找回密码
 立即注册
搜索
yeec近年来原创帖合集 本站基础知识下载汇总 yeec网站学习币充值链接 学习中心正式上线

[分享] 自制X线机探测装置

[复制链接]
郑振寰 发表于 2016-3-23 11:02 | 显示全部楼层 |阅读模式
Recently I received an email from Keith Feay in which he introduced his simple made X-ray detector. Here his building instruction:



Do-it-yourself


A simply made
X-Ray Detector
made from readily available materials


created by Keith Feay


Following a trip I made to Africa in 2008 it became obvious that many hospitals had no means of measuring X-ray equipment output. As most X-ray engineers and radiographers use fluorescent screens to check if there is an output from the tube, I wondered if this could be extended to produce some measurable results.
The detector described will work best with an oscilloscope or PC soundcard oscilloscope software but it also works with a DVM or analogue voltmeter. As I live in the UK all the parts used have been found or acquired locally. I am sure that the parts required will be available in most parts of the world.

X-ray equipment can be dangerous if not used correctly. Please use radiation protection at all times. This article is intended for those technicians and engineers who have had some training in the use and safety of x-ray machines.


Parts required
• Fluorescent screen from an X-ray cassette.
On my trip to Africa I found several damaged cassettes in all x-ray departments. Most
were beyond repair but had good fluorescent screens.
• Small Photo Voltaic cell.
A good source of these is the cheap garden lights sold in UK £1 shops.
Any small cell (up to 5cm square will do) its voltage output in daylight should be around
2.5 volts but this is not critical.
• A box to put the whole assembly into.
Plastic medicine boxes are useful for this any are usually available in most hospitals.
• Twin cable and suitable plugs for the oscilloscope / meter.
Screened cable is best but if not available the twist 2 thin wires together.


Assembly
Test your photocell in daylight to ensure that you get an output. Mark the positive and negative poles.
Test your fluorescent screen to make sure that it glows when exposed to x-rays. Most glow blue or green and the brightness depends on their chemical composition.
The screen is placed fluorescent side against the light sensitive side of the photocell and held in place with electrical tape. The cloth Gaffer tape or Tesa tape is best but any electrical tape will do as its purpose is to keep out light.
Measure the output again in strong light the reading should be zero if all light has been successfully excluded.
Place the whole assembly in a suitable box with the lead exiting through a hole drilled in the box.
[tr]Connect the detector to a meter or scope.
Place the detector under the X-ray tube. At about 500mm from the focal spot.
The distance is not critical and may need to be closer with a low output unit.
Switch the meter / scope to the suitable voltage setting. (Try 200mV)
Set the x-ray machine to 70kV 50mA and 1 second or whatever you can on your machine.
Make an Exposure whilst watching the meter / scope.
You should get a reading of approximately 100 to 200 mV depending on the screen and the cell used.
This is not a precision instrument but will when used with an oscilloscope give you a reasonably indication of the output from an x-ray machine.

Complete solar light on the left. Cells on the right.

Completed test box on left used to check a portable x-ray unit.

本帖子中包含更多资源

您需要 登录 才可以下载或查看,没有账号?立即注册

×
看贴要回是本分,有问必答是人才,解决问题回贴是公德.
医疗设备维修.维修咨询(请尽可能在论坛提问),协助维修,上门服务.
电话:13991827712

yeec维修网视频培训资料购买链接:
BeckmanCoulter DXA系列培训资料
Ortho VITROS 系列培训资料
Ortho enGen_ThermoFisher TCA 实验室自动化系统培训资料
Roche Cobas 实验室自动化系统培训资料
Roche Cobas modular系列分析仪培训资料
Horiba-ABX Yumizen系列培训资料
DiaSorin Liaison系列培训资料
Advia2120培训资料
Inpeco-Aptio系列培训资料
Atellica Solution系列培训资料
Siemens Immunoassay系列培训资料 西门子化学发光系列
SIEMENS Advia系列培训资料 西门子生化系列
Toshiba/Abbott系列培训资料 东芝雅培生化系列
Abbott Architect 系列培训资料 雅培生化化学发光系列
ACL TOP 系列培训资料 沃芬TOP血凝系列
BeckmanCoulter Immunoassay系列培训资料 贝克曼化学发光系列
BeckmanCoulter DXH 系列培训资料 贝克曼DXH血球系列
BeckmanCoulter自动样品处理系统介绍性培训资料 贝克曼前后处理流水线系列
BeckmanCoulter AU系列培训资料 贝克曼AU生化系列
BeckmanCoulter DXC系列培训资料 贝克曼DXC生化系列
LaboSpect003/008/AS 7100/7180分析仪培训资料
Horiba-ABX系列培训资料 Horiba-ABX血球系列
Sysmex 血凝系列培训(CA/CS)
Sysmex 尿液分析系列培训(UF1000/5000/UC3500)
Sysmex 血球系列培训(KX21/POCH/XS/XT/XE)
Sysmex XN系列培训(XN-L/XN1000/XN2000/XN3000/XN9000)
Sysmex HISCL系列培训
可直接淘宝店铺购买https://yeec.taobao.com,或咨询手机/微信:13991827712,QQ:67708237
 

 楼主| 郑振寰 发表于 2016-3-23 11:14 | 显示全部楼层
这是一位英国工程师在非洲援外的时候,利用废旧X线机的感光屏(荧光屏)、光电池和示波器(接笔记本电脑上的声卡接口上的示波器也行,万用表也行)制作的移动X线机球管探测装置。
感光屏后面接光电池,引线最好采用屏蔽线,没有条件的,双引线不要接触。
由于移动X线机根据曝光条件的不同,X线的光斑大小也不同,照射距离也不同,所以检查的时候提前做好条件设置。
曝光条件作者选择的是75KV 50MAS,示波器的档位选择在200mv。根据你的光电池选择不同,输出会在100-200mv之间波动,说明球管正常,否则意味着有问题。
这个制作的目的是检查球管是否出现,功率是否正常,但又没有正规的检查装置的情况下,这样做可以保护操作/维修人员。
看贴要回是本分,有问必答是人才,解决问题回贴是公德.
医疗设备维修.维修咨询(请尽可能在论坛提问),协助维修,上门服务.
电话:13991827712
gb5808 发表于 2016-5-8 18:22 | 显示全部楼层
自制X线机探测装置
您需要登录后才可以回帖 登录 | 立即注册

本版积分规则

QQ|申请友链|手机版|小黑屋|加入QQ群|yeec维修网

GMT+8, 2026-11-13 21:49 , Processed in 0.318431 second(s), 35 queries .

Powered by Discuz! X3.5

© 2001-2026 Discuz! Team.

快速回复 返回顶部 返回列表