找回密码
 立即注册
搜索
yeec近年来原创帖合集 本站基础知识下载汇总 yeec网站学习币充值链接 学习中心正式上线

[分享] 自制显微镜灯泡改换LED

[复制链接]
郑振寰 发表于 2016-3-23 13:23 | 显示全部楼层 |阅读模式
Do-it-yourself

Microscope Conversion
to LED-Light

created by Frank Weithöner


This instruction shows how to convert a standard microscope with light bulb to LED light. Only a cheap mobile phone charge can be used as the power supply because of the much lower power consumption of the LED light. Defective original power supplies do not have to be repaired or exchanged. Because a LED does not break, light bulbs are no longer needed.
The whole LED system costs less than 20 €.



The idea behind
In our workshop we often encounter problems with broken power supplies of microscopes. The power supplies are usually switch mode power supplies that can not be repaired easily. To supply the light bulb with an external standard power supply is an alternative but such power supplies which deliver 2 A (at 12 V) or even 4 A (at 6 V) are either not available or too expensive, and a dim function is still missing...
In other cases the bulbs were blown and these often very special bulbs are not available.
In the past these microscopes could not be repaired. But now we convert these microscopes to LED illumination. That works perfectly.
White ultra-bright LEDs, especially with a lens, are as bright as common 20 W-bulbs, but they only need a fraction of the current. The current needed is so small that common mobile phone chargers can be used as a power supply. Mobile phone chargers are very cheap and commonly available.



Advantages
The conversion to LED light brings some interesting advantages:

•  The light is very bright.
     A 1 W-LED in combination with a lens is as bright as a conventional 20W-light bulb.

•  The light is real white.
     Light bulbs produce a yellowish light which has to corrected by a blue filter. LED light is
     real white and also does not change the colour when dimming.

•  LEDs last for ever. No spare bulbs are needed any more.
     The life expectancy of LEDs is about some ten thousand hours. They are also insensitive
     against shocks and vibrations.

•  The power consumption is much lower.
     The power consumption is so low that cheap power supplies can be used.

•  Microscope can run with batteries for hours.
     A small (rechargeable) battery pack can provide power for some working hours.
     Sometimes the battery pack can even be integrated in the microscope.

•  Microscope and LED gets less warm.
     Less power consumption also means less heat. The microscope does not get warm any
     more.
•  The conversion is cheaper than a repair of a power supply.
     The whole system with power supply can be build for less than 20 €.



Disadvantages
I admit a white high-power-LED is not (yet) available everywhere. But because they are universal in the usage and only one model fits for all microscope types (and for other purposes) it make sense to order some pieces from Europe or US. In Europe a 1 W-LED with integrated lens costs about 10 €.



Considerations
After testing different LED types in different microscopes I came to the following conclusion:

Ultra-bright LEDs are available in standard 5 mm housings (No.1). Six to twelve LEDs produce a light which is bright enough (No.2). But the illumination with just six LEDs could be more homogeneous. The beam angle is not perfect but acceptable. One LED costs
0.20 - 1 €.

Beside the standard LEDs there are Luxeon high power LEDs (No.3).
1 W, 3 W and even 5 W-types exist. The LED is mounted on a small PCB with an aluminium board. The aluminium acts as a heat sink. This make mounting and cooling easy. The beam angle is 110°. An additional optic has to be used. Such LEDs cost 5 - 10 €.

A LED (or LED combination) of one 1 W is enough. The beam angle should be no more than 15°. 10° are perfect.

For me the best solution is the Luxeon Star/O (No.4). It is a 1 W-LED with an intergrated lens of 10°. The price is about 10 €.


1   super bright LED in standard 5 mm
     housing

2   first try with 6 x 5 mm LEDs


3   Luxeon 1 W high power LED


4   Luxeon Star/O with optic


The electronic circuit
The LEDs can not connected directly to an external power supply. An electronic control is needed.
LEDs are controlled by the current and not by the voltage. The current has to be stable and the voltage will be adjusted automatic. This current stabilizer must run with any common mobile phone chargers. That means a wide-range voltage input of 5 - 15 V is demanded. The mobile phone charger has to deliver 400 mA which is always the case.
The specifications of a Luxeon LED are: 350 mA at 3.42 V. The needed electronics has to stabilize the output current to 350 mA whatever the input voltage is. The circuit is quite simple but tricky.


The circuit diagram.

T1 and T2 make the current stabilizer. The voltage divider around the pot reduces the bias for T2 in order to dim the LED down.
Because the LED current is stabilized the supply voltage is uncritical. Every mobile phone charger can be used.


Transistor T2 controls the LED. Together with R3 it is in series with the LED. That means that the LED current also flows through the transistor and the resistor. Parallel to the resistor is the BE path of the transistor T1. Because the BE-voltage drop of a transistor is always 0.7 V also the voltage across the resistor is fixed to 0.7 V. But when the voltage across the resistor is fixed and the resistor anyway then also the current through the resistor must be fixed. The current is stabilized.

The value of the current depends on resistor R3.
R3 = 0.7 V / 0.35 A             R3 = 2 Ω
After testing in practice I chose 1.8 Ω.

The power loss of the resistor is almost 1/4 Watt.
P = 0.7 V x 0.35 A              P = 0.245 W
In theory a 1/4 Watt resistor is OK but it will get hot. With a 1/2 Watt resistor we are on the safe side.

The bias voltage for T2 is created by the voltage divider R1, the pot and R2. The values are uncritical because the current is already limited. R2 is responsible for the lowest brightness and ensures that the pot has a nice variation range.

T1 is a universal low power transistor of type BC 546 or similar.

T2 must be a bigger one. The size of T2 depends on the voltage drop across CE. It is the difference between the power supply voltage and the LED voltage (3.42 V) and the 0.7 V voltage drop across R3. T2 has to be bigger, the higher the power supply voltage is.
V(CE) = V(in) - V(LED) - V(R3)
The transistor voltage drop V(CE) multiplied by the LED current makes the transistor power loss. Current and power loss have to be considered when choosing the transistor type.
A BD 139 for example is big enough. I took the even bigger BD 243 just because it was lying around.

A small heat sink is never wrong. In this case mounting T2 directly to the housing of the microscope is even a better idea. The heat can be conducted away over the housing and the other devices can be mounted directly to the transistor. No PC Board is needed.


When using a heat sink or the metal housing of the microscope be sure that the transistor is mounted insulated by using a glimmer insulator. Usually the metal part of the transistor is connected to the collector. But in our circuit the collector is not connected to the ground.
Make sure that also the screw is insulated from the transistor.


Mechanical preparations
The original lens with the lamp housing is still needed. The LED will be mounted directly on the bottom plate. Make sure that the LED is placed exactly there, where the filament of the light bulb was before.
The blue filter disk is not needed any more. It can be taken out of the lamp housing.


Base of the microscope. The stand with the mechanic and the optics is removed.

The lens housing with the lens is seen in the centre.
At the right is the new pot for dimming and below the lens housing a small toggle switch for on/off can be seen.
A combined pot/switch would be nicer but was not available.

View to the base from the inside.

Pot and switch are already mounted but not wired. For the transistor a nice place was found. An existing hole with a threat could be used. The paint from the surface of the microscope housing was removed before mounting. The transistor now has a smooth and plane thermal connection.

The microscope gets an power supply socket at the back (up).

The wiring is finished. Only the LED has to be mounted on the bottom plate.
The parts are mounted directly at the transistor and at the pot. A PCB is not needed.

A detailed view to the wiring.

Make sure that T2 is insulated from the metal housing. Also use thermal paste for mounting.

When the mechanical work is finished the wiring takes only 15 minutes.



Experiences
All users where very satisfied with the conversion. I never heard a complain about the light. It seems that a 1 W-LED with a focus lens is really enough for all examinations. Even with the 100x oil-objective which is used for malaria examination the brightness is not only high enough but it is even brighter than the original illumination.



Update 2015


The 700 mA version
Five years have passed since I wrote the above article. In the meantime the LEDs got cheaper and more powerful. So I decided to upgrade the microscope illumination with a more powerful LED. The result is a microscope illumination with a 700 mA LED, which is twice as much as before. The power supply now is a 5 V USB power supply. As an example, here the conversion of an Olympus CH-2.




Differences to the 350 mW version
The LED is similar to the 1 W version. The voltage is the same, but the current now is 700 mA instead of 350 mA. The lens is also identical.
The circuit diagram of the current stabilizer remains the same. Only the value of resistor R3has to be changed due to the higher current. Also a heat sink for T2 and the LED is advisable now.
The calculation for the new version is easy. The LED voltage is the same (≈3.4 V) but the LED current is twice as much. That means, also the current through T2 and R3 will be two-times bigger. Because the voltage drop across R3 is still 0.7 V, the resistor itself has to be two-times smaller then. That is easy to do, I simply take two of the resistors I still have in parallel. That halves the resistance and doubles the wattage, what is also necessary.
The BD237 as T2 is still fine (2 A) but I took a BD243 (6 A) just because I have many of them. This time the transistor definitely needs a heat sink because the power loss across the transistor is twice as much and it gets really warm.

The original wire potentiometer is removed and the new one is mounted in the same place so that the original dial can be used. The user will not feel a difference.

The mounting plate is also used as a heat sink for T2.

This applies in particular to the LED. An LED does not get hot as a halogen bulb but still there is quite some heat that has to be conducted away. Make a test run. When the heat sink gets so hot that you can not touch it any more, take a bigger one. Do not forget the mounting insulation and thermal paste.
I took quite a big transistor heat sink. That is bigger than needed but in case of the Olympus it fits perfectly and later alignments of the LED were not necessary.

The new LED needs a proper heat sink.

The measurements for the bore holes have to be made very carefully and precise in order to make sure that the focus is in really in the centre. A simple brightness test as described below can be very helpful.

In the end you have to take anyway a heatsink or a piece of metal which fits into your microscope and positions the LED perfectly. Therefore some metalwork will be probably necessary. Make a test with the heatsink in place before you drill the holes for the fixing screws.



The power supply
To keep the power loss small, the input voltage has to be as small as possible. 5 V or 6 V are perfect. The power loss is small and T2 has still 'margin' to control.

The heat (power loss) across T2 depends on the voltage drop of CE. If we use a 5 V power supply the voltage drop is just 1.6 V (5 V-3.4 VLED). Multiplied by the current of 700 mA makes just 1.1 W of heat.
But if you would use a 12 V power supply the power loss of the transistor increases to 9.6 W!
This time I used a 5 V USB power supply which I installed in the microscope housing.



Testing
A simple way of measuring changes in the brightness: A LDR glued on a microscope slide.

At the end a brief comment on testing the brightness.
I do not have a measurement instrument for the brightness. But what I have is a LDR which I glued with epoxy glue in the centre of a microscope slide and connected it to an ohmmeter. With this method I do not get an absolute measurement value of the brightness but I can see changes and can compare values. That is enough to align the LED and compare the brightness with the original microscope lamp.
So, my non-scientific measurement result in case of the Olympus CH-2 is: Wow! It's much, much brighter than the original...

本帖子中包含更多资源

您需要 登录 才可以下载或查看,没有账号?立即注册

×
看贴要回是本分,有问必答是人才,解决问题回贴是公德.
医疗设备维修.维修咨询(请尽可能在论坛提问),协助维修,上门服务.
电话:13991827712

yeec维修网视频培训资料购买链接
BeckmanCoulter DXA系列培训资料
Ortho VITROS 系列培训资料
Ortho enGen_ThermoFisher TCA 实验室自动化系统培训资料
Roche Cobas 实验室自动化系统培训资料
Roche Cobas modular系列分析仪培训资料
Horiba-ABX Yumizen系列培训资料
DiaSorin Liaison系列培训资料
Advia2120培训资料
Inpeco-Aptio系列培训资料
Atellica Solution系列培训资料
Siemens Immunoassay系列培训资料 西门子化学发光系列
SIEMENS Advia系列培训资料 西门子生化系列
Toshiba/Abbott系列培训资料 东芝雅培生化系列
Abbott Architect 系列培训资料 雅培生化化学发光系列
ACL TOP 系列培训资料 沃芬TOP血凝系列
BeckmanCoulter Immunoassay系列培训资料 贝克曼化学发光系列
BeckmanCoulter DXH 系列培训资料 贝克曼DXH血球系列
BeckmanCoulter自动样品处理系统介绍性培训资料 贝克曼前后处理流水线系列
BeckmanCoulter AU系列培训资料 贝克曼AU生化系列
BeckmanCoulter DXC系列培训资料 贝克曼DXC生化系列
LaboSpect003/008/AS 7100/7180分析仪培训资料
Horiba-ABX系列培训资料 Horiba-ABX血球系列
Sysmex 血凝系列培训(CA/CS)
Sysmex 尿液分析系列培训(UF1000/5000/UC3500)
Sysmex 血球系列培训(KX21/POCH/XS/XT/XE)
Sysmex XN系列培训(XN-L/XN1000/XN2000/XN3000/XN9000)
Sysmex HISCL系列培训
可直接淘宝店铺购买https://yeec.taobao.com,或咨询手机/微信:13991827712,QQ:67708237
 

 楼主| 郑振寰 发表于 2016-3-23 13:36 | 显示全部楼层
这也是德国同行的DIY,将显微镜的灯泡改为LED。
他上面所讲的LED价格贵的惊人,10欧元,还是1W的。在淘宝上,3w5瓦的而且带光罩的才5块钱。
他的思路是:
原来显微镜的灯泡是经过变压器直接接卤素灯。把变压器和卤素灯去掉,变压器换成手机充电器,正负极接自制的调光电路再控制LED。
LED寿命长,亮度高。
之前他改造的是老式的显微镜,15年改造了奥林巴斯的CH-2,说明他的思路是经得起考验的。
上面给出了调光电路图,两个三极管组成的恒流调光电路,其中T2需要接散热片,LED也需要接散热片。具体的电阻电容选择上面有计算公式。
在CH-2改造中可以清晰的看到,左边是原来灯泡的电路,右边是改造后的电路。

其实,也不必这么复杂,借助原来的变压器,直接接一个全桥整流器和滤波电容,然后接Lm317进行电流或电压调整,不必拆一个手机充电器,电路还简单的许多。
要注意的是,LED模组在淘宝上购买的时候,1W大约是350mA,3W大约是1A,5W大约是1.3A,无论是手机充电器还是LM317的选择都要注意电流,不同规格的LM317和充电器电流有限制,0.5A-2A不等(LM317最大只能到1.5A)。
在淘宝上购买大功率LED,都有发光颜色的选择,也有功率的选择,还有透镜的选择。一般选择白色光、3W带透镜的。5W太亮,调节起来比较困难。
还要注意的是,淘宝上的LED大多数供电电压在2.3-3V 之间,汽车上的鹰眼是为了配合电瓶而定制的12V,是个例外。所以在进行调光电路设计的时候,要注意最终的电压不能超过,否则会烧毁LED。其实就是分压电阻的选择而已,相信都会计算。
还有一点,淘宝上的LED都可以选择带铝基板散热装置,一个配套的铝基板也就块八毛的,整个配套下来不过5、6块钱。加上调压电路、整流或手机充电器,整套费用不到50元。
看贴要回是本分,有问必答是人才,解决问题回贴是公德.
医疗设备维修.维修咨询(请尽可能在论坛提问),协助维修,上门服务.
电话:13991827712
东海风云 发表于 2016-4-9 20:38 | 显示全部楼层
谢谢分享!
周四平 发表于 2016-5-9 14:13 | 显示全部楼层
值得学习和期待
您需要登录后才可以回帖 登录 | 立即注册

本版积分规则

QQ|申请友链|手机版|小黑屋|加入QQ群|yeec维修网

GMT+8, 2026-6-17 20:23 , Processed in 0.792536 second(s), 33 queries .

Powered by Discuz! X3.5

© 2001-2026 Discuz! Team.

快速回复 返回顶部 返回列表