找回密码
 立即注册
搜索
yeec近年来原创帖合集 本站基础知识下载汇总 yeec网站学习币充值链接 学习中心正式上线

CX3 阻挡探测硬件安装

[复制链接]
郑振寰 发表于 2005-10-30 21:50 | 显示全部楼层 |阅读模式

注册登录才能更好的浏览或提问。

您需要 登录 才可以下载或查看,没有账号?立即注册

×

*** Submitted for Approval ***

CX3 Obstruction Detection Hardware Installation

Last Edited 05/01/2001 07:25:52 PM

Product: Synchron CX

Chapter: CX PRO

Section: CX PRO Upgrade Kit - Installation Instructions

Summary

The purpose of this document is to coach the installation of the Obstruction Detection and Correction (ODC) hardware on the CX3 side of a CX7 or CX9.

Kit Components

CX3 Obstruction Detection Hardware - For CX7 or CX9 Systems - These parts are found in the CX7 Upgrade Kit (p/n 476892).

Part No. Description Quantity

456162 Biohazard Label 1

475505 ODC transducer 1

475523 Tube - CX3 Sample Probe 1

475526 Tube - E-CAM 1

475460 CX3 Sample Wheel Cover 1

475487 MMC - CX3 1

977230 Vial - Bag of 12 1

977421 Fitting Adapter 2

Special Notes

Once the hardware is installed, be sure to enable the ODC feature via software.

Calibration of the ODC transducer must also be done after enabling the feature.

Installation Instructions

This procedure presumes the CX4 ODC hardware and drain tube are installed on the system. If the CX4 ODC hardware is not installed, click here to go to the installation instructions for the CX4 ODC hardware ==> .

Step

Action

1

Remove the CX3 sample wheel cover.

2

Install new CX3 Sample Wheel cover (p/n 475460). The cover is keyed so it can be installed one way. Place the biohazard label (p/n 456162) on the cover to the left of the opening for the vial.

3

Install one vial from the bag of vials (p/n 977230) into the round opening in the cover. This vial is the clot detection spit cup.

NOTE

Vial should not have a lid on it.

4

Shutdown the instrument through the System Idle screen.

5

Remove the CX3 MMC board. Install the new CX3 MMC board (p/n 475487).

6

Locate the allen head screws on the right side of the CX3 frame. Remove the middle allen head screw and retain for reinstallation.

7

Install the ODC assembly so the elbow fitting is away from the wall. Use the screw removed in step 7 to secure the assembly to the frame.

8

The ODC cable can be routed to the board by one of two methods.

Method 1 involves routing the cable through an opening at the rear of the CX3 card cage. Some engineers may have problems with this method and should use Method 2.

Method 2 involves removing the top right rear panel to provide clearance for routing the tube.

The following picture illustrates how the cable is routed and attached to the board regardless of the method used for routing the cable.

Method 1 for Routing CX3 ODC cable to board

8A

Look along the right wall of the CX3 side of the instrument. Locate the opening at the rear of the wall.

8B

Route the CX3 ODC cable through this opening. Insert the plug in the connector on the bottom of the CX3 MMC board.

Method 2 for Routing CX3 ODC cable to board

8AA

Move the instrument away from the wall and remove the top right panel to gain access to the CX3 card cage.

8BB

Route the CX3 ODC cable out the back of the instrument and up to the card cage. Insert the plug in the connector on the bottom of the CX3 MMC board.

8CC

Reinstall the rear panel and put the instrument back in its original location.

9

Remove line 1 from the sample probe. Locate the end of line 1 which is on the right side of the E-cam. Disconnect the tubing at the fitting. Remove line 1 (including fitting) from e-cam to the sample probe.

10

Install one adapter fitting (p/n 977421) to the end of the E-cam tube (p/n 475526) which doesn't have the barbed fitting. Install the adapter end into the port on the top of the CX3 ODC assembly. This fitting will be finger tight. Connect the other end of the E-cam tube (end with barb fitting) to the end of line 1 coming out of the E-cam.

Do not use a wrench to tighten the adapter fitting to the ODC assembly.

11

Install one adapter fitting (p/n 977421) to either end of the CX3 Sample probe tube (p/n 475523). Install the adapter fitting in the front port of the ODC assembly. This fitting will be finger tight. Route the other end of the tube to the CX3 sample probe and attach tubing to the CX3 sample probe.

12

Locate the tee in the drain tube on the CX4 ODC assembly. Attach the tube with the luer lock fitting to the drain fitting on the side of the CX3 ODC assembly. Route the tube so the CX3 sample probe movement is not affected.

13

Return to Upgrade Kit Installation Instructions ==>

看贴要回是本分,有问必答是人才,解决问题回贴是公德.
医疗设备维修.维修咨询(请尽可能在论坛提问),协助维修,上门服务.
电话:13991827712

yeec维修网视频培训资料购买链接
BeckmanCoulter DXA系列培训资料
Ortho VITROS 系列培训资料
Ortho enGen_ThermoFisher TCA 实验室自动化系统培训资料
Roche Cobas 实验室自动化系统培训资料
Roche Cobas modular系列分析仪培训资料
Horiba-ABX Yumizen系列培训资料
DiaSorin Liaison系列培训资料
Advia2120培训资料
Inpeco-Aptio系列培训资料
Atellica Solution系列培训资料
Siemens Immunoassay系列培训资料 西门子化学发光系列
SIEMENS Advia系列培训资料 西门子生化系列
Toshiba/Abbott系列培训资料 东芝雅培生化系列
Abbott Architect 系列培训资料 雅培生化化学发光系列
ACL TOP 系列培训资料 沃芬TOP血凝系列
BeckmanCoulter Immunoassay系列培训资料 贝克曼化学发光系列
BeckmanCoulter DXH 系列培训资料 贝克曼DXH血球系列
BeckmanCoulter自动样品处理系统介绍性培训资料 贝克曼前后处理流水线系列
BeckmanCoulter AU系列培训资料 贝克曼AU生化系列
BeckmanCoulter DXC系列培训资料 贝克曼DXC生化系列
LaboSpect003/008/AS 7100/7180分析仪培训资料
Horiba-ABX系列培训资料 Horiba-ABX血球系列
Sysmex 血凝系列培训(CA/CS)
Sysmex 尿液分析系列培训(UF1000/5000/UC3500)
Sysmex 血球系列培训(KX21/POCH/XS/XT/XE)
Sysmex XN系列培训(XN-L/XN1000/XN2000/XN3000/XN9000)
Sysmex HISCL系列培训
可直接淘宝店铺购买https://yeec.taobao.com,或咨询手机/微信:13991827712,QQ:67708237
 

广东陈威 发表于 2012-8-22 08:37 | 显示全部楼层
学习,谢谢!
您需要登录后才可以回帖 登录 | 立即注册

本版积分规则

QQ|申请友链|手机版|小黑屋|加入QQ群|yeec维修网

GMT+8, 2025-4-5 04:57 , Processed in 1.123087 second(s), 34 queries .

Powered by Discuz! X3.5

© 2001-2025 Discuz! Team.

快速回复 返回顶部 返回列表