FM AD function. (CPU F). ADC INTR 001 = A/D timer load counter error call from interrupt state. ADC INTR 002 = A/D timer mode load hold register test error call from interrupt state. ADC 4.01 0xx = More then 20% over range on channel xx Channel xx Explanation 0 MAIN PC BOARD TEMPERATURE 1 MOTOR CURRENT 2 VALVE CURRENT 3 CONDUCTIVITY CONTROL ACETATE/TOTAL 4 DEGASSING PRESSURE 5 HEATING VESSEL TEMPERATURE 6 CONDUCTIVITY CONTROL BCM OPTION 7 TEMPERATURE BCM CONDUCTIVITY TRANSDUCER 8 UF TRANSDUCER COIL CURRENT (POSITIVE SYNC) 9 HIGH PRESSURE GUARD (HPG) 10 DIALYSIS PRESSURE (PD) 11 AUX INPUT 12 AUX INPUT 13 AUX INPUT 14 OFFSET VOLTAGE AD-SYSTEM 15 REFERENCE VOLTAGE AD-SYSTEM FCA 4.01 001 = A/D timer load counter error call from ad-task. FCA 4.01 002 = A/D timer mode load hold register test error call from ad-task. FCB 4.01 xxx = AD conversion stop for xxx sec. FCN 4.01 000 = Case error. The program is not working properly. FCN 4.01 001 = Timer error. FCN 4.01 002 = AD converter offset out of limits. -10 mV < offset (= 0.0 mV) < + 10 mV FCN 4.01 003 = AD converter reference out of limits. 45.45 mV < reference ( = 49.50 mV) < 54.89 mV FCN 4.01 004 = AD-system timeout 1 sec. FCN 4.01 005 = AD buffer full. FCN 4.01 006 = AD buffer full and empty.
FM CPU板A/D转换的问题,检查下板上的接头,看下是否有液体溅到板上面。不行的话还是得找厂家工程师支持。 |