找回密码
 立即注册
搜索
yeec近年来原创帖合集 本站基础知识下载汇总 yeec网站学习币充值链接 学习中心正式上线

[转发]RT-PCR操作详细步骤

[复制链接]
郑振寰 发表于 2006-3-16 09:53 | 显示全部楼层 |阅读模式

注册登录才能更好的浏览或提问。

您需要 登录 才可以下载或查看,没有账号?立即注册

×

Protocol for Real-Time RT-PCR


This protocol describes the detailed experimental procedure for real-time RT-PCR using SYBR Green I as mentioned in Xiaowei Wang and Brian Seed (2003) A PCR primer bank for quantitative gene expression analysis. Nucleic Acids Research 31(24): e154; pp.1-8. Please refer to this paper and the PrimerBank Help page for more background information. This protocol is also available in Microsoft Word format here. Here is a one-page succinct protocol. The procedure begins with reverse transcription of total RNA. The cDNA is then used as template for real-time PCR with gene specific primers. You may need to modify this protocol if you use different reagents or instruments for real-time PCR.


Time required

cDNA synthesis: 2 hours.
real-time PCR: 2 hours.
Dissociation curve analysis: 0.5 hour.

Reagents and Equipments

Oligonucleotide Primers. Gene specific primers are retrieved from PrimerBank (https://pga.mgh.harvard.edu/primerbank/). These primers are ordered from the MGH DNA Core facility (https://dnacore.mgh.harvard.edu/synthesis/index.shtml). All the primers are desalted and both UV absorbance and capillary electrophoresis are used to assess the quality of primer synthesis.
Mouse total liver RNA (Stratagene).
Mouse total RNA master panel (BD Biosciences / Clontech).
SYBR Green PCR master mix, 200 reactions (Applied Biosystems).
Optical tube and cap strips (Applied Biosystems).
SuperScript First-Strand Synthesis System for RT-PCR (Invitrogen).
25 bp DNA ladder (Invitrogen).
ABI Prism 7000 Sequence Detection System (Applied Biosystems).
ABI Prism 7000 SDS software (Applied Biosystems).
3% ReadyAgarose Precast Gel (Bio-Rad).
Agarose gel electrophoresis apparatus (Bio-Rad).

Detailed procedure

Reverse Transcription

Reverse Transcription is carried out with the SuperScript First-Strand Synthesis System for RT-PCR. The following procedure is based on Invitrogen’s protocol.

1. Prepare the following RNA/primer mixture in each tube:

Total RNA 5 mg
random hexamers (50 ng/ml) 3 ml
10 mM dNTP mix 1 ml
DEPC H2O to 10 ml

2. Incubate the samples at 65°C for 5 min and then on ice for at least 1 min.
3. Prepare reaction master mixture. For each reaction:

10x RT buffer 2 ml
25 mM MgCl2 4 ml
0.1 M DTT 2 ml
RNAaseOUT 1 ml

4. Add the reaction mixture to the RNA/primer mixture, mix briefly, and then place at room temperature for 2 min.
5. Add 1 ml (50 units) of SuperScript II RT to each tube, mix and incubate at 25°C for 10 min.
6. Incubate the tubes at 42°C for 50 min, heat inactivate at 70°C for 15 min, and then chill on ice.
7. Add 1 ml RNase H and incubate at 37°C for 20 min.
8. Store the 1st strand cDNA at -20°C until use for real-time PCR.

Real-time PCR

1. Normalize the primer concentrations and mix gene-specific forward and reverse primer pair. Each primer (forward or reverse) concentration in the mixture is 5 pmol/ml.
2. Set up the experiment and the following PCR program on ABI Prism SDS 7000. Do not click on the dissociation protocol if you want to check the PCR result by agarose gel. Save a copy of the setup file and delete all PCR cycles (used for later dissociation curve analysis). Please note the extension steps are slightly different from described in our paper.

1) 50°C 2 min, 1 cycle
2) 95°C 10 min, 1 cycle
3) 95 °C 15 s -> 60 °C 30 s -> 72 °C 30 s, 40 cycles
4) 72°C 10 min, 1 cycle

3. A real-time PCR reaction mixture can be either 50 ml or 25 ml. Prepare the following mixture in each optical tube.

25 ml SYBR Green Mix (2x)
0.5 ml liver cDNA
2 ml primer pair mix (5 pmol/ml each primer)
22.5 ml H2O
OR 12.5 ml SYBR Green Mix (2x)
0.2 ml liver cDNA
1 ml primer pair mix (5 pmol/ml each primer)
11.3 ml H2O

4. After PCR is finished, remove the tubes from the machine. The PCR specificity is examined by 3% agarose gel using 5 ml from each reaction.
5. Put the tubes back in SDS 7000 and perform dissociation curve analysis with the saved copy of the setup file.
6. Analyze the real-time PCR result with the SDS 7000 software. Check to see if there is any bimodal dissociation curve or abnormal amplification plot.

Troubleshooting

Here I listed a few major causes for real-time PCR failures. Please read the PrimerBank Help page for more details.

Little or no PCR product. Poor quality of PCR templates, primers, or reagents may lead to PCR failures. First, please include appropriate PCR controls to eliminate these possibilities. Some genes are expressed transiently or only in certain tissues. In our experience, this is the most likely cause for negative PCR results. Please read literature for the gene expression patterns. One caveat is that microarrays are not always reliable at measuring gene expressions. After switching to the appropriate templates, we obtained positive PCR results in contrast to the otherwise negative PCRs (see our paper for more details).

Poor PCR amplification efficiency. The accuracy of real-time PCR is highly dependent on PCR efficiency. A reasonable efficiency should be at least 80%. Poor primer quality is the leading cause for poor PCR efficiency. In this case, the PCR amplification curve usually reaches plateau early and the final fluorescence intensity is significantly lower than that of most other PCRs. This problem may be solved with re-synthesized primers.

Primer dimer. Primer dimer may be occasionally observed if the gene expression level is very low. If this is the case, increasing the template amount may help eliminate the primer dimer formation.

Multiple bands on gel or multiple peaks in the melting curve. Agarose gel electrophoresis or melting curve analysis may not always reliably measure PCR specificity. From our experience, bimodal melting curves are sometimes observed for long amplicons (> 200 bp) even when the PCRs are specific. The observed heterogeneity in melting temperature is due to internal sequence inhomogeneity (e.g. independently melting blocks of high and low GC content) rather than non-specific amplicon. On the other hand, for short amplicons (< 150 bp) very weak (and fussy) bands migrating ahead of the major specific bands are sometimes observed on agarose gel. These weak bands are super-structured or single-stranded version of the specific amplicons in equilibrium state and therefore should be considered specific. Although gel electrophoresis or melting curve analysis alone may not be 100% reliable, the combination of both can always reveal PCR specificity in our experience.

Non-specific amplicons. Non-specific amplicons, identified by both gel electrophoresis and melting curve analysis, give misleading real-time PCR result. To avoid this problem, please make sure to perform hot-start PCR and use at least 60°C annealing temperature. We noticed not all hot-start Taq polymerases are equally efficient at suppressing polymerase activity during sample setup. The SYBR Green PCR master mix described here always gives us satisfactory results. If the non-specific amplicon is persistent, you have to choose a different primer pair for the gene of interest. You are also encouraged to report bad primers to Xiaowei Wang (xwang@molbio.mgh.harvard.edu).

看贴要回是本分,有问必答是人才,解决问题回贴是公德.
医疗设备维修.维修咨询(请尽可能在论坛提问),协助维修,上门服务.
电话:13991827712

yeec维修网视频培训资料购买链接
BeckmanCoulter DXA系列培训资料
Ortho VITROS 系列培训资料
Ortho enGen_ThermoFisher TCA 实验室自动化系统培训资料
Roche Cobas 实验室自动化系统培训资料
Roche Cobas modular系列分析仪培训资料
Horiba-ABX Yumizen系列培训资料
DiaSorin Liaison系列培训资料
Advia2120培训资料
Inpeco-Aptio系列培训资料
Atellica Solution系列培训资料
Siemens Immunoassay系列培训资料 西门子化学发光系列
SIEMENS Advia系列培训资料 西门子生化系列
Toshiba/Abbott系列培训资料 东芝雅培生化系列
Abbott Architect 系列培训资料 雅培生化化学发光系列
ACL TOP 系列培训资料 沃芬TOP血凝系列
BeckmanCoulter Immunoassay系列培训资料 贝克曼化学发光系列
BeckmanCoulter DXH 系列培训资料 贝克曼DXH血球系列
BeckmanCoulter自动样品处理系统介绍性培训资料 贝克曼前后处理流水线系列
BeckmanCoulter AU系列培训资料 贝克曼AU生化系列
BeckmanCoulter DXC系列培训资料 贝克曼DXC生化系列
LaboSpect003/008/AS 7100/7180分析仪培训资料
Horiba-ABX系列培训资料 Horiba-ABX血球系列
Sysmex 血凝系列培训(CA/CS)
Sysmex 尿液分析系列培训(UF1000/5000/UC3500)
Sysmex 血球系列培训(KX21/POCH/XS/XT/XE)
Sysmex XN系列培训(XN-L/XN1000/XN2000/XN3000/XN9000)
Sysmex HISCL系列培训
可直接淘宝店铺购买https://yeec.taobao.com,或咨询手机/微信:13991827712,QQ:67708237
 

您需要登录后才可以回帖 登录 | 立即注册

本版积分规则

QQ|申请友链|手机版|小黑屋|加入QQ群|yeec维修网

GMT+8, 2026-9-16 10:31 , Processed in 0.284223 second(s), 31 queries .

Powered by Discuz! X3.5

© 2001-2026 Discuz! Team.

快速回复 返回顶部 返回列表