在G-service里面校准下冲洗分离管路的检测器。步骤如下:
- Exit the ADVIA Centaur CP system software to desktop.
- Turn the ADVIA Centaur CP instrument power off.
Wait 10 seconds.
- Turn the ADVIA Centaur CP instrument power on.
- Access the Service Software:
- At the desktop, double-select ADVIA Centaur CP Tools.
- Double-select gService.
- At the Service Login window, enter the FSE user name and password.
The ADVIA Centaur CP Service Software window displays.
- At the ADVIA Centaur CP Service Software window, select Wash Station.
- At the Wash Station Main Driver window, select Home Wash Station.
- Select Exercise Wash Station.
- At the Exercise Wash Station window, in the Pumps area, verify that the Aspiration Pump On field is set to 1500 ms.
- Select Aspiration Pump On and wait a few seconds.
- Repeat these steps three more times.
- Select Exit.
- At the Wash Station Main Driver window, select Calibrate Sensors.
- At the Calibrate Sensors window, select Calibrate Aspirate Probes Bubble Detectors.
- When prompted, ensure that the waste tubing is clear of fluids, and then select Yes.
- Close all Service Software windows.
不过总报试剂针离线,怎么会不影响工作呢?仪器运行时报试剂针离线机器会停止的! |