample cover was removed during run. Samples will not be dispensed.
000265SRQ - TransportsReagent cover was removed during run. Reagents will not be dispensed.
000280SRQ - FluidsUnrecoverable uninitiated move for water pump
000281SRQ - FluidsUnrecoverable uninitiated move for aspirate/wash probe ACS water bottle lid opened
000283SRQ - FluidsACS water bottle fluid level low
000284SRQ - FluidsACS water bottle fluid level high
000285SRQ - FluidsVacuum is lost. Vacuum sensor error, or Vacuum remained too low to continue priming
000286SRQ - FluidsVacuum sensor error
000287SRQ - FluidsVacuum level high
00028ASRQ - FluidsVacuum sensor error
00028CSRQ - FluidsExtended Operation Module error: ACS water bottle too full.
00028DSRQ - FluidsThe EOM waste bottle is full. Please empty.
00028FSRQ - FluidsThe Extended Operation Module is not active because the EOM button is in the off position.
000290SRQ - FluidsThe Extended Operation Module is active.
000292SRQ - FluidsThe EOM is not active because the floor stand lid has been opened.
0002FFSRQ - GeneralSRQ conditions lost due to queue overflow
0004XXSequencer Timeout ErrorRun canceled due to sequencer timeout
看贴要回是本分,有问必答是人才,解决问题回贴是公德.
医疗设备维修.维修咨询(请尽可能在论坛提问),协助维修,上门服务.
电话:13991827712
maxyan 发表于 2007-3-19 11:01 | 显示全部楼层

thanks!

白梅 发表于 2007-11-20 15:30 | 显示全部楼层

有AXSYM的错误代码吗?

谢谢啦

顶风 发表于 2007-11-20 16:23 | 显示全部楼层
没有这方面的帖子,但本站有Axsym的维修手册,那上面有机器的故障代码表。积分足够即可下载。如果是医院的用户就没有必要看了,反正都是代理公司免费维修的。
您需要登录后才可以回帖 登录 | 立即注册

本版积分规则

QQ|申请友链|手机版|小黑屋|加入QQ群|yeec维修网

GMT+8, 2025-4-2 19:58 , Processed in 5.645775 second(s), 39 queries .

Powered by Discuz! X3.5

© 2001-2025 Discuz! Team.

快速回复 返回顶部 返回列表