Summary Synchron system configuration can be changed by either loading software or via QNX. The QNX procedure is described below.
Procedure The function set_dinstr provides the same function as set_instr for QNX version 2.0. This function allows the field engineer to change system type (CX7, CX5, CX4), language, and chemistry configuration without loading software. Unlike the CX function, all files required by this function are located on the hard drive.
LIMITATIONS This procedure should only be used during installation.
Reason: This procedure sets the database to the original default database (Software Set: v6.0.10.77 or Synchron Console: v6.0.00.03). However, if a chemistry update has been performed, the software is not notified the original database is in use. The operator will have incorrect chemistry database information displayed on the System Parameters screen. This limitation currently exists with the set_instr function used on the Synchron CX product line.
Step Action
1 Break the boot sequence to access QNX.
2 Change to the /cx7/bin directory by typing:
cd /cx7/bin <Enter>
3 Start the configuration change by typing:
set_dinstr <Enter>
4 The following information appears on the screen:
Continuing with the program will clear metering information
Do you still wish to continue? (y/n)
If y was selected and metering was not initialized, the following messages will appear on the screen:
/cx7/bin/meter*.dat: No such file or directory Message appears if metering wasn't initialized.
Please wait. Getting original instrument settings.
Starting setup task.........
Metering has been disabled
CURRENT SELECTIONS:
Information displayed at this time will depend on current console configuration. Settings below represents CX PRO default console configuration.
Instrument Type: CX7 PRO
CO2: Enabled
Language: English
CX3 Configuration: BUN3, GLU3, CRE3, NA, K, CL, CO2, CA(ISE), TP3
WARNING: CHANGING INSTRUMENT TYPES WILL DELETE ALL DATABASES EXCEPT FOR THE CHEMISTRY AND CALIBRATOR.
SET INSTRUMENT TYPE
CONVERT THE CX5 ALIGNMENT DATA TO CX5 PRO ALIGNMENT (ONLY USED WHEN UPGRADING CX5PTS TO CX5 PRO)
CHANGE LANGUAGE
EDIT THE INSTRUMENT SERIAL NUMBER
EXIT
Please enter number (1 - 5)
5 Click the links below to see what will appear when each of the 3 set_dinstr options are selected:
1) SET INSTRUMENT TYPE ==>
2) CONVERT THE CX5 ALIGNMENT DATA TO CX5 PRO ALIGNMENT (ONLY USED WHEN UPGRADING CX5PTS TO CX5 PRO)==>
3) CHANGE LANGUAGE ==>
4) EDIT THE INSTRUMENT SERIAL NUMBER ==>
6 When configuration has been changed, exit set_dinstr by entering 5 Exit.
If the serial number was changed, the following message will appear on the screen.
Re-starting setup task ...
Updating the serial number
Please wait.....
The following message always appears when exiting this function:
If operation has completed successfully then please reboot the instrument.
If the operation failed, please see the manual for further instructions.
7 To prevent a Fatal Note regarding the metering data from occurring during boot, type the following command:
rm setup_data3 <Enter>
8 If the original CE/Delta console had Software Set 6.2 or higher installed, return to the console setup procedure to copy alignment data from the backup diskette to the hard drive via QNX.
If the original CE/Delta console had Software Set 5.X installed, type exit and press the <Enter> key at the QNX prompt and the boot process will pick up at the blue copyright screen.
Option 1 - Set Instrument Type When SET INSTRUMENT TYPE is selected, the following information will appear on the screen:
Click here to return to instructions ==>
CODE Instrument
0 CX7 PRO ==> CLICK HERE TO LINK TO OPTIONS ()
1 CX4 PRO ==> CLICK HERE TO LINK TO OPTIONS ()
2 CX5 PRO ==> CLICK HERE TO LINK TO OPTIONS ()
3 CX3 PRO ==> CLICK HERE TO LINK TO OPTIONS ()
4 Exit
Please enter instrument code (0-4)
If the CX7 PRO is selected, the following information will appear on the screen:
Click here to return to top
The instrument has been set to CX7 PRO
WARNING: CHEMISTRY CONFIGURATIONS WITHOUT CO2 REQUIRE THE CO2 PATCHOUT KIT - CPU DIP SWITCH (S1-7) SETTING IS IGNORED.
WARNING: CHANGING CX7 CONFIGURATION DELETES ALL DATABASES.
CX7 PRO Configuration
0 No Change
1 BUN, GLU, CRE, CA(CUP), NA, K, CL
2 BUN, GLU, CRE, CA(CUP), NA, K, CL, CO2
3. BUN, GLU, CRE, NA, K, CL, Ca (ISE), TP
4. BUN, GLU, CRE, NA, K, CL, Ca (ISE), CO2, TP
Please enter configuration code (0-4)
After the desired configuration is selected, appropriate files will be copied and the operator will be returned to the screen described in step 4 above. The new instrument configuration will be reflected under CURRENT SELECTIONS
CURRENT SELECTIONS:
Instrument Type: CX7 PRO
CO2: Enabled
Language: English
CX3 Configuration: Information here will be based on selection made above.
If the CX4 is selected, appropriate files will be copied and the operator will be returned to the screen described in step 4 above. The new instrument configuration will be reflected under CURRENT SELECTIONS.
Click here to return to top
CURRENT SELECTIONS:
Instrument Type: CX4 PRO
CO2: Disabled
Language: English
This is a CX4 instrument configuration.
If the CX5 PRO is selected, the following information will appear on the screen:
Click here to return to top
The instrument has been set to CX5 PRO
WARNING: CHEMISTRY CONFIGURATIONS WITHOUT CO2 REQUIRE THE CO2 PATCHOUT KIT - CPU DIP SWITCH (S1-7) SETTING IS IGNORED.
WARNING: CHANGING CX5 CE CONFIGURATION DELETES ALL DATABASES.
CX5 PRO Configuration
0 No Change
1 NA, K, CL
2 NA, K, CL, CO2
3. NA, K, CL, Ca (ISE)
4. NA, K, CL, CO2, Ca (ISE)
Please enter configuration code (0-4)
After the desired configuration is selected, appropriate files will be copied and the operator will be returned to the screen described in step 4 above. The new instrument configuration will be reflected under CURRENT SELECTIONS
CURRENT SELECTIONS:
Instrument Type: CX5 PRO
CO2: Disabled/Enabled
Language: English
The ISE Configuration: Information here will be based on selection made above.
If the CX3 PRO is selected, the following information will appear on the screen:
Click here to return to top
The instrument has been set to CX3 PRO
WARNING: CHEMISTRY CONFIGURATIONS WITHOUT CO2 REQUIRE THE CO2 PATCHOUT KIT - CPU DIP SWITCH (S1-7) SETTING IS IGNORED.
WARNING: CHANGING CX3 CONFIGURATION DELETES ALL DATABASES.
CX3 PRO Configuration
0 No Change
1 BUN, GLU, CREA, CA (CUP), NA, K, CL
2 BUN, GLU, CREA, CA (CUP), NA, K, CL, CO2
3. BUN, GLU, CREA, NA, K, CL, Ca (ISE), TP
4. BUN, GLU, CREA, NA, K, CL, CO2, Ca (ISE), TP
Please enter configuration code (0-4)
After the desired configuration is selected, appropriate files will be copied and the operator will be returned to the screen described in step 4 above. The new instrument configuration will be reflected under CURRENT SELECTIONS
CURRENT SELECTIONS:
Instrument Type: CX3 PRO
CO2: Disabled/Enabled
Language: English
The CX3 Configuration: Information here will be based on selection made above.
Option 2 - CX5 Alignment Data This option is not implemented.
Click here to return to instructions ==>
Option 3 - Change Language The following information will appear on the screen when the Change Language option is selected:
Click here to return to instructions ==>
3) Change Language
0) English
1) French
2) German
3) Italian
4) Spanish
5) Exit
Please enter number (0-5)
After the desired language is selected, appropriate files will be copied and the operator will be returned to the screen described in step 4 above. The new language will be reflected under CURRENT SELECTIONS.