WBC的阈值是不可调的,现将维修手册上的一段描述摘抄下来以供参考:
wbc mode
1.run a sample of wbc mode calibrator through the probe as a patient
2.press ALT+D to display the dianosyics screen .page down to view the wbc mode 2.the wbc mode 2 value must match the assayed value with a tolerance of +4.168/-0.0,if adjustment is required.
3.rerun the wbc mode calibrator and adjust R153 on the discriminator board to match the assayed value on the bottle of wbc mode calibrator. |