人家报架移动错误,你死盯着爪子做什么?
手册上明明告诉你,一个架子测试完成,接着做下一个架子的时候,左右试管探测器没有检测到,所以停止了。
Rack Move Error 2 Description: A rack is moved even the rack shift operation is not performed during the analysis. Purpose: To ensure the rack is properly positioned before or after the shift operation. Method: If Tube Position sensor (Right) or Tube Position sensor (Left) is shaded during from completing rack shift to starting next rack shift. System monitors at a constant timing (every 0.1 second) except from starting to move the hand until the hand movement completes. Error Type: C Help [OK]: The error is recovered.
任何一本手册都只会告诉你错误发生机理,至于怎么修那是维修人员的全面理解的事情,手册不可能面面俱到,也不可能把工程师当白痴。 |